Theo,
Why do you want to set Xhi to 999 for stand-by?
Were you considering an alternative? You see with 999 on Xhi the beam will be hitting some graphite wall of the tube, also C2 4000 is quite powerful field. Actually with Xhi 999, you even don't need to touch the C2 as beam will not make to the C2 aperture.
Two years ago we were using C2 4000 for our "standby" (TaskEnd) procedure, but we had moved to recommended by Cameca C2 2500 as that is closer to 1900 (around this C2 value the most commonly used 20nA current is produced with all apertures). We had noticed that the smaller C2 difference, when switching the currents, the lesser drift of current appears -- condensers needs time to stabilize.
For our standby procedure we do not deflect beam, but "park" the beam on the 0,0,0 position of the stage, thus beam is not hitting neither Faraday plate, neither the tube wall - with such procedure We got so clean column that after 1.5 year of our previous tip life, when service was exchanging the tip and opening the column - engineer was highly surprised how clean the tube was (previously, it was dirty like hell, when beam was been left deflected). Beam is also deflected by closing EP6 automatically. (and we were closing EP6 as TaskEnd previously, now we leave it open)