I’ve attached a couple of plots that show graphically how useful the blank correction can be. I have a rutile standard (NMNH 120812) that I know contains vanadium, but I’ve had a great deal of difficulty verifying the amount precisely due to interference from Ti Kβ1,3, Ti Kβ5, associated satellites, and also the Ti K absorption edge. Of course, this is a worst-case scenario in which the Ti concentration is very high and the V concentration is very low. Recently, I acquired a high-purity synthetic TiO2 standard, and I’ve overlaid its spectrum (using LiFL) on that of the natural rutile in the first plot below. In the second plot, I’ve essentially done a manual blank correction by subtracting the high purity TiO2 spectrum from that of the natural rutile (difference illustrated in green). There is absolutely no way to perform an overlap correction with confidence in this case, and the blank correction provides the only useful result (variable between ~0.2 and ~0.3 wt% V2O3 in this case), especially since natural rutile is generally close to end-member composition. In this particular case, the rutile contains ~98.9 wt% TiO2. I hope that someone will find this useful for both the general case and for this specific case. I often find that people minimize the importance of interference from Ti at the V Kα peak position when using LiF.
Hi Brian,
Looking through my old data I realized I've never really looked closely at this specific interference. So yesterday I acquired a test run analyzing a bunch of Ti standards for vanadium and today I took a look at the data. It's more interesting than I thought it might be... there is something going on that I don't understand. Interestingly I get very similar results using both off-peak and MAN background corrections, so I don't think the issue is extrapolation across an absorption edge.
Conditions were 15 keV, 30 nA and 60 seconds on-peak and 60 seconds off-peak. The background positions for V Ka on LiF are here:
And for V ka on PET here:
Where the red line is the on-peak position for V Ka, the purple lines are the old off-peak positions and the green lines are the off-peak position utilized for the acquisitions.
I acquired both Ti Ka and V Ka on both LiF and PET crystals just for fun and acquired 5 points on each standard. That way we can look at LiF data only, PET data only, or aggregate the intensities from both the LiF and PET crystals. For example, looking at the off-peak data on LIF crystals first I get this for TiO2 when no interference correction is applied:
ELEM: Ti Ti-D V V-D Sr Fe Cr Mn O SUM
XRAY: (ka) (ka) (ka) (ka) () () () () ()
AVER: 59.956 --- .551 --- .000 .000 .000 .000 40.000 100.506
SDEV: .073 --- .010 --- .000 .000 .000 .000 .000 .083
SERR: .033 --- .005 --- .000 .000 .000 .000 .000
%RSD: .12 --- 1.84 --- .00 .00 .00 .00 .00
As you can see I get about 0.5% wt% apparent V in TiO2. Turning on the interference correction and using TiO2 as the interference standard, we of course get a perfect correction because it's analyzing itself:
ELEM: Ti Ti-D V V-D Sr Fe Cr Mn O SUM
XRAY: (ka) (ka) (ka) (ka) () () () () ()
AVER: 59.988 --- .000 --- .000 .000 .000 .000 40.000 99.988
SDEV: .073 --- .009 --- .000 .000 .000 .000 .000 .082
SERR: .033 --- .004 --- .000 .000 .000 .000 .000
%RSD: .12 --- 9536.47 --- .00 .00 .00 .00 .00
Now if we look at another Ti standard, say SrTiO3 we obtain this with the interference correction on:
ELEM: Ti Ti-D V V-D Sr Fe Cr Mn O SUM
XRAY: (ka) (ka) (ka) (ka) () () () () ()
AVER: 26.329 --- .010 --- 47.740 .000 .000 .000 26.150 100.228
SDEV: .223 --- .005 --- .000 .000 .000 .000 .000 .222
SERR: .100 --- .002 --- .000 .000 .000 .000 .000
%RSD: .85 --- 50.80 --- .00 .00 .00 .00 .00
Not terrible, but still 2 standard deviations from zero. And next we can do an extrapolation to a material containing a larger concentration of Ti such as TiC as seen here:
ELEM: Ti Ti V V Sr Fe Cr Mn C
TYPE: ANAL ANAL ANAL ANAL SPEC SPEC SPEC SPEC SPEC
BGDS: LIN EXP LIN EXP
TIME: 60.00 --- 60.00 --- --- --- --- --- ---
BEAM: 30.03 --- 30.03 --- --- --- --- --- ---
ELEM: Ti Ti-D V V-D Sr Fe Cr Mn C SUM
XRAY: (ka) (ka) (ka) (ka) () () () () ()
AVER: 78.622 --- .012 --- .000 .000 .000 .000 20.000 98.634
SDEV: .259 --- .011 --- .000 .000 .000 .000 .000 .250
SERR: .116 --- .005 --- .000 .000 .000 .000 .000
%RSD: .33 --- 92.30 --- .00 .00 .00 .00 .00
Actually not too bad, just over a standard deviation from zero. Next we'll look at the results from the PET crystals and that's where the wheels really come off...