Here's a possibly interesting method for determining sulfur concentrations *and* the approximate oxidation states *at the same time* on a microprobe.
As many of you know, depending on the oxidation state of sulfur, the emission peak on a WDS spectrometer varies enough to make it necessary to accurately position the spectrometer on the actual emission peak position. Basically there is roughly a 10% change in intensity between the pyrite and anhydrite sulfur peak positions.
Determining the degree of peak shift in a basaltic glass when the sulfur concentrations are around 1000 PPM is difficult, partly because a high enough precision wavescan requires considerable time, causing sample damage, and also possible oxidation of the sulfur in the glass, which causes further peak shifting. Specifically we've found it necessary to acquire sulfur wavescans on our basaltic glasses for around an hour each, and even then we've found it necessary to have PFE increment the stage position a few microns every minute or so to avoid further oxidizing the sulfur.
The basic idea being that once we determine the actual emission peak position for a given glass sample, we can set the spectrometer to that position and collect our sulfur intensities at the peak position for accurate quantitative analysis, regardless of the sulfur primary standard.
But earlier this week two graduate students, Dan Rasmussen and Michele Muth, came up with an idea for determining the sulfur peak position by essentially creating a "multi-collector" microprobe for characterizing reduced to oxidized sulfur "species", by tuning each of our 5 PET spectrometers to cover the range of sulfur oxidation peak positions.
So we first tuned all the spectrometers to our pyrite primary standard (-1 valance), spectrometer offset equals 0, and knowing that the pyrrhotite sulfur peak (-2 valence) is shifted to the right +4 units (in Cameca units), and anhydrite is shifted to the left by 30 units, we adjusted our 5 WDS spectrometers as seen here:
+4 0 -10 -20 -30
Spectrometer 1 being the spectrometer tuned to the pyrrhotite sulfur peak, spectrometer 2 tuned to the pyrite peak and spectrometer 5 tuned to the anhydrite sulfur peak position, with spectrometers 3 and 4 tuned to intermediate peaks positions between pyrite and anhydrite.
The idea being that the spectrometer that produces the highest sulfur concentration will therefore probably be the correct concentration, and also give us some information on what the oxidation state of the sulfur is, depending on which spectrometer it is. Furthermore, instead of counting an hour or so for a high precision wavescan, we only need to count for a few hundred second to get excellent sensitivity for our ~1000 PPM sulfur measurements!
When we then analyzed our ND70 SIMS glass standard for all 5 spectrometers, which reportedly has a sulfur concentration of 900 PPM, we obtained the following results:
Un 24 ND70, Results in Elemental Weight Percents
SPEC: Si Al Fe Mg Ca Na K Ti P Mn O F
TYPE: SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC
AVER: .000 .000 .000 .000 .000 .000 .000 .000 .000 .000 .000 .000
SDEV: .000 .000 .000 .000 .000 .000 .000 .000 .000 .000 .000 .000
ELEM: S S S S S
BGDS: LIN LIN LIN LIN LIN
TIME: 160.00 160.00 160.00 160.00 160.00
BEAM: 49.62 49.62 49.62 49.62 49.62
ELEM: S S S S S SUM
XRAY: (ka) (ka) (ka) (ka) (ka)
492 .072 .073 .069 .077 .073 .364
493 .069 .075 .073 .070 .074 .362
494 .071 .076 .068 .068 .064 .347
495 .072 .072 .070 .062 .071 .347
496 .070 .074 .070 .069 .072 .354
497 .071 .072 .070 .064 .066 .344
498 .073 .070 .069 .065 .074 .352
499 .073 .076 .072 .069 .069 .358
500 .073 .069 .070 .067 .062 .341
501 .071 .071 .072 .074 .073 .360
AVER: .071 .073 .070 .068 .070 .353
SDEV: .001 .002 .002 .004 .004 .008
SERR: .000 .001 .001 .001 .001
%RSD: 2.00 3.28 2.40 6.43 6.12
STDS: 730 730 730 730 730
STKF: .5044 .5044 .5044 .5044 .5044
STCT: 152.32 451.70 499.97 127.50 166.44
UNKF: .0007 .0007 .0007 .0007 .0007
UNCT: .22 .65 .70 .17 .23
UNBG: .10 .18 .23 .08 .11
ZCOR: 1.0000 1.0000 1.0000 1.0000 1.0000
KRAW: .0014 .0014 .0014 .0014 .0014
PKBG: 3.23 4.58 4.08 3.24 3.09
Remember spectrometers 1, 3, 4, and 5 were detuned from the pyrite peak position to accomodate the range of sulfur oxidation states from pyrrhotite to anhydrite.
Anyway, none of the spectrometers gave us a concentration close to 900 PPM, but of course we analyzed sulfur on all 5 spectrometers so 99%+ of the matrix is missing (that's why the ZCOR is 1.0000 because the software thinks this is a pure sulfur sample, with a total of .353 weight percent!
So let's specify a nominal basaltic glass composition, so the matrix correction can perform its physics magic, and now we obtain the following results:
Un 24 ND70, Results in Elemental Weight Percents
SPEC: Si Al Fe Mg Ca Na K Ti P Mn O F
TYPE: SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC SPEC
AVER: 23.750 7.441 9.203 4.046 7.947 1.944 .158 1.109 .087 .170 43.964 .045
SDEV: .000 .000 .000 .000 .000 .000 .000 .000 .000 .000 .000 .000
ELEM: S S S S S
BGDS: LIN LIN LIN LIN LIN
TIME: 160.00 160.00 160.00 160.00 160.00
BEAM: 49.62 49.62 49.62 49.62 49.62
ELEM: S S S S S SUM
XRAY: (ka) (ka) (ka) (ka) (ka)
492 .089 .090 .085 .095 .091 100.314
493 .085 .093 .091 .087 .091 100.311
494 .088 .093 .083 .084 .079 100.294
495 .088 .089 .087 .077 .088 100.294
496 .086 .091 .087 .085 .088 100.302
497 .088 .089 .087 .079 .082 100.290
498 .091 .086 .085 .080 .092 100.300
499 .090 .094 .089 .085 .085 100.307
500 .090 .086 .086 .082 .077 100.286
501 .088 .087 .089 .091 .090 100.310
AVER: .088 .090 .087 .084 .086 100.301
SDEV: .002 .003 .002 .005 .005 .010
SERR: .001 .001 .001 .002 .002
%RSD: 2.01 3.28 2.40 6.43 6.12
STDS: 730 730 730 730 730
STKF: .5044 .5044 .5044 .5044 .5044
STCT: 152.32 451.70 499.97 127.50 166.44
UNKF: .0007 .0007 .0007 .0007 .0007
UNCT: .22 .65 .70 .17 .23
UNBG: .10 .18 .23 .08 .11
ZCOR: 1.2340 1.2340 1.2340 1.2340 1.2340
KRAW: .0014 .0014 .0014 .0014 .0014
PKBG: 3.23 4.58 4.08 3.24 3.09
Notice two things, first our ZCOR values for sulfur Ka are now 1.2340 (as opposed to 1.0000), and spectrometer 2, which was tuned to the pyrite peak position (spectrometer offset equals 0), now gives us 900 PPM *and* this is the highest concentration measured, so we might assume that this is the correct spectrometer position for this sample's oxidation state.
Does this mean that SIMS standard glass ND70 is quite reduced with a valence similar to pyrite? I do not know (has anyone out there done EXAFS or XANES work on this glass standard?), but I will say these results are intriguing and maybe there is some value is setting up one's microprobe as a sulfur "multi-collector" using all 5 spectrometers.