Author Topic: Intermittent Socket Timeout Errors  (Read 6845 times)
Owen Neill
  • Global Moderator
    • Electron Microbeam Analysis Laboratory, University of Michigan
« on: March 01, 2017, 06:16:04 PM »
Dear colleagues,
In recent combined automated EDS/WDS acquisitions on our microprobe, we have been seeing automations intermittently aborted by the following error:
ERROR in ThermoEDSSpectraWriteSystem : Error number  5 returned from Thermo EDS interface (Socket timeout error) when calling function TEVB_PutSpectrum
Description of conditions follows:
- Probe is a JEOL 8500, using PFE (v11.8.3) to acquire WDS data and ThermoNSS (v4.1.85) to acquire EDS data.
- No images are being acquired, just EDS and WDS data. We don't use the Thermo interface for imaging via PFE.
- PFE and NSS are running on separate computers. I can connect to NSS from the PFE computer via the TestEDS app without problems, and outwith the intermittent timeout errors, EDS acquisition via PFE appears to work normally. All DLL's appear to be in the correct folders and function properly.
- This error appears intermittently - sometimes long automations (>12 hours) will complete without aborting due to this error, while at other times the automation will abort after only a couple of points.
- The "Measure beam current before EDS acquisition" box is not checked, as per this thread: http://probesoftware.com/smf/index.php?topic=136.msg2150#msg2150
- Re-booting both the Thermo and PFE computers seems to help prevent the error in post-reboot automations, although I can't say this with certainty.
- Both PFE and Thermo computers are running 64-bit Win7.
I was wondering if any other users had encountered similar problems, and if so, how they dealt with them. Thanks in advance for any info, it's much appreciated.
OKN
  • Global Moderator
Re: Intermittent Socket Timeout Errors
« Reply #1 on: March 01, 2017, 09:46:27 PM »
Owen,
I see what appears to be  the same intermittent error. I have PFE and NSS on the same PC. I used to have two separate PCs, and I'd say I saw the issue more often, but i can't be sure.
Gareth
« Reply #2 on: March 01, 2017, 10:43:15 PM »
Owen,
Gareth
Owen and Gareth: what version of NSS are you using?  I'm using v. 4 of NSS on the same computer with PFE and I'm not seeing any socket timeout issues, though I think I might have seen some a few years ago with an earlier version of NSS.
Owen Neill
  • Global Moderator
« Reply #3 on: March 02, 2017, 08:36:49 AM »
Thank you both for following up. I'm running NSS v4.1.85 on the Thermo side.
OKN
  • Global Moderator
« Reply #4 on: July 14, 2017, 02:57:13 AM »
Was there a conclusion to this topic?
If there was some action taken to potentially minimise the occurrence of time-out errors,  perhaps it could be documented here?
Gareth
« Reply #5 on: July 14, 2017, 08:49:47 AM »
Have you been seeing any timeout issues with the version of NSS that you are currently running?
I've recently started running the new Thermo Pathfinder v. 1.2, but I haven't done any long integrated WDS-EDS acquisitions yet.
« Reply #6 on: July 18, 2017, 04:26:07 PM »
Hi Gareth,
I'm running a 700+ point grid with EDS spectrum acquisition overnight using the latest Thermo Pathfinder 1.2.72 software so we'll see if I get any socket timeouts.
But I ran EDS spectrum acquisition on standards and wavescans all day and did not see anything amiss so far.
By the way, I noticed that Thermo has an auto mode for maximum energy cutoff, so I will add that to the software tonight.
« Reply #7 on: July 19, 2017, 02:37:33 PM »
Hi Gareth,
I ran 500+ WDS points with full spectrum EDS acquisitions using Thermo Pathfinder last night and it was still running fine when I came in this morning.
So either the problem is specific to your network setup, or it's a problem that was fixed in Pathfinder.
« Last Edit: July 19, 2017, 03:35:40 PM by John Donovan »
SteveSeddio
  • Moderator
  • Setting the mood for microanalysis...
« Reply #8 on: July 19, 2017, 03:13:15 PM »
Hi Gareth and Owen,
Could you please confirm that the errors you are seeing are only regarding the calling of the function TEVB_PutSpectrum?
-Steve
kthompson75
  • Post Doc
« Reply #9 on: July 22, 2017, 04:43:46 PM »
Time out errors do occur. Typically they are not caused by the software application that times out. Instead, they are generally system related. We typically allow 10 seconds for a response to be sent and acknowledged after a command is issued. If the system gets busy doing something else, that 10 seconds can elapse pretty quickly. The result is a time-out. These time-outs are necessary because if the system has become truly disconnected, the user will regain control after the time-out.
We can always extend the timeout duration of certain functions if requested.
SteveSeddio
  • Moderator
« Reply #10 on: April 13, 2018, 10:56:47 AM »
These socket timeout issues should no longer be a problem in the Pathfinder 1.3.22 release.
emma_fisi
« Reply #11 on: October 19, 2018, 01:10:01 PM »
I have Pathfinder 1.3 v22 on my 8530F, and I'm having a couple of issues, one of which appears to involve a socket. A couple of times when starting a run, I get the following:
"ERROR in ThermoEDSSpectraStart : Error number  4 returned from Thermo EDS interface (Error writing to socket) when calling function TEVB_EDSAcqRunning"
I restart the run, and the error does not reappear. Just letting you know in case the same error mentioned earlier has resurfaced...
I'll start a new thread for the other issue.
SteveSeddio
  • Moderator
« Reply #12 on: October 22, 2018, 02:59:04 PM »
Emma,
Has the socket error persisted after implementing Probeman's suggestion (in our off-board email thread)?
emma_fisi
« Reply #13 on: October 23, 2018, 10:46:32 AM »
Hi Steve,
The problem has not recurred - I will certainly let you know if it does ;)
Em