If you want to verify the presence of boron using LDE2, then you need to increase the dwell time, perhaps to 1 s, and bump the beam current way up, maybe to 200 nA or so -- tourmaline can handle it, though I'd defocus the beam to 10 microns.
Brian
Thanks for the advice Brian. I'll try it.
I was brought an unknown mineral and asked to scan for Boron. So I first tried to see the B peak on a tourmaline (I assume the user who gave it to me correctly identified it optically; I don't have an EDS
).
I'm trying to wrap my head around what's possible with LDE2. At 10nA there appears to be peaks at about 190-ish and 195 in various blanks, and the tourmaline (if those are in fact peaks and not noise). At 30nA some of the "peaks" begin to merge. I'll see what 200nA kicks out.
Have you tried to quantify B in tourmaline with LDE2?
Deon.
Hi Deon,
I've never analyzed for boron in tourmaline because generally it's reasonable to assume wt% B
2O
3 that gives 3 B
3+ per 29 anhydrous oxygens. Like John pointed out, the absorption correction when using BN as a boron standard will be outrageously large and will likely lead to inaccurate wt% B
2O
3 (not to mention problems with variable peak position, variable peak shape, and poor counting statistics). Many tourmalines fall at least
roughly within the dravite-schorl solid solution. For the dravite end-member (NaMg
3Al
6(BO
3)
3Si
6O
18(OH)
4), wt% B
2O
3 is 10.89, and wt% H
2O is 3.76. For end-member schorl (NaFe
3Al
6(BO
3)
3Si
6O
18(OH)
4), wt% B
2O
3 is 9.91, and wt% H
2O is 3.42.
One approach is simply to analyze for Si, Al, Ti, Cr, Fe, Mn, Mg, Ca, Na, K, F, and maybe Cl (not likely to be present in measurable quantity) and adjust assumed wt% B
2O
3 and wt% H
2O (taking halogens into account if present) and see if you can produce a reasonable-looking tourmaline formula unit. Keep in mind, though, that Al
3+ is somewhat variable, as it can substitute for Si
4+ in a coupled substitution that also replaces a divalent cation with
VIAl
3+. Significant Fe
2O
3 may be present as well. Also, Li
+ can substitute on the octahedral sites, and this can complicate interpretation of the analysis -- it also requires adding octahedral Al
3+ to maintain charge balance such that elbaite (Na(
VIAl,Li)
3Al
6(BO
3)
3Si
6O
18(OH)
4) may contain in excess of ~43 wt% Al
2O
3.
Maybe you already know about this, but a good on-line resource is the
Handbook of Mineralogy, as it gives representative analyses of most of the listed minerals.
Brian