Either way, opting for all P10 detectors to minimize running costs would result in an EPMA very severely limited in its capabilities.
*FYI, Cameca used high pressure P10 detectors instead of sealed Xenon detectors.
As Cameca was mentioned here, as user of Cameca SX100 and SXFiveFE with only P10 detectors I share here my toughs. TLDR; The advantage of high pressure P10 detector is that intensities are long-term and short-term stable, where Xe detector is only short-term stable, and low-P P10 is only long-term stable:
detector gas | [Low P P10] | [HighP P10] | [Xe] |
long-term stable | yes | yes | no |
short-term stable | no | yes | yes |
Low Pressure P10 requires re-calibration after seasonal changes, also they are clearly very sensitive to atmospheric front passing (drop or raise in pressure). If compared with low pressure P10 det., then, yes, sealed Xe detector looks immune to atmospheric changes, and that could be kind of advantage. However, Xe detector and P10 low pressure are aimed to different energies, and Xe is not used for low energy X-rays.Furthermore, it looks less advantageous when compared with high pressure P10 detector, which intensities are both - long and short time stable. The intensities on high pressure P10 detectors keep stable for few/many years - which makes the instrument more efficient as it needs less frequent (re-)calibrations - there is more time available for analysis. In contrast, with constant Xe escaping from Xe detector - the calibrations probably needs to be (re-)done quite often for all higher energy X-ray lines. Thus I would chose high pressure P10 detector any moment instead of Xe detector for LiF and PET, and would consider a probe without such an option to be defective-by-design. Such defect could only be compensated with
lifetime free-of-charge unlimited Xe detector replacement contract or/and a significant discount (should be same price as P10 option) to compensate the time required for repeated re-calibrations. The only advantage compared with low pressure P10 is that recalibrations of Xe detector can be planned periodically in advance, there with low pressure P10 detectors the (re-)calibrations will follow local weather events.
BTW, if outlet pressure of P10 gas from detectors could be regulated to absolute value instead relative to the atmosphere (or atmosphere pressure of lab is stabilized), low pressure P10 could have also improved short-term stability. Maybe probe developers/builders could offer rather some ready-built-in solution for that... that would be some value-added worth to pay more.
P.S. I scratched my eyes twice, as p/b given by that Virtual WDS looks at disadvantage for Xenon. That is indeed correct, as too much signal is not good too, as it will over-saturate counting system. Higher intensity comes with downsides as it gets easy to go over 200 kcps for major elements. In conjunction with large XTALs it is easy to get intensities which hitherto sold counting electronics were not designed to cope with. Xe detector makes sense for trace analysis, but for major elements that could be kind of hindrance.