If there's one thing we can be sure of in this world, it's that pure synthetic quartz is one atom of Si to two atoms of O (SiO2). That yields a theoretical composition of Si 46.74 wt% O 53.26 wt%.
Now in the case of our synthetic quartz, I have the following composition noted in our standard database:
St 14 SiO2 synthetic
TakeOff = 40.0 KiloVolt = 15.0 Density = 2.650 Type = oxide Mount = alkali, hydrous, carbonate, sil-sul, misc
Specimen from ESPI, 99.99%, EPMA (UCB): Al2O3 ~ 0.01%
Catalog #K4699M
ICP-MS (Alan Konig, PPM) Ti=1.42, Li=3.18, Cu=1.26, Mn=2.93, Pb=0.11, Zn=2.09, K=0.03
Flame AA (Chris Lewis, PPM):
Al=15 +/- 5
Fe=6 +/- 3
Mn=1.5 +/- 0.3
Na=5 +/- 3
Li= 2.3 +/-
Oxide and Elemental Composition
Average Total Oxygen: 53.255 Average Total Weight%: 100.000
Average Calculated Oxygen: 53.258 Average Atomic Number: 10.805
Average Excess Oxygen: -.003 Average Atomic Weight: 20.029
ELEM: SiO2 Al2O3 O
XRAY: ka ka ka
OXWT: 99.994 .009 -.003
ELWT: 46.740 .005 53.255
KFAC: .4101 .0000 .2664
ZCOR: 1.1397 1.2239 1.9993
AT% : 33.332 .004 66.665
24 O: 12.000 .001 24.000
Because we have around a few tens of PPM of Al in the matrix (from hydrothermal growth) depending on the technique, it's not exactly stoichiometric. But as to the Si content we can have pretty darn good accuracy just by assuming single crystal stoichiometry.
Now I'm not claiming these are the best analyses that could be performed, as I was in a rush trying to get some results before our instrument undergoes a week of maintenance and the beam was not as stable as it should be.
And also I'm not claiming these synthetic silicates are perfectly accurate compositions, as I've simply assumed formula stoichiometry on them, though the NIST glasses should be reasonably accurate. The natural end member diopside is also probably reasonable accurate from assuming stoichiometry, but does have some small tremolite inclusions... hey, it's a natural material!
Now because these analyses were performed on an LPET crystal at 30 nA, we had lots of counts coming into the detector, so when extrapolating from these high count rates in SiO2 to lower Si concentrations in silicates and glasses, the dead time calibrations really do matter. But in any event, assuming our dead time calibrations are reasonably accurate (that is, up to date!), we might ask: do our modern matrix corrections really allow us to extrapolate from pure SiO2 to various silicates and glasses?
The matrix corrections in these various secondary standards ranges from ~20 to 30%. Here is a screen shot of the Evaluate application showing a number of silicates and glass standards using SiO2 as the primary standard:

What do you think? Here's the key:
263 = synthetic Fe2SiO4 (Oak Ridge, Boatner)
275 = synthetic Mn2SiO4 (Oak Ridge, Boatner)
273 = synthetic Mg2SiO4 (Inst. Solid State Physics, H. Takei)
160 = NIST K-412 mineral glass
162 = NIST K-411 mineral glass
215 = BIR-1G basaltic glass (USGS)
358 = natural end member diopside, 600 PPM Fe (Chesterman)
In fact all these secondary standards show 2% or better relative accuracy when using SiO2 as a primary standard. I think this data bodes well for our high purity synthetic standards global efforts as described here:
https://probesoftware.com/smf/index.php?topic=1415.0Do you have the Probe for EPMA software? It might be instructive to run a bunch of your silicate and oxide standards and calculate the results in the Evaluate application. It won't tell you whether it's your dead time calibrations, or your standard compositions, that are inaccurate, but the first one is easily remedied:
https://probesoftware.com/smf/index.php?topic=1160.0and the latter we're working on:
https://probesoftware.com/smf/index.php?topic=1442.0Please join us in our efforts to place microanalysis on a more sound footing by joining the FIGMAS group and/or contacting Will Nachlas at the University of Wisconsin to see how you can participate in these efforts.
Oh, and please run some dead time calibrations on your instrument! Your analyses will be better for it no matter what standards you utilize.