I was going to respond with a suggestion to put your entire instrument on a 240v 12 KW UPS because that's what we have on our SX100 (it only runs for 15 minutes, but within a few minutes of a power outage the diesel generators will kick on and they will run for days):
https://probesoftware.com/smf/index.php?topic=1288.msg10726#msg10726
but then I saw your post here:
https://probesoftware.com/smf/index.php?topic=1288.msg10727#msg10727
So, you are doing the same thing apparently? Why then do you need a separate power backup for your diffusion pump?
Without diesel generator the UPS for SX100 is useless (in diff pump prevention). We have beefy UPS supplying only to our SXFiveFE where it makes sense (and that I mentioned in those other topics). However our SX100 is not under UPS protection, and neither it would make a difference.
There are few crucial differences between SXFive(FE) and SX100 which makes UPS relevant or irrelevant.
SXFiveFE:
* has Schottky tip - can be destroyed with abrupt cutting of supply
* has column-gun motorized separation valve
* has turbomolecular pump
* uses valve(s) to separate the vacuum between chamber and spectrometers which requires pressure and energy to stay closed (loosing power will automatically open valves unify spectrometer and chamber vacuum).
SX100:
* W/LaB6
* manual column-gun valve
* diffusion pump
* motorized valve to separate or unify chamber and spectrometer vacuum directly mechanically connected with diffusion pump inlet valve - power is needed to switch it to opened or closed states.
Case with power-failure with no UPS for SXFiveFE:
0. Power down.
1. Tip gets off abruptly (for fractions of second being hot without any field containment), ion pumps releases Ar and some other lousily buried stuff - high risk of tip contamination.
2. motorized column-gun separation valve can't be closed as there is no power thus gun and chamber spaces keeps being connected.
3. turbomolecular pump starts naturally to slow down - the chamber vacuum gradually approaches primary vacuum during that... but because of point 2 UHV space of gun is getting contaminated by primary vacuum eventually.4. valves open between spectrometer and chamber unifying the pressureIn case of P10 gas leak to spectrometer - P10 contaminates Gun chamber at that point.Thus UPS (and corresponding software solution) for SXFiveFE is not just optional - it is a must.
See the difference with presented case with power-failure (lets say for 4 hours) and UPS present and equipped with communication and preventive measures (it is not enough to have an UPS. The UPS communication with PC and the state of UPS needs to be accessible to custom made daemon which will correctly prevent damages in correct steps or pull out from prevention sequence):
0. Power down, UPS keeps the system fully operational for 2 hours, (but Air conditioning is not under UPS, thus it is not wise to keep tip fully operational for 2 hours).
1. 30 min after power down, decrease tip heat by 0.2 A (custom script)
2. 1h - initiate full safe tip shutdown (takes about 10 min), (at older versions close the column-gun valve with corresponding script, at newest firmware it does automatically after gun is shutdown). (there is exe for gun shutdown in newest peaksight, thus can be executed with independent UPS software)
3. if it is possible trip the chiller - that will trigger preventive shutting of electronics - two biggest power consumers out - UPS got additional few hours which it can keep whole vacuum system running. As there is loss of communication with SX at that moment PC should be also triggered to shut down which will save even more battery.
After 4 hours power is back: the vacuum was kept fully functioning. After turning FEG back on in one day system is ready back to be used.
In case power would not come back - eventually all pumps would go off, but as all valves which needs to be shut and all valves which needs to be open are correct - It would take just additional few hours to return back to functional state, and again in one day system would be ready back.
In case of no UPS - it would require baking of UHV part, and would take
at least 2 days to get back to functional state. Also it has some additional risks when engaging baking procedure.
Now lets look to sequence of events during power failure at SX100 (without UPS):
0. Power down
1. PC goes instantly off.
1. in about 0.5 second significant voltage drop triggers electronics to shutdown and cut off supply to other systems (SX100 is capable to go through very short sub second outages without a hitch due to most of integrated PSU's containing very huge capacitors).
2. Gun is forcibly off (still steaming hot, if it is LaB6 - that can be kind of catastrophic), diffusion pump heater is off (but oil does not cool instantly, it is still actively pumping for few seconds after), roughing pumps are off, ion pumps are off.
3. Pressure starts to raise in the spectrometers (as those have biggest leaks), however bottom of diffusion pump is still hot enough to boil the oil (and there is no cooling of diffusion pump as water chiller is off too).
4. The backing buffer is separated from roughing pump with electromagnetic valve which gets closed with power down. Thus diffusion pump has good vacuum reserve from backing side for cooling down. However inlet is in danger as pressure raises there.
5. And there comes this mentioned 100mA 12V battery, which powers single event of diff pump inlet valve closing, which prevents hot diffusion pump from interaction with gases from deteriorating vacuum in the chamber. In case of no leaks between spectrometer and chamber there is not such a big deal, but otherwise there is risk of deterioration of diffusion oil and its partial breakdown in reaction with contaminants from spectrometer (i.e. P10 gas).
To precise up what kind of battery I am blabbering here about, look to the pictures below.
The placement of 12V battery in SX100, here it is already taken out:

and this is how battery looks from closer:

marking on two lines are as follows:
10 VB 10
PC9602 REF 125306
vendor marking in background: SAFT
Now lets look how UPS with 15 min capability would help here (without diesel generator):
0. Power goes Down, UPS takes over.
1. PC can switch the Gun off. And that's all what can be achieved! (PC can also shutdown itself to prevent data-loss, albeit 15 min of LaB6 being off could make a difference for it to survive.)
2. Can't Halt vacuum (so that diff pump could cool down safely), can't do anything as all other vacuum states requires manual switching of gun-chamber valve to position 1.
3. after 15min the sequence of event is the same as in the list of events above for case without UPS... and finally it comes again to the moment where 100mA 12V battery is needed to prevent the diff pump deterioration by closing the diff pump inlet valve.
Of course if after all this, soon (~5-10min) the power will come back the valve will close automatically due to hard restart of vacuum system, thus there is a bit less of danger. The worst case (highest risks) without functional 12V battery is with long outage. So Probeman, lucky You, that Your lab/institution has well thought power backup system. At least someone at your institution thinks in advance.
Anyway we had for many years the UPS for SX100 as it helps in situations when power failure happens during work hours, but we had 3 consecutive break downs of UPS within single year, and just gave up after with UPS for SX100 (especially that it does not brings any crucial prevention).
And we are heading off-topic. 12V battery is important to prevent the diff pump oil from weak vacuum or contamination from spectrometer. The valve is complicated it is kind of double - it closes inlet of diffpump and at same time opens the inlet to spectrometers (equalizing vac between spectrometer and chamber). It is important to not deteriorate diffusion oil and probably during many last years (probe was assembled in factory in 1998) it could during some power outages.
However my main problem is Old diffusion pump itself. And as I got only single (and a bit off-topic) answer, and looking to the Cameca design I fear it is mounted for a life-time with no oil control in mind. I.e. to check the oil level the pump needs to be vented to air pressure first, and in the way how it is mounted and connected in SX100 - it looks that venting of it was never seen as any of options. Also lack of feedback means that no one probably had seen any maintenance work being done on those pumps during service visits.