PHA is hard to teach.
Hi Ben,
I agree!
My position has been to just teach the users to generally accept the default detector (high voltage) bias value and then only adjust the gain to get the PHA peak around 3-5 volts on the JEOL, or 2-3 volts on the Cameca and leave detector in differential mode with the baseline/window values at 0.5/9.5 volts for JEOL and 0.5/4.5 volts for Cameca. Assuming one is tuning on a high concentration of the element in question.
My reasoning is that if one has a high intensity of the element, the PHA peak can only shift slightly to the right as the intensity drops (e.g., minor/trace elements).
The fact that there might be an escape peak present is OK, as those photons do represent the photons of interest. The idea being that if one can, one should always include the escape peak in the PHA range because that improves one's statistics. Trying to exclude the escape peak will often introduce a non-linear response from the detector as the escape peak is partially removed, due to the shifting of the peak from different count rates.
Also, if one attempts to try and set tight baseline/window values to avoid high order Bragg interferences, again one will often only introduce non-linear responses from the detector, due to the shifting of the peak from different count rates.
I find it better to "let them all in" and then the interference correction sort things out!
The only time I find I actually need to mess with the baseline/window settings is for an interference of say, Na ka II on O ka. Simply because it's difficult to find an interference standard that contains Na, but no oxygen...
Of course, once we all go to pin diode detectors for our WDS spectrometers things may get easier!
Just me two cents.
john