I have recently been contacted by a post-doc asking about accuracy issues in EPMA and how they might relate to problems they are seeing in the P/T results in pyroxene systems. These are excellent questions! The following is part of my response to these questions.
I didn't realize that the Na concentrations for these OPX/CPX thermo/barometers were so low. Knowing this now, here's the deal with EPMA accuracy (assuming no instrumental problems!):
1. For major concentrations what matters are the standard accuracy and the matrix correction accuracy. I don't think the matrix correction accuracy is a problem with modern software (see below). But the heterogeneity in the Smithsonian standards is a well known problem. In fact I think Ed Vicenzi has characterized some of these issues in the past.
2. For trace concentrations what matters are the background corrections (and interference corrections if necessary). It is interesting to note that labs using JEOL EPMA instruments (which has roughly an 80% market share) have no way to correct for spectral interferences. That is unless they utilize software from Probe Software.
Bottom line: the above two points make sense if you consider that the matrix corrections are multiplicative and scales (as does standard accuracy) with concentration (so the smaller the concentration the smaller the absolute accuracy error). That is, a 1% relative error in accuracy is 1 wt% at 100 wt%, but only .1 wt% at a 10 wt% concentration, and only 1 PPM at 100 PPM.
While the background correction is subtractive and is a constant source of error (so the smaller the concentration the larger the accuracy error). That is, a 100 PPM error in the background estimation is a 10% error at a 1000 PPM concentration, but it's a 100% error at 100 PPM and a 1000% error at 10 PPM. This is worth thinking about a bit- it's not obvious. If it was obvious I wouldn't be having to rant about this for so many years! 😁
For minor elements it's a bit of both considerations (standards/matrix and also backgrounds/interferences), but mostly background correction, with some influence from the matrix correction. And also spectral interferences depending on the elements in question. For this system I don't think spectral interferences are an issue, as the only other elements that would interference seriously with Na are Zn and a little from P (but these should elements be present in very low concentrations in most augite compositions, I think?).
Some might suggest a low Na pyroxene standard might fix this accuracy problem and that is the traditional, but wrong approach in my opinion. For highest minor/trace element accuracy in pyroxenes what we should want, is an approximately matrix matched standard with a *zero* Na concentration. A non-zero minor/trace (and homogeneous) element standard is very difficult to obtain and to characterize.
What is much easier (and better) is a "blank" standard roughly similar to ones unknown (or even a glass with similar elements) that has *zero* (say, less than 1 PPM) Na present. This we can determine easily (e.g., ICP-MS, SIMS)
and then our minor/trace accuracy is as good as our measurement precision (think about this, as it's a gift from the science gods). Basically we are simply testing our background correction accuracy by using a suitable blank standard run as an unknown. This is all spelled out in the Donovan et al., 2011 Amer. Min. paper.
In the Probe for EPMA software, this "blank" correction is applied iteratively during the matrix correction for highest accuracy. It can also be subtracted out after the fact in an Excel spreadsheet manually, as long as the composition doesn't change too much from the blank correction. This is usually the case.
I know this is not what we were taught in grad school, but it's the way forward in this current mess of heterogeneous natural standards. As an example, I couldn't find a pyroxene analysis, but here's an orthoclase analysis with around .8 wt% Na showing all the various matrix corrections in the CalcZAF software:
Elemental Weight Percents:
ELEM: Na Si K Al Mg Ca Ti Mn Fe P O H Ba TOTAL
1 .768 30.520 12.862 8.625 .000 .014 .001 -.005 1.355 .000 45.798 .000 .054 99.991 Armstrong/Love Scott (default)
2 .764 30.621 12.864 8.571 .000 .014 .001 -.006 1.383 .000 45.798 .000 .054 100.064 Conventional Philibert/Duncumb-Reed
3 .768 30.419 12.861 8.623 .000 .014 .001 -.005 1.335 .000 45.798 .000 .054 99.869 Heinrich/Duncumb-Reed
4 .767 30.577 12.863 8.594 .000 .014 .001 -.006 1.356 .000 45.798 .000 .054 100.017 Love-Scott I
5 .768 30.547 12.863 8.621 .000 .014 .001 -.006 1.355 .000 45.798 .000 .054 100.015 Love-Scott II
6 .756 30.879 12.866 8.561 .000 .014 .001 -.006 1.375 .000 45.798 .000 .054 100.298 Packwood Phi(pz) (EPQ-91)
7 .763 30.668 12.862 8.497 .000 .014 .001 -.005 1.343 .000 45.798 .000 .054 99.995 Bastin (original) Phi(pz)
8 .768 30.561 12.862 8.547 .000 .014 .001 -.006 1.378 .000 45.798 .000 .054 99.978 Bastin PROZA Phi(pz) (EPQ-91)
9 .767 30.616 12.863 8.554 .000 .014 .001 -.006 1.377 .000 45.798 .000 .054 100.037 Pouchou and Pichoir-Full (PAP)
10 .764 30.630 12.863 8.570 .000 .014 .001 -.006 1.378 .000 45.798 .000 .054 100.067 Pouchou and Pichoir-Simplified (XPP)
AVER: .765 30.604 12.863 8.576 .000 .014 .001 -.006 1.364 .000 45.798 .000 .054 100.033
SDEV: .004 .119 .001 .041 .000 .000 .000 .000 .017 .000 .000 .000 .000 .109
SERR: .001 .038 .000 .013 .000 .000 .000 .000 .005 .000 .000 .000 .000
MIN: .756 30.419 12.861 8.497 .000 .014 .001 -.006 1.335 .000 45.798 .000 .054 99.869
MAX: .768 30.879 12.866 8.625 .000 .014 .001 -.005 1.383 .000 45.798 .000 .054 100.298
As we can see, the variance in the Na matrix corrections are very small (40 PPM), so the main problem with Na concentrations in the cpx/opx barometer is the heterogeneity of the Smithsonian "standard", and the background corrections, where I'm guessing many of us haven't tested our ability to measure zero Na in a roughly similar pyroxene matrix.
I did a quick search in my standard database to the Kakanui augite compositions and here is what I came up with:
![](https://probesoftware.com/smf/gallery/395_18_10_21_1_10_32.png)
So this Dahlheim glass is a close match, but this was a glass produced by Dan Weill at Oregon, so yes, I have a fair amount of it (which I shared with Paul Carpenter at Washington University), but is the Na content truly zero? I don't know, but it could be characterized using ICP-MS/SIMS... anyone interested in doing this for us?
The other close match to an augite composition is the NIST K-412 glass which nominally contains no Na, but I seem to have reported some 500-600 PPM of Na in this material some decades ago. Is that accurate? I have no idea. NIST does not report it in their certificate. Has anyone out there measured trace elements in K-412 (or K-411) using ICP-MS or SIMS?