The Athens 2022 EPMA just concluded and there were questions about how to correct data for modeled secondary fluorescence effects.
This post here provides one example:
https://probesoftware.com/smf/index.php?topic=58.msg5603#msg5603It should be noted however, that ideally one should perform a matrix correction after the concentrations have been modified after this subtraction as the "apparent" concentration of the fluoresced element is of course not actually present in the matrix. However, so long as the subtracted concentration is a fraction of a percent the effect should be minimal, particularly on the measured element itself.
In fact the CalcZAF software actually includes this matrix iteration in the secondary fluorescence correction based on defined distances from the boundary. See here for details:
https://probesoftware.com/smf/index.php?topic=58.msg223#msg223Unfortunately this SF subtraction/iteration GUI not yet implemented in the Probe for EPMA software, though soon perhaps.
But it should also be pointed out that the modeled secondary fluorescence concentrations from Fanal or Penepma are only accurate for EDS detectors (without any Bragg defocusing) and WDS detectors in the two directions parallel to the spectrometer orientation. This is described by Ben Buse in the paper linked to this post here:
https://probesoftware.com/smf/index.php?topic=1447.msg10718#msg10718To avoid this WDS defocusing effect on the secondary fluorescence from a boundary, one should orient the sample so the boundary is parallel to the spectrometer orientation, hence one of the reasons we are trying to document this spectrometer orientation in JEOL and Cameca microprobes.