A new feature PHA acquisition feature is now available for Cameca (SX100 and SXFive) instruments. But first a minor "mea culpa"...
Last year I installed the new Thermo WDS mapping option for my NSS EDS system. Works very well, see here:
http://probesoftware.com/smf/index.php?topic=77.msg284#msg284This is because the Sx100 has TTL outputs available to connect to the Thermo WDS input BNC cables. Subsequently I learned that the Cameca SXFive is different from the Sx100 is this respect and no longer provides TTL pulse outputs for WDS mapping by 3rd party applications. For some reason I mistakenly associated this change in the SXFive TTL output with the implementation of the Cameca PHA multi channel analyzer (MCA) acquisition hardware, but of course the Sx100 has always had this PHA MCA capability so I don't know where I got that idea...
Subsequently when I learned that JEOL, had also dropped their WDS TTL output capability in their new 8230/8530 instruments I wrongly assumed that they had also implemented this new PHA MCA hardware for PHA scanning acquisition. But that was not the case... so although I've coded this new PHA MCA acquisition capability for both Cameca and JEOL instruments, it only works for JEOL in demo mode!
So what is PHA MCA? Well the final result is no different than with the traditional baseline/window scan method, but it is *much* faster. Right now the SX100 function does not return the integration times so only raw counts are displayed, but we will soon implement a method that can return the elapsed time also, so we can display cps for the PHA scans as we currently do.
So like an EDS spectrum acquisition, one sees the same type of acquisition for the PHA scan, in that it improves the full spectrum precision over time (at least the way Probe Software has implemented the PHA MCA call).
Note that PFE will also perform this PHA MCA acquisition now in demo mode so even JEOL users can see for yourself. What is nice about the demo mode is that it checks the emission line energy and if it is high enough to ionize Ar, it also shows the Ar escape peak correctly in the PHA MCa scan (no not for Xe yet, remember this is just demo mode!). Nice for teaching in off-line demo mode though.
Anyway, JEOL does not currently offer this very nice capability but for SX100/SXFive users, simply update to the latest PFE (10.2.9) and obtain the latest Cameca driver from Probe Software (v. 1.9.0).
Then simply edit the Probewin.ini file for the PHAHardwareType as shown here:
[pha]
PHAHardwareType=1 ; 0 = trad. PHA, 1 = MCA PHA acquisition
Here's what is looks like during acquisition:

Yeah, no visual difference once it's acquired as seen here, though a note is made of the PHA acquisition type:

Note that although the default PHA acquisition mode can be specified in the Probewin.ini file as shown above, one can also change it for the current probe run, by simply checking this box here: