As you know, TDI point analyses display a number of TDI fit parameters in Probe for EPMA as shown here:
ELEM: Na Fe Ca Si O SUM
208 14.643 10.321 3.802 28.711 42.282 99.758
209 14.711 10.788 3.838 29.091 42.887 101.315
210 15.123 10.710 3.728 28.991 42.851 101.404
211 13.979 10.394 3.872 28.746 42.140 99.130
AVER: 14.614 10.553 3.810 28.885 42.540 100.402
SDEV: .474 .230 .061 .186 .385 1.136
TDI%: 53.503 -3.738 -2.765 -4.833 --- percent relative TDI correction (what is displayed below in the maps)
DEV%: .4 .7 1.3 .2 --- percent TDI fit deviation
TDIF: HYP-EXP LOG-LIN LOG-LIN LOG-LIN --- TDI fit type
TDIT: 105.75 104.25 105.75 106.00 --- total TDI acqusition time
TDII: 203. 65.1 29.1 427. --- TDI intensity zero intercept (cps per nominal beam)
TDIL: 5.31 4.18 3.37 6.06 --- TDI intensity zero intercept (log cps per nominal beam)
But for TDI scanning data (in CalcImage) we have no equivalent summary of the TDI fit parameters (until now). Then Anette von der Handt pointed out to me that we could at least display the TDI correction for each pixel as a map...

I think it is very cool that we can now see how the "grid" TDI effect in the quant Na and K maps are corrected for on a pixel by pixel basis.
Basically if you are calculating TDI scanning corrections for x-ray maps, CalcImage will now automatically calculate and save TDI correction percent maps as GRD files. One will need to open them (from the File | Open menu), and then utilize the Output Currently Displayed Images menu in CalcImage for output to Surfer.
For these maps, because the host quartz pixels have concentrations of Na and K close to zero, the percent TDI corrections can be very large due to statistics, so for the maps above I set the max Z to values that better show the TDI corrections within the melt inclusion.
But what about Si Ka?

I find it interesting that the TDI correction size for Si ka are similar for both the quartz host and the ryolite melt inclusion- except for the edge artifacts, which are undoubtedly effects from the beam overlapping on both compositions, but why do the TDI corrections in these edge pixels seem to show a spectrometer orientation effect (spectro 4 on the Cameca is oriented in the lower left)?
For example, in the quant maps, Si (in the upper right of the image below) shows distinct edge effects, but no apparent spectrometer orientation effect... or maybe the TDI edge artifact trend (above) is due to the direction of the slow scan which on the Cameca is from top to bottom...

Maybe it's got nothing to do with either the spectrometer orientation or the slow scan direction, but any other ideas?
john