I am sorry but unfortunately I see different picture. Could you look at addition file (PictureSnap image vs image in Analog Signal Image Data window)
1. If you load PictureSnap from the Automate window it displays the digitized points from the Automate window (whether they have been acquired or not). These will have the digitized sample number and digitized point number displayed (starting with 1 for each digitized sample).
2. If you load PictureSnap from the Run menu, then it will display points already acquired in the run. These labels are using the acquired sample number and the (consecutive) line number from the run.
3. The Display Analog Signal Image dialog only displays points already acquired in the run. These labels are using the acquired sample number and the (consecutive) line number from the run.
#2 and #3 will display acquired point labels the same way.
If you are asking to have the beam size displayed for acquired points in the PictureSnap window, that is not available because the point labels in PictureSnap needs to be consistent for both digitized points and acquired points, but digitized points do not yet have a beam size defined.