Good afternoon, everyone:
Over the last short while, I have ventured into the realm of monazite chemical-age dating using PfE software, and have done a lot (not exhaustive, however) of background reading of topics in the forum, which have been very helpful and informative. I am not an expert with this method at all, however.
A student using the PfE software noted that he got age results that are about 15% lower than expected for an age-standard we have in the lab (a GSC standard). In going through his file, there are a few samples where he didn't specify concentrations for the 'major' elements, so I would expect age results from these samples will be off significantly, but I am still searching for answers for his other unknowns, where he assigned the correct values for majors using the "Specified Concentrations" option. (Are some default values used for matrix-corrections where concentrations of majors are not specified?)
First of all, unfortunately, we don't have a well-characterized YPO4 (xenotime) standard in the lab, so we had to resort to a YAG standard for correcting the Pb_Y peak overlap, which is not ideal owing to the difference in chemical states of Y between YPO4 and YAG and associated peak-shifts. If anything, shouldn't one expect apparent calculated ages to be higher than accepted for an age-standard (and for unknowns) using YAG for the Pb_Y versus those obtained using a YPO4 standard?
Secondly, we don't have an anhydrous Pb-phosphate standard for Pb, we have a PbS (galena) and a Pb-chromate (crocoite) standards. My predecessor had used crocoite in the past, in the inherited methodology. Is there an inherent shift of the PbMa peak similar to that of the interfering Y-peak that I have missed that would account for a 15% difference in observed PbMa intensity that would carry through into the age calculation?
Thirdly, owing primarily to tight time constraints in trying to adapt someone else's protocol from an old version of JEOL software (i.e., UNIX-based, circa 2004) (and where I suspect the main fault is going to be found in the 15% error in age determination, mentioned above), only 1 upper and 1 lower off-peak intensities were measured, instead of the recommended, more rigorous backgound-modelling methods widely discussed in this forum (and in the scientific literature). I realize that rigorous background determination for unknowns is more critical in obtaining good analyses for this technique than is standardization/calibration.
My recommendation to the student was to re-acquire minor-element data using a multi-point background model, while I track down a good piece of YPO4 for the Pb_Y overlap correction. The student then asked if it was possible to use already acquired wavescans and associated DAT files to extract background values and somehow plug those values into the age calculation. I pointed out that the analytical conditions used during the wavescan acquisition are not the same as those used for the minor element determination, and that, even though the intensities are quoted as being normalized to cps/20 nA for both wavescan and quantitative data, the reliability of the counting statistics for the wavescans is relatively low, owing to the huge difference in dwell times between the two types of measurements. That being said, I wouldn't know where in the software one would put in 'normalized (to 20 nA)' background values from multiple points taken from wavescans, nor even if it is possible (nevermind advisable!) to do this manually.
Any helpful hints (and strong boots to the rear!) in the right direction are most appreciated!
Best regards
Dan MacDonald
Probe servant
Dalhousie University