The purpose of simulating WDS polygonization tail artifacts is simply to create a more realistic looking spectrum with WDS spectral overlaps from the long tails we observe in our instruments. So yesterday I decided to test how well these simulated tails could used to demonstrate the quantitative spectral interference corrections in Probe for EPMA simulations.
But first let me mention that the simulated WDS spectra generated in Probe for EPMA are synthesized by summing pre-calculated Penepma spectra for *pure elements* at various electron beam energies:
https://probesoftware.com/smf/index.php?topic=202.msg9589#msg9589This is in contrast to the EDS simulated spectra generated in Probe for EPMA which are created by running Penepma in real time based on the actual compositions. My impression is that running Penepma in real time is roughly equivalent to an actual beam current of a few nano-amps. These simulated EDS spectra can then be quantified (when run long enough for sufficient statistics) for any elements in the spectrum.
Unfortunately we cannot do the same for WDS scans because the scan would vary in statistical significance during the simulated scan "acquisition". So we resort to pre-calculated pure element simulations performed at 1 eV intervals for WDS resolution. In addition these WDS spectra for each analyzed element are re-normalized for the element in question to generate a quantitative intensity for that element, but other secondary emission lines in the spectra from other elements cannot be considered quantitative since currently, no absorption correction is applied to the spectra when they are summed. In other words the pure element spectra are quantitative, but not necessarily the compound spectra.
Anywho, I decided to perform a simulation for Ti, V and Cr using pure element standards and a secondary standard of a Ti, Al, V, Cr alloy (SRM 654b) to see just how bad the interference correction is when performed on these simulated spectra with the newly added polygonization artifacts. For reference the composition of SRM 654b is shown here:
St 654 NBS SRM-654b
TakeOff = 40.0 KiloVolt = 15.0 Density = 5.000 Type = alloy
NBS (NIST), Ti base alloy
Elemental Composition
Average Total Oxygen: .000 Average Total Weight%: 100.000
Average Calculated Oxygen: .000 Average Atomic Number: 21.491
Average Excess Oxygen: .000 Average Atomic Weight: 45.783
ELEM: Ti Al V Fe Sn Cu Ni Cr Si Mo Zr
XRAY: ka ka ka ka la ka ka ka ka la la
ELWT: 88.974 6.340 4.310 .230 .023 .004 .028 .025 .045 .013 .008
KFAC: .8825 .0449 .0420 .0021 .0002 .0000 .0003 .0002 .0004 .0001 .0001
ZCOR: 1.0082 1.4110 1.0270 1.0837 .9563 1.0886 1.0434 1.1243 1.2290 1.0746 1.1283
AT% : 85.041 10.758 3.873 .189 .009 .003 .022 .022 .073 .006 .004
First here is the simulated WDS spectra for V Ka (interfered by Ti Kb):

Note that the 4 wt% V emission is quite interfered by the 89 wt%Ti (Kb) emission. The Cr Ka emission is also somewhat interfered by the vanadium Kb emission as seen here:

An example of what I call a "cascade" interference, where A interferes with B, and B interferes with C, which requires an iterative solution for quantitative results (the situation where two primary analytical emission lines *both* interfere with each other (e.g., Ba La and Ti Ka), I call "self-interfering" or "pathological" interferences, which also requires an iterative solution).
Using instrumental measurements on this same alloy produces excellent results as shown many years ago in this paper (see last line in table 1 on page 26):
https://epmalab.uoregon.edu/publ/Improved%20Interference%20(Micro.%20Anal,%201993).pdfOK, so first is the simulated "analysis" of the alloy *without* an interference correction:
St 654 Set 1 NBS SRM-654b
TakeOff = 40.0 KiloVolt = 15.0 Beam Current = 30.0 Beam Size = 0
(Magnification (analytical) = 20000), Beam Mode = Analog Spot
(Magnification (default) = 600, Magnification (imaging) = 200)
Image Shift (X,Y): .00, .00
NBS (NIST), Ti base alloy
Number of Data Lines: 5 Number of 'Good' Data Lines: 5
First/Last Date-Time: 08/11/2021 09:32:18 AM to 08/11/2021 09:35:22 AM
Average Total Oxygen: .000 Average Total Weight%: 100.360
Average Calculated Oxygen: .000 Average Atomic Number: 21.492
Average Excess Oxygen: .000 Average Atomic Weight: 45.810
Average ZAF Iteration: 2.00 Average Quant Iterate: 2.00
St 654 Set 1 NBS SRM-654b, Results in Elemental Weight Percents
ELEM: Ti V Cr Al Fe
TYPE: ANAL ANAL ANAL SPEC SPEC
BGDS: LIN LIN LIN
TIME: 20.00 20.00 20.00 --- ---
BEAM: 30.01 30.01 30.01 --- ---
ELEM: Ti V Cr Al Fe SUM
116 89.094 4.914 .089 6.340 .230 100.666
117 88.600 5.037 .108 6.340 .230 100.315
118 88.645 4.936 .100 6.340 .230 100.251
119 88.728 5.012 .076 6.340 .230 100.386
120 88.608 4.952 .052 6.340 .230 100.182
AVER: 88.735 4.970 .085 6.340 .230 100.360
SDEV: .207 .052 .022 .000 .000 .187
SERR: .093 .023 .010 .000 .000
%RSD: .23 1.05 25.80 .00 .00
PUBL: 88.974 4.310 .025 6.340 .230 99.879
%VAR: -.27 15.32 239.36 .00 .00
DIFF: -.239 .660 .060 .000 .000
STDS: 522 523 524 --- ---
STKF: .9940 1.0000 .9988 --- ---
STCT: 344.27 340.70 339.74 --- ---
UNKF: .8805 .0484 .0008 --- ---
UNCT: 304.95 16.49 .26 --- ---
UNBG: .49 .53 .94 --- ---
ZCOR: 1.0078 1.0268 1.1233 --- ---
KRAW: .8858 .0484 .0008 --- ---
PKBG: 627.81 31.97 1.28 --- ---Note that the V concentration is high by about 0.5 wt% and the Cr is high by about 160 PPM. Now the same analysis but *with* an interference correction applied:
St 654 Set 1 NBS SRM-654b
TakeOff = 40.0 KiloVolt = 15.0 Beam Current = 30.0 Beam Size = 0
(Magnification (analytical) = 20000), Beam Mode = Analog Spot
(Magnification (default) = 600, Magnification (imaging) = 200)
Image Shift (X,Y): .00, .00
NBS (NIST), Ti base alloy
Number of Data Lines: 5 Number of 'Good' Data Lines: 5
First/Last Date-Time: 08/11/2021 09:32:18 AM to 08/11/2021 09:35:22 AM
Average Total Oxygen: .000 Average Total Weight%: 99.749
Average Calculated Oxygen: .000 Average Atomic Number: 21.482
Average Excess Oxygen: .000 Average Atomic Weight: 45.781
Average ZAF Iteration: 2.00 Average Quant Iterate: 4.00
St 654 Set 1 NBS SRM-654b, Results in Elemental Weight Percents
ELEM: Ti V Cr Al Fe
TYPE: ANAL ANAL ANAL SPEC SPEC
BGDS: LIN LIN LIN
TIME: 20.00 20.00 20.00 --- ---
BEAM: 30.01 30.01 30.01 --- ---
ELEM: Ti V Cr Al Fe SUM
116 89.113 4.287 .084 6.340 .230 100.053
117 88.619 4.413 .103 6.340 .230 99.705
118 88.664 4.312 .095 6.340 .230 99.641
119 88.747 4.387 .072 6.340 .230 99.775
120 88.627 4.328 .047 6.340 .230 99.573
AVER: 88.754 4.345 .080 6.340 .230 99.749
SDEV: .207 .053 .022 .000 .000 .186
SERR: .093 .024 .010 .000 .000
%RSD: .23 1.22 27.23 .00 .00
PUBL: 88.974 4.310 .025 6.340 .230 99.879
%VAR: -.25 .82 221.58 .00 .00
DIFF: -.220 .035 .055 .000 .000
STDS: 522 523 524 --- ---
STKF: .9940 1.0000 .9988 --- ---
STCT: 344.27 340.70 339.74 --- ---
UNKF: .8805 .0423 .0007 --- ---
UNCT: 304.95 14.42 .24 --- ---
UNBG: .49 .53 .94 --- ---
ZCOR: 1.0080 1.0270 1.1242 --- ---
KRAW: .8858 .0423 .0007 --- ---
PKBG: 627.81 28.07 1.26 --- ---
INT%: ---- -12.59 -5.68 --- ---
Now with the interference applied the V concentration is quite good, but the Cr concentration is under corrected and still a bit high. I'm going to say that this first attempt at adding WDS polygonization artifacts in these simulated WDS spectra is pretty good, but not quite quantitative enough at the 1000 PPM level and below. I will continue to evaluate and let you all know what else I find.
In summary I would say these polygonization simulations are good enough for teaching and training, which of course is the whole purpose of the simulation mode in Probe for EPMA. Please feel free to try some WDS simulations of your own, and share your results here in this topic.