Yes I had never tried MPB, as You probably are aware we have no PFE license. However, I am looking forward and evaluating features at fundamental level which are in PFE, especially that Cameca is not going to advance its software any more further.
If I would be using PFE, where circular interference correction is possible, I would use single background measurements explicitly (and probably MAN too). In that case I would be advocating that "One size fits all".
I'm all for "rules of thumb", but there are always exceptional circumstances and various approaches to them as I hope you will agree. Remember, we don't know, what we don't know.
The approach taken by Jercinovic et al., is to utilize MPB background measurements separately from the on-peak measurements and apply those background measurements to the on-peak measurements using the "Nth Point" methods in Probe for EPMA. This has the effect of vastly improving sample statistics much like the MAN method, without incurring additional sample damage. The paper I cited above doesn't go into this variation of method in super detail but I think it gets a mention.
OK, I hope you have the opportunity to try Probe for EPMA some day. The good news is that it has something for everybody, but a person can certainly utilize only those options and methods that they prefer. It's definitely a "team effort"!
So we get -100 PPM +/- 90 PPM, which is still within 2 sigma of zero, but a bit more negative than I'd like to see. I think Jon Fellowes comments above might also be relevant in this situation.
interesting... was the interference correction measurement acquired also with the same slope?
Yes, I selected all the samples and switched them from 2 point linear to high side only for all V Ka measurements. It's just two clicks of the mouse!

OK, OK, maybe 4 clicks!

Now as I looked into my Wavescans I got an a bit counter-intuitive idea: Would not V Ka measurement in this case be better to do on LPET instead of LLIF? Ti Kb1 will be much broader and it will dominate the interference with V Ka, and Ti Kb'' and Ti Kb5 will be more diffused with less influence. I call it counter-intuitive as that would use worse spectral resolution as an advantage.
Measuring V Ka in SrTiO3 using a PET crystal without an interference correction yields an apparent vanadium concentration of ~1.39 wt% (13,900 PPM) as seen here (TiO2 as the primary standard):
St 851 Set 2 SrTiO3 (strontium titanate), Results in Elemental Weight Percents
ELEM: Ti Ti V V Sr Fe Cr Mn O
TYPE: ANAL ANAL ANAL ANAL SPEC SPEC SPEC SPEC SPEC
BGDS: LIN EXP LIN HIGH
TIME: --- 60.00 --- 60.00 --- --- --- --- ---
BEAM: --- 30.03 --- 30.03 --- --- --- --- ---
ELEM: Ti-D Ti V-D V Sr Fe Cr Mn O SUM
XRAY: (ka) (ka) (ka) (ka) () () () () ()
179 --- 26.153 --- 1.398 47.740 .000 .000 .000 26.150 101.441
180 --- 26.222 --- 1.388 47.740 .000 .000 .000 26.150 101.501
181 --- 26.215 --- 1.372 47.740 .000 .000 .000 26.150 101.477
182 --- 26.144 --- 1.397 47.740 .000 .000 .000 26.150 101.431
183 --- 26.173 --- 1.395 47.740 .000 .000 .000 26.150 101.459
AVER: --- 26.182 --- 1.390 47.740 .000 .000 .000 26.150 101.462
SDEV: --- .035 --- .011 .000 .000 .000 .000 .000 .028
SERR: --- .016 --- .005 .000 .000 .000 .000 .000
%RSD: --- .14 --- .77 .00 .00 .00 .00 .00
PUBL: n.a. 26.100 n.a. n.a. 47.740 n.a. n.a. n.a. 26.150 99.990
%VAR: --- .31 --- --- .00 --- --- --- .00
DIFF: --- .082 --- --- .000 --- --- --- .000
STDS: --- 922 --- 923 --- --- --- --- ---
This was a measurement I did a number of years ago, so when I get a chance I might try it again, though I'd be more interested finding a synthetic high purity ilmenite standard to try these tests on! The same data with the interference correction (and a high off-peak only measurement) produces this result:
St 851 Set 2 SrTiO3 (strontium titanate), Results in Elemental Weight Percents
ELEM: Ti Ti V V Sr Fe Cr Mn O
TYPE: ANAL ANAL ANAL ANAL SPEC SPEC SPEC SPEC SPEC
BGDS: LIN EXP LIN HIGH
TIME: --- 60.00 --- 60.00 --- --- --- --- ---
BEAM: --- 30.03 --- 30.03 --- --- --- --- ---
ELEM: Ti-D Ti V-D V Sr Fe Cr Mn O SUM
XRAY: (ka) (ka) (ka) (ka) () () () () ()
179 --- 26.181 --- -.010 47.740 .000 .000 .000 26.150 100.061
180 --- 26.251 --- -.023 47.740 .000 .000 .000 26.150 100.117
181 --- 26.243 --- -.040 47.740 .000 .000 .000 26.150 100.093
182 --- 26.173 --- -.011 47.740 .000 .000 .000 26.150 100.051
183 --- 26.202 --- -.015 47.740 .000 .000 .000 26.150 100.077
AVER: --- 26.210 --- -.020 47.740 .000 .000 .000 26.150 100.080
SDEV: --- .036 --- .012 .000 .000 .000 .000 .000 .026
SERR: --- .016 --- .005 .000 .000 .000 .000 .000
%RSD: --- .14 --- -61.81 .00 .00 .00 .00 .00
PUBL: n.a. 26.100 n.a. n.a. 47.740 n.a. n.a. n.a. 26.150 99.990
%VAR: --- .42 --- --- .00 --- --- --- .00
DIFF: --- .110 --- --- .000 --- --- --- .000
STDS: --- 922 --- 923 --- --- --- --- ---
We now obtain -200 PPM with a variance of 120 PPM, so again within 2 standard deviations of zero! Again a little more negative than I'd like to see... Frankly I think Jon Fellowes is correct about the bonding environment and that we should seek a high purity synthetic FeTiO3 material for a more accurate test of correcting the Ti Kb interference on V ka in ilmenites (though TiO2 should work absolutely fine as an interference standard for correction of Ti Kb on V Ka in rutiles). I did find this article on synthesizing FeTiO3 from 1978:
https://www.sciencedirect.com/science/article/pii/0022024878903123It's sort of funny to find this paper, because back in the 80s (starting out as a fledgling microanalyst), I had written Takei about his synthesis of fayalite (Fe2SiO4) and he very nicely sent me a little crystal which I examined. The only problem was that it had tiny little blebs of Fe metal. But then (I can't remember the circumstances exactly) Lynn Boatner at Oak Ridge sent me a synthetic fayalite which I have in several of my standards mounts at Oregon and I think I even have a tiny scrap left in a bottle in the standard materials collection. The material from Oak Ridge is perfect, and I wrote him a few years ago asking if he still had any additional material to spare or at least the recipe for this synthesis, but he never responded.
Finding significant (commercially produced) quantities of Fe2SiO4 and FeTiO3 would be fantastic. I think Will Nachlas, Aurelein Moy and John Fournelle are working on obtaining these high purity end member materials for use as microanalytical standards.
It sure would be nice.