OK, now on to Probe for EPMA.   We think the implementation of the ferrous/ferric calculation (from Droop, 1987) is pretty good, but please download the latest version (12.7.1) and let us know what you think.  The CalcZAF implementation is described here:

https://probesoftware.com/smf/index.php?topic=691.msg8592#msg8592

So let's take an example of hematite. Calculating all iron as FeO we get these results:

Un    3 Hematite, Results in Elemental Weight Percents

ELEM:       Fe       O
TYPE:     ANAL    CALC
BGDS:      LIN
TIME:    10.00     ---
BEAM:    30.00     ---

ELEM:       Fe       O   SUM
    29  68.992  19.766  88.758
    30  69.126  19.805  88.931
    31  68.758  19.699  88.457

AVER:   68.959  19.757  88.715
SDEV:     .186    .053    .240
SERR:     .108    .031
%RSD:      .27     .27
STDS:       39     ---

STKF:    .6810     ---
STCT:   228.06     ---

UNKF:    .6566     ---
UNCT:   219.89     ---
UNBG:      .48     ---

ZCOR:   1.0503     ---
KRAW:    .9642     ---
PKBG:   456.58     ---

Un    3 Hematite, Results in Oxide Weight Percents

ELEM:      FeO       O   SUM
    29  88.758    .000  88.758
    30  88.931    .000  88.931
    31  88.457    .000  88.457

AVER:   88.715    .000  88.715
SDEV:     .240    .000    .240
SERR:     .138    .000
%RSD:      .27  -91.65
STDS:       39     ---

Note the low totals (~88%) due to the missing ferric oxygen.  Now we specify the ferrous/ferric calculation and the mineral cations and oxygen (2 and 3 respectively for hematites) as seen here:



and re-calculating we now obtain these results:

Un    3 Hematite, Results in Elemental Weight Percents
 
ELEM:       Fe       O
TYPE:     ANAL    CALC
BGDS:      LIN
TIME:    10.00     ---
BEAM:    30.00     ---

ELEM:       Fe       O   SUM
    29  70.168  30.155 100.323
    30  70.305  30.213 100.518
    31  69.930  30.052  99.982

AVER:   70.134  30.140 100.274
SDEV:     .190    .081    .271
SERR:     .109    .047
%RSD:      .27     .27
STDS:       39     ---

STKF:    .6810     ---
STCT:   228.06     ---

UNKF:    .6566     ---
UNCT:   219.89     ---
UNBG:      .48     ---

ZCOR:   1.0682     ---
KRAW:    .9642     ---
PKBG:   456.58     ---

Ferrous/Ferric Calculation Results:
        Ferric/TotalFe   FeO        Fe2O3    Oxygen from Fe2O3
    29       1.000        .000     100.323      10.052
    30       1.000        .000     100.518      10.071
    31       1.000        .000      99.982      10.017

AVER:        1.000        .000     100.274      10.047

Un    3 Hematite, Results in Oxide Weight Percents

ELEM:      FeO       O   SUM
    29  90.271  10.052 100.323
    30  90.447  10.071 100.518
    31  89.965  10.017  99.982

AVER:   90.228  10.047 100.274
SDEV:     .244    .027    .271
SERR:     .141    .016
%RSD:      .27     .27
STDS:       39     ---

Our totals are now much better but because we included this excess oxygen in the matrix correction as suggested by Brian Joy, even more impressive is that our Fe went from 68.959 to 70.134 wt%!

We also added this ferrous/ferric output to the User Specified Output options as seen here:



so now these results can be output to Excel as usual:

« Last Edit: August 19, 2019, 05:56:38 pm by John Donovan »
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Probeman

  • Emeritus
  • *****
  • Posts: 1975
  • Never sleeps...
    • John Donovan
Re: Specifying Unanalyzed Elements For a Proper Matrix Correction
« Reply #106 on: September 11, 2019, 02:38:22 pm »
Here's a "real world" example of a titanium magnetite with the ferrous/ferric excess oxygen calculation in Probe for EPMA using the method of Droop (1987) to calculate the excess oxygen from ferric iron. This excess oxygen is then added into the matrix correction for improved accuracy. The output below also includes oxides and sum of cations also as specified in the Calculation Options dialog.

Un   30 8400 Mgt Trav In-Out
TakeOff = 40.0  KiloVolt = 15.0  Beam Current = 50.0  Beam Size =    0
(Magnification (analytical) =  20000),        Beam Mode = Analog  Spot
(Magnification (default) =     1000, Magnification (imaging) =    100)
Image Shift (X,Y):                                       -2.00,   3.00

Formula Based on Sum of Cations = 3.00   Oxygen Calc. by Stoichiometry
Number of Data Lines:   6             Number of 'Good' Data Lines:   6
First/Last Date-Time: 07/23/2019 02:12:20 PM to 07/23/2019 02:29:56 PM

Average Total Oxygen:       28.684     Average Total Weight%:  100.294
Average Calculated Oxygen:  28.684     Average Atomic Number:   20.135
Average Excess Oxygen:        .000     Average Atomic Weight:   31.696
Average ZAF Iteration:        3.00     Average Quant Iterate:     4.00

Oxygen Calculated by Cation Stoichiometry and Included in the Matrix Correction
Excess Oxygen From Ferric Iron Calculated and Included in the Matrix Correction
Charge Balance Method of Droop (1987), Total Cations= 3.00, Total Oxygens= 4.00

Un   30 8400 Mgt Trav In-Out, Results in Elemental Weight Percents
 
ELEM:       Fe      Mg      Si      Ti       V      Mn      Cr      Al       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    CALC
BGDS:      MAN     MAN     MAN     MAN     LIN     LIN     MAN     MAN
TIME:    40.00  140.00   40.00   80.00   30.00   30.00   88.00  141.00     ---
BEAM:    50.32   50.32   50.32   50.32   50.32   50.32   50.32   50.32     ---

ELEM:       Fe      Mg      Si      Ti       V      Mn      Cr      Al       O   SUM 
    72  60.751   1.862    .019   6.903    .309    .403    .054   1.270  28.664 100.234
    73  60.644   1.863    .075   6.872    .333    .407    .059   1.258  28.657 100.170
    74  60.946   1.871    .021   6.884    .295    .405    .056   1.255  28.732 100.464
    75  60.505   1.856    .028   6.918    .323    .420    .056   1.260  28.584  99.950
    76  60.954   1.876    .016   6.951    .335    .422    .043   1.251  28.776 100.623
    77  60.805   1.858    .023   6.949    .301    .404    .044   1.249  28.693 100.325

AVER:   60.768   1.864    .030   6.913    .316    .410    .052   1.257  28.684 100.294
SDEV:     .175    .008    .022    .032    .017    .008    .007    .008    .066    .235
SERR:     .071    .003    .009    .013    .007    .003    .003    .003    .027
%RSD:      .29     .41   73.90     .47    5.40    2.06   12.65     .61     .23
STDS:      395      12      14      22      23      25      24     396     ---

STKF:    .6779   .4736   .4101   .5547   .6328   .7341   .6400   .0469     ---
STCT:   6812.2 53482.2  7089.5 14540.1 13304.5 22081.3  4697.4  3264.8     ---

UNKF:    .5649   .0085   .0002   .0703   .0033   .0038   .0006   .0072     ---
UNCT:   5676.3   964.8     3.7  1841.7    69.6   114.7     4.3   503.0     ---
UNBG:     20.5    79.1     5.5    52.0    25.2    49.5     9.6    48.0     ---

ZCOR:   1.0758  2.1821  1.4034   .9839   .9547  1.0761   .8885  1.7401     ---
KRAW:    .8333   .0180   .0005   .1267   .0052   .0052   .0009   .1541     ---
PKBG:   277.35   13.19    1.68   36.42    3.77    3.32    1.45   11.48     ---
INT%:      .00    ----    ----    ----  -17.39    -.03  -10.90     .00     ---

Ferrous/Ferric Calculation Results:
        Ferric/TotalFe   FeO        Fe2O3    Oxygen from Fe2O3 
    72        .458      42.390      39.749       3.983
    73        .455      42.482      39.494       3.957
    74        .460      42.324      40.100       4.018
    75        .456      42.333      39.459       3.954
    76        .458      42.468      39.953       4.003
    77        .458      42.388      39.828       3.990

AVER:         .458      42.397      39.764       3.984

Un   30 8400 Mgt Trav In-Out, Results in Oxide Weight Percents

ELEM:      FeO     MgO    SiO2    TiO2    V2O3     MnO   Cr2O3   Al2O3       O   SUM 
    72  78.156   3.087    .041  11.515    .454    .520    .079   2.400   3.983 100.234
    73  78.019   3.090    .161  11.464    .491    .526    .086   2.377   3.957 100.170
    74  78.407   3.103    .044  11.484    .433    .523    .081   2.371   4.018 100.464
    75  77.839   3.078    .059  11.539    .476    .542    .082   2.382   3.954  99.950
    76  78.418   3.110    .034  11.594    .492    .545    .063   2.363   4.003 100.623
    77  78.225   3.080    .050  11.591    .442    .522    .065   2.359   3.990 100.325

AVER:   78.177   3.091    .065  11.531    .465    .530    .076   2.375   3.984 100.294
SDEV:     .225    .013    .048    .054    .025    .011    .010    .015    .025    .235
SERR:     .092    .005    .020    .022    .010    .004    .004    .006    .010
%RSD:      .29     .41   73.90     .47    5.40    2.06   12.65     .61     .64
STDS:      395      12      14      22      23      25      24     396     ---

Un   30 8400 Mgt Trav In-Out, Results Based on Sum of 3 Cations

ELEM:       Fe      Mg      Si      Ti       V      Mn      Cr      Al       O   SUM 
    72   2.381    .168    .001    .315    .013    .016    .002    .103   3.921   6.921
    73   2.377    .168    .006    .314    .014    .016    .002    .102   3.921   6.921
    74   2.384    .168    .002    .314    .013    .016    .002    .102   3.922   6.922
    75   2.378    .168    .002    .317    .014    .017    .002    .103   3.921   6.921
    76   2.380    .168    .001    .316    .014    .017    .002    .101   3.922   6.922
    77   2.382    .167    .002    .317    .013    .016    .002    .101   3.923   6.923

AVER:    2.380    .168    .002    .316    .014    .016    .002    .102   3.922   6.922
SDEV:     .002    .000    .002    .001    .001    .000    .000    .001    .001    .001
SERR:     .001    .000    .001    .001    .000    .000    .000    .000    .000
%RSD:      .10     .24   73.96     .46    5.41    2.06   12.76     .74     .02

Note that if we did *not* include the calculation of excess oxygen in the matrix correction our oxides and formula would look like this:

Un   30 8400 Mgt Trav In-Out, Results in Oxide Weight Percents

ELEM:      FeO     MgO    SiO2    TiO2    V2O3     MnO   Cr2O3   Al2O3       O   SUM 
    72  77.687   3.107    .037  11.434    .451    .517    .074   2.412    .000  95.718
    73  77.554   3.109    .158  11.383    .487    .523    .080   2.388    .000  95.684
    74  77.933   3.123    .041  11.402    .430    .520    .076   2.383    .000  95.908
    75  77.374   3.098    .056  11.458    .472    .539    .076   2.393    .000  95.467
    76  77.946   3.130    .031  11.512    .489    .542    .058   2.375    .000  96.083
    77  77.755   3.100    .047  11.509    .439    .519    .059   2.371    .000  95.799

AVER:   77.708   3.111    .062  11.450    .461    .527    .071   2.387    .000  95.776
SDEV:     .221    .013    .048    .054    .025    .011    .010    .015    .000    .210
SERR:     .090    .005    .020    .022    .010    .004    .004    .006    .000
%RSD:      .28     .41   78.00     .47    5.41    2.06   13.52     .61 -167.33
STDS:      395      12      14      22      23      25      24     396     ---

Un   30 8400 Mgt Trav In-Out, Results Based on Sum of 3 Cations

ELEM:       Fe      Mg      Si      Ti       V      Mn      Cr      Al       O   SUM 
    72   2.379    .170    .001    .315    .013    .016    .002    .104   3.376   6.376
    73   2.375    .170    .006    .313    .014    .016    .002    .103   3.379   6.379
    74   2.382    .170    .002    .313    .013    .016    .002    .103   3.374   6.374
    75   2.376    .170    .002    .316    .014    .017    .002    .104   3.378   6.378
    76   2.378    .170    .001    .316    .014    .017    .002    .102   3.376   6.376
    77   2.380    .169    .002    .317    .013    .016    .002    .102   3.377   6.377

AVER:    2.378    .170    .002    .315    .014    .016    .002    .103   3.377   6.377
SDEV:     .002    .000    .002    .001    .001    .000    .000    .001    .002    .002
SERR:     .001    .000    .001    .001    .000    .000    .000    .000    .001
%RSD:      .10     .24   78.05     .46    5.41    2.06   13.63     .74     .06

Of course the oxygen formula is low, but note the effect of this "missing oxygen" on the concentrations of the other elements. E.g., without the excess oxygen in the matrix correction, the Fe and Ti concentrations went down, but the Mg and Al concentrations went up!
« Last Edit: September 11, 2019, 03:26:57 pm by Probeman »