Author Topic: Monte carlo simulation of BS coefficient  (Read 4727 times)
« on: April 27, 2016, 01:10:57 PM »
I've just run a quick comparison of the BS coefficient calculated using Win X-ray, casino and penepma for Ag and Fe metal (initially to check upbound fraction in penepma-res.dat was the bs coefficient-but then the results where interesting.)
Results as follows - best agreement Win X-ray/Penepma whereas casino considerable differ. My guess Win X-ray Penepma the best. (Although casino was much faster even than win x-ray at least through wine on linux.)
« Last Edit: July 22, 2020, 05:22:26 AM by Ben Buse »
Re: Monte carlo simulation of BS coefficient
« Reply #1 on: April 27, 2016, 01:27:45 PM »
Excellent!
It would be awesome to compare these calculations to actual measurements, but this is not easy as one has to apply a small bias voltage to an isolated sample to prevent additional signal from secondary electrons.
Attached below is a paper (you must be logged in to see attachments!) where I replicated Heinrich's BSE measurements in my SX51 at UC Berkeley.
« Reply #2 on: April 27, 2016, 03:56:40 PM »
Here is the post with the Au BSE coefficients calculated for different densities attached:
http://probesoftware.com/smf/index.php?topic=126.msg508#msg508
The reason I mention the Au BSE calculations on different densities is because some time ago I found that putting in different densities in Casino caused it to produce significantly different BSE values. I pointed this out to Raynald Gauvin and he went "sacre bleu"!      :D
I assume that this was fixed however, but I haven't tested it.
« Last Edit: April 27, 2016, 04:11:46 PM by Probeman »
« Reply #3 on: April 28, 2016, 03:24:56 AM »
I've just stumbled across a database of experimental data from literature which David Joy compiled for SE and BSE yields (whilst looking for something else - as is the way)
http://web.utk.edu/~srcutk/htm/interact.htm
For Ag values are 0.394 (1); 0.415(2), 0.423(3),0.409(35)
For Fe values are 0.278(1) and 0.298 (13)
numbers in brackets are the references in the report
Which are closest to PENEPMA and Win X-ray. I was using casino V 2.48
hdemers
« Reply #4 on: April 28, 2016, 07:11:52 AM »
For BSE simulation, you should use CASINO v3 and not version 2. Also you need to specify which elastic cross section models you use. In CASINO v3, from a limited testing, I found Mott-Browning model to closer match with the Joy database, but the ELSEPA model should be better. PENELOPE use ELSEPA, but I am not sure which settings they have use to generate the cross section, so it may explain the difference between CASINO ELSEPA and PENELOPE ELSEPA.
Regards,
Hendrix
hdemers
« Reply #5 on: April 28, 2016, 08:35:21 AM »
It have to be an old version of casino. I just try it with casino v2 and v3 and not difference with different densities:
Casino v3 20 kV Au
rho = 1:      0.477801 (0.0015)
rho = 19.3: 0.478413 (0.0015)
Casino v2 20 kV Au
rho = 1:      0.449739 (0.0015)
rho = 19.3: 0.449722 (0.0015)
In estimate 3 sigma variation is in ().
Regards,
Hendrix
« Reply #6 on: April 28, 2016, 09:53:47 AM »
Casino v3 20 kV Au
Casino v2 20 kV Au
Regards,
Hendrix
Hi Hendrix,
Thanks for testing.  Yes, I assumed it was fixed many years ago when I showed it to Raynauld.  I think it was v. 1.x of Casino actually now that I think of it.
Many people seem to think intuitively that density should affect backscatter in bulk samples, but the physics says no!
« Reply #7 on: April 28, 2016, 12:11:54 PM »
Regards,
Hendrix
Thanks that's very helpful