Hi everyone,
Recently a customer asked me to analyze a bunch of Ti-Pt alloys, each being made of 2 to 4 different phases. Aside of the small size of the particle, I quickly realized that different matrix correction procedure yield pretty different results. The Ti and Pt weight-% content can easily vary by 2-3 wt-% each depending on the correction procedure used (and these variations are not covariant: when Ti increase with a matrix correction, the Pt-content sometimes goes up, sometimes goes down). Here is an example in one phase:
Elemental Weight Percents (average of 5 points, each very similar):
ELEM: N O Ti Pt TOTAL
1 .033 .405 32.423 65.961 98.823 Armstrong/Love Scott (default)
2 -.527 .203 30.865 73.592 104.133 Conventional Philibert/Duncumb-Reed
3 -.557 .248 32.523 70.250 102.464 Heinrich/Duncumb-Reed
4 -.376 .292 31.481 67.185 98.582 Love-Scott I
5 -.307 .324 31.757 67.929 99.702 Love-Scott II
6 .150 .298 29.716 71.031 101.195 Packwood Phi(pz) (EPQ-91)
7 .021 .336 33.631 70.589 104.577 Bastin (original) Phi(pz)
8 -.597 .251 31.637 68.291 99.582 Bastin PROZA Phi(pz) (EPQ-91)
9 -.119 .297 31.479 69.211 100.868 Pouchou and Pichoir-Full (Original)
10 -.132 .312 31.777 69.357 101.314 Pouchou and Pichoir-Simplified (XPP)
This can be problematic as the key for this study is to determine the Ti to Pt (atoms) proportion in each phase. This remind me of the Si-Ir example John Armstrong & Paul Carpenter once presented to me, and in their case the "accurate" value was actually obtain... only when using an Si-Ir standard instead of pure Si and pure Ir standards ("classical" example of the use of CalcZAF here - see also Lehigh School of Microscopy!). Unfortunately in my case, I don't have access to a certified / known composition Pt-Ti alloy, and I now wonder which matrix correction is the "best" / most accurate.
And this is only a small part of the iceberg, as a side-kick of this project would have been to analyze nitrogen and oxygen. Nitrogen is (almost) out of the question here, as the count rate is very low, the expected content is low (most being < 1000 ppm), and N Ka interfere with Ti Ll (=> large error, high correction and very high detection limit around 5000-8000 ppm depending on the counting time). Oxygen is another beast, and I got decent count rate. However, we were surprised by the high O wt-% content. While some phases appear poor in O (as expected: samples were prepared in anoxic environment), other shows up to 1 wt-% O and more (example above show ~0.3%).
Therefore three questions to you:
(a) What ZAF/phi-rho-z correction would you recommend for Ti-Pt alloys, assuming only pure Ti and pure Pt standards are available? Analytical conditions involve Ti Ka and Pt Ma, at 15 keV, 100-200 nA (assumes no problem with PHA / high count rate, each standard yield around or less than 100 cps/nA with the selected setup, and almost half less in the unknown => around 10000 to maximum 20000 cps - should be OK?).
(b) Same as (a), but this time if we further consider the measurement of light element (N and O) at these conditions: which matrix correction to choose (curiously enough, Armstrong yield significantly higher O-content: 0.4% vs. ~0.2-0.3 for all others)?
(c) Without commenting on the (very) bad choice of voltage (15 keV) for light element (*), the high oxygen content appears surprising. Is it possible the samples got heavily oxidized? If so, is this possible that some phases got more O-rich than other (in the same sample!)?
Many thanks in advance, and Happy Thanksgiving!
Julien
(*) with my current instrument / ill (?) BSE, I am loosing all necessary contrast to differentiate the different phases in these Pt-Ti alloys.