It is a long story why I do this. The short version is, I analyze Al in a sample, but sample and standards are coated with Al. And the thickness of the Al coating on standards and unknown is all different.
Philipp and I discussed his situation and we now understand each other better.
Basically he has samples (standards and unknowns) with different thicknesses of Al coating. Unfortunately he needs to analyze for Al in his unknown samples! Ugh!
![Angry >:(](https://probesoftware.com/smf/Smileys/default/angry.gif)
Probe for EPMA handles this situation of different coating thicknesses for standards and unknowns just fine (it even handles the situation of different standards with different coating materials and thicknesses), but because he is trying to
measure the aluminum concentration in a sample which is coated with aluminum, he also wanted to find a way to "subtract" some of the Al Ka signal (that corresponds to the "excess" Al Ka signal from the Al coating, before the quantification of the Al in the actual sample!
So he thought that he might export the k-ratios from PFE and then edit the CalcZAF input file for this purpose. The problem he ran into is that unlike Probe for EPMA, CalcZAF isn't quite as smart as PFE is about handling standards with different coatings...
The problem that Gian ran into a few months ago was that CalcZAF was re-setting the coating parameters on the standards (back to 20 nm of carbon) when loading in a new unknown sample from the CalcZAF input file. This was fixed, but because each sample in a CalcZAF input file *could* have different standards, I re-load the standard parameters for the new sample which then resets the standard coating parameters back to the last coating specified for the previous standard. This worked fine for Gian because his standards were coated differently than his unknowns, but all the standards were coated with the same thickness.
Not so for Philipp! His standards (for some reason) are coated with different thicknesses of Al. But then I realized that Philiip can simply do his aluminum intensity "tweaking" in Probe for EPMA by utilizing the "specified APF" parameter in the PFE Elements/Cations dialog to account for the "extra" Al signal from his coatings. This way he can utilize the extra power of PFE where one can specify each unknown and/or standard with different coating parameters and all is good.
Whew!
He will still need to be careful about fluorescence of the Al coating from his standard and unknown materials, but at least it is possible in PFE to specify different coating materials and thicknesses for each standard or unknown.
john