from what I learned so far and talking to more experienced users of "non-traditional" coating materials I am ready to make the plunge (after careful vetting of our new coater, see below).
Besides analysis of carbon itself, carbon has also very high absorbance of O ka, N ka, so there are various other applications for non-carbon coatings. I think the absolutely crucial part for all metal coatings will be how thin as well as how accurate and reproducible you can lay down the coating, especially if you don't always want to coat the standards together with the samples. There is much, much less room for error relative to carbon coatings.
Also, as far as I understand, the attraction of iridium specifically for a lot of people is that there is a specific thickness (0.6 nm??) that gives you the same absorbance as carbon so you could mix and match carbon and iridium coating theoretically (PFE also has a coating thickness correction, btw).
You might want to look further into the work of Kat Crispin and John Armstrong about this or contact Kat. From their abstracts at M&M 2013:
"Results of this study indicate that an Ir-coat of 0.4 nm thickness is sufficient to provide sufficient conductivity on an insulating sample for analysis; although for samples containing beam-sensitive materials, a minimum thickness of 1 nm is necessary to avoid element migration (Table 1). Coatings of less than 2 nm require minimal or no additional offline corrections, making this an ideal substitute for traditional carbon coatings and require no special preparations of standards (i.e. uncoated or carbon coated standards can be directly compared). However, the K-lines of boron to oxygen and the L-lines of calcium to chromium are heavily absorbed by iridium, requiring correction for Ir thicknesses over 1 nm."
(Armstrong & Crispin, 2013, Ultra-thin Iridium as a Replacement Coating for Carbon in High Resolution Quantitative Analyses of Insulating Specimens. Microanal. 19 (Suppl 2), 1070)