Correct. I have not moved any slits etc. all counts for 60 sec on peak 10 sec off. Pha parameters set so that i could get good results at high and low counts. No obvious charging that I could detect.
All done according to the pdf I was surprised at the count rates also, si metal counts were ~2500 cps/nA on tap sp1, orthoclase was ~600 cps/nA. All measurements at 15kV.
Hi Dan,
OK, thanks.
Remember, you only need to check the PHA settings at the *highest* (expected) count rate in INTEGRAL mode. Also when you calculate your "predicted" count rates, you should use the count rates observed at the lowest beam currents (in cps/nA) for the primary standard and just multiple by the beam currents. After all, it's the high count rates on the primary standard (Si metal in this case) that is stressing the dead time correction model in the k-ratios.
I know this goes against everything we were taught back in the day, but if the PHA peak is fully above the baseline level at the highest (expected) count rate, then at lower count rates the PHA peak will merely shift to the right and
still be counted in INTEGRAL mode as described here:
https://probesoftware.com/smf/index.php?topic=1466.msg11450#msg11450Also be sure your beam is defocused to 5 or 10 um or so. I assume that none your your crystals are TAPL crystals? That would explain the lower count rates. But really it's not the count rates that bother me, and your k-ratios look fairly constant, though I don't understand why they are going non-linear at these relatively low count rates.
But what really bothers me is that your spectrometers are giving such different k-ratios. That is, we don't really care what the absolute values of the k-ratios are, after all we are measuring Si Ka in Orthoclase relative to Si metal, and of course there is a significant emission (energy) peak shift between the two materials, especially for Si Ka.
But regardless of what that k-ratio value is observed to be (at the lowest count rates where the dead time correction is insignificant), we still should be seeing that *same* k-ratio on all our spectrometers. Your plot shows k-ratios from ~0.17 to ~ 0.25 for Si ka in these two materials and that is a significant difference as you have noted on your plot. Though if you utilized a mix of TAP and PET crystals that could explain some of the differences. Can you plot the PET and TAP constant k-ratios in separate plots (maybe that is what you did to begin with)?
![Embarrassed :-[](https://probesoftware.com/smf/Smileys/default/embarrassed.gif)
Yes, it could also be sample (or stage) tilt. Scott Boroughs raised that question to me here:
https://probesoftware.com/smf/index.php?topic=1466.msg11329#msg11329as he saw similar (though smaller) variations in his simultaneous constant k-ratios, which he felt might also be due to sample (or stage) tilt in that it appeared systematic with respect the spectrometer orientations around the instrument. And I responded in the next post that this is something, which if appropriately characterized, we could compensate for in the absorption correction by changing the "effective" takeoff angle for each spectrometer.
As you can see in my response, Probe Software had modified the underlying physics code in CalcZAF to handle different spectrometer take off angles for each spectrometer and utilizing that in the absorption correction. I haven't heard back from Scott if a change in the effective takeoff angle due to sample (or stage) tilt would compensate for this.
More disturbing is the possibility that the variation in these simultaneous constant k-ratios could be the result of differences due to asymmetrical Bragg diffraction. In this case, each crystal could demonstrate a different effective takeoff angle. The problem is knowing which spectrometer has the correct effective takeoff angle! And for that determination we would need to worry about differences in the emission peak position, differences in the carbon coating, native oxide thickness, etc, etc. All the usual suspects we deal with in any quantitative analysis...
PS you might want to try Ti Ka on TiO2 and Ti metal as the emission peak shifts will be significantly less for these materials, though you'll still have the native oxide layer issue with Ti metal. But again, the absolute value of the k-ratios do not matter, only that they remain constant as a function of count rates!