Good afternoon, John:
Thank you very much for all your helpful advice, I am most grateful to have it.
I have a question arising from some trials with the method of having the spectrometer peak before every analysis and 'letting it rip'. When I did this for a small batch of points, I manually updated the on-peak value (MgKa in CH3TAPH) by doing a manual peaking. When I ran the points with the automatic peaking, the first point had an L-value I had put in, i.e., 107.314 from a manual peaking, and when it did the automated peaking, it displayed the centroid value at 107.3119 but used the L-value 108.1088 which is way over on the flank of the peak. That first analysis gave a 'correct' net intensity vale of about 46,994 counts (i.e., extremely close to what I had thought it would be based on the manual peaking). The second and third points of the run gave similar peaking results and in each case but the centroid value was ignored and the program used the 108.109 L-value and didn't update the on-peak value to that of the centroid. Also, the second and third points yielded net intensity values of 1720 counts and 1689 counts, respectively, compared with the first point at 46994 counts. These 2 lower net intensities make perfect sense when looking at the peaking graph, as they are the intensities taken at the L-value of108.109. Also, when I did a re-peaking of MgKa after the run had completed, the following error message showed up:
Starting spectrometer peaking procedure for Mg ka on spectro 3...
Mg ka Spectro 3, StartPk: 108.109, StartI: 1772.5, HiPeak: 63.5, LoPeak: 21.5, StartP/B: 41.71
ROM Scan completed on spectro 3 with 50 total points
Spectro 3 Fine Scan FAILED- Calculated peak centroid at 107.3095, is too far offset from original peak position of 108.1088
ROM Scan completed on spectro 3 with 150 total points
Spectro 3 Coarse Scan FAILED- Calculated peak centroid at 107.314, is too far offset from original peak position of 108.1088
Spectro 3 FAILED- Coarse peaking did not succeed, will use initial position of 108.1088
Mg ka Spectro 3, StopPk: 108.109, StopI: 1675.0, HiPeak: 63.5, LoPeak: 21.5, StopP/B: 39.41
ROM Gaussian Peak Center Results (cps):
Element Spectr Peaked StartPk StopPk Offset StartI StopI StartPB StopPB
Mg ka 3 TAPH No 108.109 108.109 -.31 1772.5 1675.0 41.71 39.41
My questions are as follows: 1. What is this L-value 108.109 supposed to be, i.e., where did the program get this number? It seems to be close to the theoretical value of 107.7985 in your periodic table (107.513 in JEOL's L-value table). 2. Is the automatically determined L-value supposed to update to that of the calculated centroid, and if not, why not, as it yields the 'correct' net intensities. 3. Regarding the error message that showed up, I see that the peaking procedure started at the incorrect 108.109 L-value (as expected), found the correct centroid and rejected it because it was too far off from the initial (but wrong) peak position. The centroid is the correct position, however. So I am not sure how to get around this issue, because my whole aim is to see how the peak position and associated intensities drift over time, but that will be (almost?) impossible if the on-peak position isn't correct, as the wrong intensities will be collected at that default L-value of 108.109 every single time, because the software will reject the centroid value which is correct, even if it is offset by what is an unacceptable amount to the peaking software. How can I get around this and what am I doing wrong?
Thank you very much for all your kind help and patience.
Best regards,
Dan