Author Topic: Spreadsheet of sin theta values, background interferences  (Read 1431 times)
« on: February 21, 2017, 06:49:29 am »
I've finally adapted the JEOL l value spreadsheet for Cameca sin theta.
It includes correction for reflective index.
Using the excel filter - you can examine the range of interest and it gives relative position of interfering lines.
In the example below - looking at Fe KA1. Elements of atomic number <30, x-rays lines with intensity > 1% and LIF range 47000 to 49000 are shown. Mn Kb1 is seen to occur at -638
« Last Edit: April 12, 2020, 10:05:12 pm by John Donovan »
Re: Spreadsheet of sin theta values, background interferences
« Reply #1 on: March 02, 2017, 01:12:38 pm »
I've just found that I was calculating sin theta wrong for the high order interferences - for I was using n*lamda - but in the xray table (as given in calczaf) the angstrom for high order interferences is the apparent wavelength (i.e. angstrom = n*lamda in table)
So I've corrected the calculation in the spreadsheet