Author Topic: BSE TDI measurements  (Read 3060 times)
« on: April 21, 2016, 07:26:12 AM »
Bit of an unconventional question, but is it possible in PfE to measure the BSE signal as part of a TDI analysis? I'm looking at the amount of ablation I get of a C coating on a heavy substrate during low kV analyses using TDI and am seeing a steady decrease in absorbed current during each analysis - this can be as large as a ~10% drop so it's not insubstantial. The faraday current is stable. I think this decrease may be due to an increasing BSE coefficient as the C is thinned (BSE imaging after analysis shows bright spots surrounded by darker contamination rings).
Mike
Re: BSE TDI measurements
« Reply #1 on: April 21, 2016, 08:21:59 AM »
Mike
Right now PFE only acquires the absorbed current as an option during the TDI acquisition:
http://probesoftware.com/smf/index.php?topic=40.msg4006#msg4006
Also the BSE signal is only acquired in PFE as an image.  Your request sounds pretty specific, so your best bet might be to utilize the Remote interface and write a macro in Excel as described here:
http://probesoftware.com/smf/index.php?board=9.0
This way you get complete control over your acquisition and data processing.
« Reply #2 on: April 22, 2016, 05:03:33 AM »
Thanks John, I knew it was a long shot but it was worth asking.
« Reply #3 on: April 22, 2016, 08:39:01 AM »
It wouldn't hard using the Remote interface.  Just call the get intensity function along with the get image function using a small (say 32 x 32 pixel) image acquisition.  You should be able to get pretty short TDI intervals this way.