Author Topic: Halogen correction  (Read 3076 times)
Halogen correction
« on: September 30, 2019, 05:06:14 PM »
I am not sure if its due to me doing something wrong, but I am finding something strange going on with the halogen correction at the moment, and I don't think I have seen it happen in the past so not sure what has changed or what I might be doing wrong now.
I am analysing apatite which I have done many times in the past, and typically I analyse a bunch of apatites as "standards" as well for QC checks. What I am finding is when I process these secondary standards it is subtracting double the amount of oxygen it should with the halogen correction enabled (ie ~3.2 oxygens). When I analyse the same apatites as "unknowns" and process them as "unknowns" it subtracts the correct amount of oxygen (ie ~1.6 oxygens). Its like there is a double flag subtracting when they are analysed as standards. This is with "calculate stoichiometric oxygen" and "display as oxides" ticked in calculation options.
As "Standard":
St  707 Set   1 Apatite, Results in Oxide Weight Percents
SPEC:        O     CO2
TYPE:     CALC    SPEC
AVER:   -3.212    .410
SDEV:     .035    .000
ELEM:      CaO     MgO    TiO2    SiO2   Al2O3     FeO     MnO      Cl       F   La2O3   Ce2O3   Nd2O3    P2O5    Na2O     SO3     SrO    Y2O3   SUM
    45 53.3862  .04260  .00000  .46065  .00042  .00000  .00696  .36325 3.64139  .51881  .70278  .17773 41.0875  .16914  .25996  .04540  .11289 98.1913
    46 53.7970  .03712  .00000  .46968  .01290  .05835  .00016  .37404 3.75368  .47391  .67576  .13349 41.4981  .18669  .19775  .05872  .06839 98.9616
    47 53.2852  .04685  .00000  .46397  .00000  .05181  .00358  .33982 3.75822  .45751  .64484  .21806 41.6324  .18523  .12736  .05559  .09889 98.5411
    48 52.5823  .05287  .00000  .47128  .00318  .03399  .00110  .37115 3.58467  .50989  .64063  .15948 41.3289  .20837  .15113  .06991  .11026 97.5169
AVER:  53.2627  .04486  .00000  .46639  .00412  .03604  .00295  .36206 3.68449  .49003  .66600  .17219 41.3868  .18736  .18405  .05741  .09761 98.3027
SDEV:   .50468  .00666  .00000  .00495  .00602  .02614  .00304  .01552  .08572  .02910  .02909  .03556  .23497  .01610  .05844  .01009  .02040  .61124
SERR:   .25234  .00333  .00000  .00248  .00301  .01307  .00152  .00776  .04286  .01455  .01455  .01778  .11749  .00805  .02922  .00505  .01020
As "unknown":
Un    3 Durango2, Results in Oxide Weight Percents
SPEC:        O
TYPE:     CALC
AVER:  -1.504
SDEV:     .034
   107 54.2091  .04725  .00000  .41866  .00020  .01547  .00000  .36244 3.32484  .44270  .61604  .24827 41.4422  .17199  .21913  .03436  .11979 100.191
   108 53.2595  .04786  .00000  .41589  .01463  .04665  .01813  .36427 3.36025  .45606  .57950  .21030 41.5254  .17116  .25589  .05795  .10014 99.3864
   109 53.3979  .05882  .00000  .44552  .01022  .00891  .00090  .39483 3.50690  .46835  .61204  .25679 41.6277  .19061  .21152  .06647  .07528 99.7669
   110 53.0930  .04944  .00000  .45288  .00395  .04572  .01845  .32910 3.26784  .45868  .65655  .26986 41.8062  .19884  .23781  .09074  .09983 99.6286
   111 53.6377  .04895  .00506  .43000  .00708  .00000  .01192  .38136 3.35012  .40432  .61868  .14233 41.9211  .18779  .29513  .07202  .09161 100.108
   112 53.4209  .05359  .00000  .45582  .00696  .04606  .00000  .35811 3.41653  .46710  .66029  .19115 41.2252  .29098  .38589  .07846  .11351 99.6510
   113 53.8312  .05010  .00000  .47780  .00915  .00000  .00000  .40240 3.44283  .56007  .63156  .14622 41.3396  .29092  .42716  .05239  .11348 100.234
   114 53.1826  .05850  .00000  .35239  .01289  .00000  .00401  .35314 3.35610  .35307  .50914  .18175 41.8094  .20547  .17516  .04905  .09443 99.2042
   115 52.7779  .05023  .00000  .34551  .00918  .04377  .00528  .41335 3.27134  .40721  .61721  .20419 41.8926  .31852  .36512  .05145  .07190 99.3740
   116 53.9716  .05552  .00000  .34099  .00507  .02027  .00819  .41868 3.40459  .43163  .59893  .19126 41.5325  .23218  .33564  .07035  .10209 100.191
AVER:  53.4781  .05203  .00051  .41354  .00793  .02269  .00669  .37777 3.37013  .44492  .60999  .20421 41.6122  .22585  .29085  .06232  .09821 99.7736
SDEV:   .43606  .00430  .00160  .04991  .00426  .02078  .00727  .02912  .07476  .05406  .04289  .04357  .23982  .05460  .08433  .01646  .01574  .38624
SERR:   .13789  .00136  .00051  .01578  .00135  .00657  .00230  .00921  .02364  .01710  .01356  .01378  .07584  .01727  .02667  .00520  .00498
Any idea what I am doing wrong...?
« Last Edit: September 30, 2019, 05:28:22 PM by John Donovan »
Re: Halogen correction
« Reply #1 on: October 01, 2019, 08:35:15 AM »
Interesting. Can you email me the MDB file?
« Reply #2 on: October 03, 2019, 07:03:27 PM »
Ok just for the benefit of anyone who might encounter the same problem, it was entirely my own stupid fault.
My apatite "standards" in my Standard.MDB file already had their excess oxygen subtracted, thus when I turned on the halogen correction in a normal run my apatite standards had a double subtraction.
In the image below, on the left is without the excess oxygen corrected in my Standard.MDB file which gives me the correct halogen subtraction if I turn the halogen correction on in a run. Below right is obviously with the excess oxygen correction, resulting in a double subtraction in a run.
« Reply #3 on: October 03, 2019, 09:07:15 PM »
You absolutely are correct in your conclusions.
Many years ago when we started CalcZAF and Probe for EPMA, we decided that it was very important, for geologists at least, to account for chemical substitution of stoichiometric oxygen by elements such as halogens (and additional oxygen from say, ferric iron).
So in the standard compositional database (Standard.exe) one can enter a standard composition with all the elements in ideal stoichchiometric proportions, but we can also enter the excess *or* deficit oxygen in those situations where ideal stoichiometric assumptions are not valid, due to the presence of, for example, ferric iron or halogens.
So yes, when you've already accounted for the excess or deficit oxygen in the standard composition database, applying the halogen correction in Probe for EPMA, results in a double correction of the oxygen-halogen equivalence.
Therefore, if one is utilizing an already halogen corrected composition for a standard, in Standard, there is no need to also apply a halogen correction in Probe for EPMA. Probe for EPMA automatically notes the excess or deficit oxygen (whether it be a magnetite or a fluor-phlogopite), and compensates for the difference in the matrix effects.
However, if one specifies a standard composition in the Standard compositional database where the excess or deficit oxygen is *not* documented, then, yes, one must apply the correction for this situation in Probe for EPMA when this standard is measured as a standard (or as an unknown), whether that situation be excess oxygen from ferric iron:
or for the replacement of oxygen by halogens:
https://probesoftware.com/smf/index.php?topic=92.msg8557#msg8557
Of course for samples measured as unknown samples, these excess or deficit oxygen corrections must always be applied because the program knows nothing (a priori) regarding the sample.
There is one exception I must take to your post however. Your characterization of this question as being "stupid" is simply incorrect. There is nothing obvious about the way in which the ferric or halogen corrections are applied in Probe for EPMA because one would need to carefully parse the physics code on Github (see the ZAFSmp procedure in ZAF.bas) to comprehend exactly how these corrections are applied:
Though please note that these procedures (regardless of the naming convention) are entirely agnostic as to exactly exactly what matrix correction models are being applied, i.e., ZAF vs. phi-rho-z!
The bottom line is: thank-you for noting this quite subtle and yet interesting aspect to the matrix correction models. To be entirely honest, when we first noticed these effects, we weren't too surprised when we noticed that including the deficit oxygen correction in apatites and phlogopites made a significant difference in the results, but we were completely surprised by the effect that excess oxygen from ferric iron had, on not only the iron concentrations, but also on the other elements.
Excellent work.
« Reply #4 on: October 11, 2019, 12:05:40 PM »
To emphasize what Ben Wade found in his investigations, here are examples of a fluor-phlogopite entered as both a "normal" standard composition where the stoichiometric oxygen is assumed to be in nominal proportions to the cations, and also where the fluor-phlogopite is entered with the "deficit" oxygen entered.
So first the fluor-phlogopite entered with the assumption of default oxygen stoichiometry:
St  282 Fluor-phlogopite (synthetic)
TakeOff = 40.0  KiloVolt = 15.0  Density =  2.879  Type = phylosilicate
Grown by S. Wones, Univ of Tenn
Average Total Oxygen:       41.775     Average Total Weight%:  103.789
Average Calculated Oxygen:  41.778     Average Atomic Number:   11.202
Average Excess Oxygen:       -.003     Average Atomic Weight:   20.822
Oxygen Equiv. from Halogen:  3.798
ELEM:     SiO2     MgO   Al2O3     K2O       F       O
XRAY:      ka      ka      ka      ka      ka      ka
OXWT:   42.791  28.700  12.100  11.180   9.020   -.003
ELWT:   20.002  17.307   6.404   9.281   9.020  41.775
KFAC:    .1502   .1224   .0432   .0818   .0249   .1837
ZCOR:   1.3317  1.4138  1.4820  1.1342  3.6217  2.2746
AT% :   14.287  14.285   4.761   4.762   9.525  52.379
24 O:    6.546   6.545   2.182   2.182   4.364  24.000
Note that it totals over 100% because in reality some of the fluorine replaces some of the stoichiometric oxygen.  So what happens when we analyze this as a standard in PFE:
St  282 Set   1 Fluor-phlogopite (synthetic)
(Magnification (default) =      600, Magnification (imaging) =    200)
Formula Based on Sum of Cations = .000   Oxygen Calc. by Stoichiometry
First/Last Date-Time: 03/08/2016 01:07:36 PM to 03/08/2016 01:08:57 PM
Average Total Oxygen:       42.236     Average Total Weight%:  105.056
Average Calculated Oxygen:  42.239     Average Atomic Number:   11.196
Average Excess Oxygen:       -.003     Average Atomic Weight:   20.818
Oxygen Equiv. from Halogen:  3.827  Halogen Corrected Oxygen:   38.409
Oxygen Equivalent from Halogens (F/Cl/Br/I), Not Subtracted in the Matrix Correction
St  282 Set   1 Fluor-phlogopite (synthetic), Results in Elemental Weight Percents
ELEM:        F      Mg      Si      Al       K       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    CALC
TIME:    20.00   20.00   20.00   20.00   20.00     ---
BEAM:    29.99   29.99   29.99   29.99   29.99     ---
ELEM:        F      Mg      Si      Al       K       O   SUM
    16   9.109  17.911  19.906   6.399   9.278  42.058 104.661
    17   9.096  18.065  19.874   6.577   9.333  42.293 105.239
    18   9.056  17.895  20.097   6.476   9.389  42.356 105.270
AVER:    9.087  17.957  19.959   6.484   9.333  42.236 105.056
SDEV:     .028    .094    .121    .090    .056    .157    .343
SERR:     .016    .054    .070    .052    .032    .091
%RSD:      .30     .52     .60    1.38     .60     .37
PUBL:    9.020  17.307  20.002   6.404   9.281  41.775 103.789
%VAR:      .74    3.75    -.21    1.25     .56    1.10
DIFF:     .067    .650   -.043    .080    .052    .461
STDS:      831     284     374     374     374     ---
STKF:    .1545   .1235   .2480   .0703   .1132     ---
STCT:    51.52   39.60   82.71   23.26   37.65     ---
UNKF:    .0251   .1271   .1495   .0436   .0823     ---
UNCT:     8.38   40.77   49.86   14.43   27.37     ---
UNBG:      .18     .70     .95     .53     .75     ---
ZCOR:   3.6155  1.4126  1.3350  1.4873  1.1342     ---
KRAW:    .1627  1.0294   .6028   .6205   .7271     ---
PKBG:    46.46   59.87   53.96   28.34   37.44     ---
We see the total is actually even higher than the assumption of stoichiometry, mostly because the matrix correction for Mg incorrectly calculates the effect due to the extra oxygen. Calculating the same as oxides we get:
St  282 Set   1 Fluor-phlogopite (synthetic), Results in Oxide Weight Percents
ELEM:        F     MgO    SiO2   Al2O3     K2O       O   SUM
    16   9.109  29.702  42.586  12.090  11.176   -.003 104.661
    17   9.096  29.957  42.518  12.428  11.242   -.003 105.239
    18   9.056  29.675  42.995  12.236  11.310   -.003 105.270
AVER:    9.087  29.778  42.700  12.252  11.243   -.003 105.056
SDEV:     .028    .156    .258    .169    .067    .000    .343
SERR:     .016    .090    .149    .098    .039    .000
%RSD:      .30     .52     .60    1.38     .60    -.14
PUBL:    9.020  28.700  42.791  12.100  11.180   -.003 103.789
%VAR:      .74    3.75    -.21    1.25     .56    -.14
DIFF:     .067   1.078   -.092    .151    .063    .000
Note that the excess/deficit oxygen is close to zero, but not a good analysis.  So now, let's turn on the halogen correction in PFE from the Analytical | Analysis Options dialog as seen here:
and analyze this standard again in PFE:
Average Total Oxygen:       38.412     Average Total Weight%:  100.872
Average Calculated Oxygen:  38.415     Average Atomic Number:   11.321
Average Excess Oxygen:       -.003     Average Atomic Weight:   21.059
Oxygen Equiv. from Halogen:  3.735  Halogen Corrected Oxygen:   38.412
Oxygen Equivalent from Halogens (F/Cl/Br/I), Subtracted in the Matrix Correction
    16   8.890  17.751  19.926   6.392   9.282  38.226 100.465
    17   8.879  17.904  19.895   6.570   9.337  38.465 101.049
    18   8.841  17.736  20.116   6.468   9.393  38.545 101.101
AVER:    8.870  17.797  19.979   6.477   9.337  38.412 100.872
SDEV:     .025    .093    .120    .090    .056    .166    .353
SERR:     .015    .054    .069    .052    .032    .096
%RSD:      .29     .52     .60    1.38     .60     .43
%VAR:    -1.66    2.83    -.12    1.13     .61   -8.05
DIFF:    -.150    .490   -.023    .073    .056  -3.363
ZCOR:   3.5293  1.4000  1.3363  1.4856  1.1347     ---
Much better. And here again as oxide concentrations:
    16   8.890  29.436  42.629  12.077  11.181  -3.746 100.465
    17   8.879  29.689  42.561  12.414  11.247  -3.742 101.049
    18   8.841  29.412  43.036  12.222  11.315  -3.726 101.101
AVER:    8.870  29.512  42.742  12.238  11.248  -3.738 100.872
SDEV:     .025    .154    .257    .169    .067    .011    .353
SERR:     .015    .089    .148    .098    .039    .006
%RSD:      .29     .52     .60    1.38     .60    -.29
%VAR:    -1.66    2.83    -.12    1.13     .61137523.80
DIFF:    -.150    .812   -.049    .137    .068  -3.735
The reason we are still slightly high with our total is because the std k-factor from the standard database is slightly wrong due to the excess oxygen assumed in the standard database!
In the next post we'll look at how this should be handled properly...
« Reply #5 on: October 11, 2019, 12:24:32 PM »
So in the previous post we saw how we should *not* be handling the oxygen halogen equivalence in our standard database compositions.  In my own standard database I have the same fluor-phlogopite standard as seen in the previous post, but this time entered (with a different standard number of course), with the missing oxygen specified, as seen here:
St  284 Fluor-phlogopite (halogen corrected)
(applied F=O equivalence)
Average Total Oxygen:       37.980     Average Total Weight%:   99.994
Average Calculated Oxygen:  41.778     Average Atomic Number:   11.324
Average Excess Oxygen:      -3.798     Average Atomic Weight:   21.063
OXWT:   42.791  28.700  12.100  11.180   9.020  -3.798
ELWT:   20.002  17.307   6.404   9.281   9.020  37.980
KFAC:    .1500   .1235   .0432   .0818   .0256   .1630
ZCOR:   1.3338  1.4016  1.4818  1.1346  3.5255  2.3298
AT% :   15.001  14.999   4.999   5.000  10.001  50.000
24 O:    7.200   7.200   2.400   2.400   4.800  24.000
Note that our total is correct because we entered -3.798 oxygen in the standard database as seen here:
Now let's leave the halogen correction in Probe for EPMA turned on, as Ben did originally and see what we get:
St  284 Set   1 Fluor-phlogopite (halogen corrected)
First/Last Date-Time: 03/08/2016 12:46:56 PM to 03/08/2016 12:48:17 PM
Average Total Oxygen:       34.141     Average Total Weight%:   95.774
Average Calculated Oxygen:  37.939     Average Atomic Number:   11.464
Average Excess Oxygen:      -3.798     Average Atomic Weight:   21.336
Oxygen Equiv. from Halogen:  3.702  Halogen Corrected Oxygen:   34.141
St  284 Set   1 Fluor-phlogopite (halogen corrected), Results in Elemental Weight Percents
     7   8.832  17.174  19.891   6.361   9.282  34.009  95.549
     8   8.813  17.127  19.932   6.382   9.412  34.078  95.743
     9   8.729  17.137  20.047   6.512   9.269  34.337  96.030
AVER:    8.791  17.146  19.957   6.418   9.321  34.141  95.774
SDEV:     .055    .025    .081    .082    .079    .173    .242
SERR:     .032    .014    .047    .047    .046    .100
%RSD:      .62     .14     .40    1.27     .85     .51
PUBL:    9.020  17.307  20.002   6.404   9.281  37.980  99.994
%VAR:    -2.54  (-.93)    -.23     .22     .43  -10.11
DIFF:    -.229  (-.16)   -.045    .014    .040  -3.839
STCT:    51.52   41.12   82.71   23.26   37.65     ---
UNKF:    .0256   .1235   .1494   .0434   .0821     ---
UNCT:     8.54   41.14   49.84   14.36   27.32     ---
UNBG:      .21     .66     .92     .52     .77     ---
ZCOR:   3.4329  1.3880  1.3354  1.4798  1.1350     ---
KRAW:    .1658  1.0004   .6025   .6173   .7256     ---
PKBG:    41.65   63.50   55.28   28.88   36.40     ---
As Ben Wade saw, we're now getting a *double* correction for the halogen effect. Once because the missing oxygen is already accounted for in the standard database and again because we turned on the halogen correction in Probe for EPMA.  And here is the same standard analyzed in PFE as oxide concentrations (and still wrong!):
St  284 Set   1 Fluor-phlogopite (halogen corrected), Results in Oxide Weight Percents
     7   8.832  28.480  42.554  12.019  11.181  -7.517  95.549
     8   8.813  28.402  42.642  12.058  11.338  -7.509  95.743
     9   8.729  28.418  42.887  12.304  11.166  -7.473  96.030
AVER:    8.791  28.433  42.694  12.127  11.228  -7.500  95.774
SDEV:     .055    .041    .173    .154    .095    .023    .242
SERR:     .032    .024    .100    .089    .055    .013
%RSD:      .62     .14     .40    1.27     .85    -.31
PUBL:    9.020  28.700  42.791  12.100  11.180  -3.798  99.994
%VAR:    -2.54  (-.93)    -.23     .22     .43   97.48
DIFF:    -.229  (-.27)   -.097    .027    .048  -3.702
What we should really be doing is specifying the deficit oxygen in the standard database composition, and simply *not* utilize the halogen correction in PFE as seen here:
Average Total Oxygen:       37.939     Average Total Weight%:   99.963
Average Calculated Oxygen:  41.737     Average Atomic Number:   11.327
Average Excess Oxygen:      -3.798     Average Atomic Weight:   21.066
Oxygen Equiv. from Halogen:  3.801  Halogen Corrected Oxygen:   34.138
     7   9.071  17.344  19.872   6.370   9.277  37.825  99.759
     8   9.050  17.296  19.913   6.391   9.408  37.885  99.943
     9   8.961  17.304  20.028   6.521   9.265  38.108 100.186
AVER:    9.027  17.315  19.937   6.427   9.317  37.939  99.963
SDEV:     .059    .026    .081    .082    .079    .149    .214
SERR:     .034    .015    .047    .047    .046    .086
%RSD:      .65     .15     .41    1.27     .85     .39
%VAR:      .08   (.04)    -.32     .36     .39    -.11
DIFF:     .007   (.01)   -.065    .023    .036   -.041
ZCOR:   3.5251  1.4016  1.3341  1.4819  1.1345     ---
Now that's better!  And again here as oxides:
     7   9.071  28.762  42.513  12.036  11.176  -3.798  99.759
     8   9.050  28.682  42.601  12.075  11.332  -3.798  99.943
     9   8.961  28.694  42.846  12.321  11.161  -3.798 100.186
AVER:    9.027  28.713  42.653  12.144  11.223  -3.798  99.963
SDEV:     .059    .043    .173    .155    .095    .000    .214
SERR:     .034    .025    .100    .089    .055    .000
%RSD:      .65     .15     .41    1.27     .85     .00
%VAR:      .08   (.04)    -.32     .36     .39     .00
DIFF:     .007   (.01)   -.138    .044    .043    .000
Voila!
OK, I probably didn't explain this well, so if anyone has any questions at all please ask away!
I'll sum up by saying that these considerations also apply to any other mineral compositions that contain halogens and those halogens replace some of the assumed stoichiometric oxygen, e.g., chlor-apatites as Ben Wade showed. Of course if you want to measure oxygen on all your minerals, that would work too, but that's another discussion!
So in summary:
1. Enter your standard compositions with any deficit (or excess oxygen) accounted for in the standard database.
2. Do not apply the halogen correction to standards when the halogen replacement of oxygen has already occurred in the standard database (this is actually Donovan's fault because he really should not have made the halogen correction a global parameter, but he's a lazy slacker, so what did you expect?).   ;)
3. Do apply the halogen correction to your unknown chlor-apatites and fluor-phlogopites!  Then you will automatically compensate for the replacement of stoichiometric oxygen by halogens, and this effect will be included in the matrix correction properly.
« Last Edit: October 11, 2019, 03:47:02 PM by Probeman »
« Reply #6 on: June 17, 2020, 11:04:40 AM »
Because there are multiple ways to do these oxygen-halogen corrections, it should be pointed out that one can use the oxygen-halogen correction on standard samples (when the oxygen-halogen correction is already accounted for in the standard composition), but then you want to be sure to use the actual oxygen value from the standard database as seen here (by not calculating oxygen by stoichiometry in PFE):
St  284 Set   4 Fluor-phlogopite (halogen corrected), Results in Elemental Weight Percents
ELEM:       Si       F      Cl      Mg      Ca       P      Al       K       O
TYPE:     ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    ANAL    SPEC
BGDS:      LIN     LIN     LIN     LIN     LIN     EXP     LIN     LIN
TIME:    20.00   20.00   20.00   30.00   30.00   20.00   20.00   20.00     ---
BEAM:    30.18   30.18   30.18   30.18   30.18   30.18   30.18   30.18     ---
ELEM:       Si       F      Cl      Mg      Ca       P      Al       K       O   SUM
   500  19.746   9.086    .006  17.146    .006    .001   6.611   9.431  37.980 100.014
   501  20.414   8.670    .002  16.990    .011   -.007   6.598   9.339  37.980  99.998
   502  20.307   9.530    .004  17.256    .001    .003   6.602   9.414  37.980 101.096
   503  19.871   8.945    .005  17.271    .005   -.004   6.609   9.387  37.980 100.068
   504  21.925   9.162    .015  15.628    .022    .002   6.227   8.441  37.980  99.400
   505  21.238   8.648    .002  16.261    .004    .002   6.174   9.078  37.980  99.388
AVER:   20.584   9.007    .006  16.758    .008    .000   6.470   9.181  37.980  99.994
SDEV:     .842    .332    .005    .669    .007    .004    .210    .385    .000    .623
SERR:     .344    .135    .002    .273    .003    .002    .086    .157    .000
%RSD:     4.09    3.68   84.10    3.99   92.35-1540.98    3.24    4.20     .00
PUBL:   20.002   9.020    n.a.  17.307    n.a.    n.a.   6.404   9.281  37.980  99.994
%VAR:     2.91  (-.15)     ---   -3.17     ---     ---    1.03   -1.07     .00
DIFF:     .582  (-.01)     ---   -.549     ---     ---    .066   -.100    .000
STDS:      160     284     285      12     285     285     160     374     ---
STKF:    .1621   .0256   .0602   .4736   .3596   .1601   .0335   .1132     ---
STCT:   1749.1    71.9  2270.1 18778.0  8626.5  3449.1  5319.3  1650.6     ---
UNKF:    .1549   .0255   .0000   .1196   .0001   .0000   .0439   .0809     ---
UNCT:   1671.0    71.8     1.8  4741.7     1.7      .0  6979.3  1179.7     ---
UNBG:      4.2     3.7    10.4    20.6    17.8     3.5    81.9     9.0     ---
ZCOR:   1.3295  3.5271  1.2181  1.4013  1.1187  1.4130  1.4733  1.1351     ---
KRAW:    .9553   .9982   .0008   .2525   .0002   .0000  1.3121   .7147     ---
PKBG:   399.72   20.80    1.18  231.91    1.10    1.00   86.24  133.02     ---
INT%:     ----     .00    ----    ----    ----    ----    ----    ----     ---
TDI%:     .249   -.837  -5.332    .678    .245    .000    .000    .000     ---
DEV%:       .2     1.8     6.2      .1     3.7      .0      .0      .0     ---
TDIF:  LOG-LIN LOG-LIN LOG-LIN LOG-LIN LOG-LIN    ----    ----    ----     ---
TDIT:    44.00   41.00   41.33   52.67   56.33     .00     .00     .00     ---
TDII:    1674.    75.1    11.2   4759.    19.4    ----    ----    ----     ---
TDIL:     7.42    4.32    2.41    8.47    2.96    ----    ----    ----     ---
St  284 Set   4 Fluor-phlogopite (halogen corrected), Results in Oxide Weight Percents
ELEM:     SiO2       F      Cl     MgO     CaO    P2O5   Al2O3     K2O       O   SUM
   500  42.245   9.086    .006  28.433    .008    .003  12.492  11.360  -3.619 100.014
   501  43.672   8.670    .002  28.174    .016   -.015  12.467  11.250  -4.239  99.998
   502  43.443   9.530    .004  28.616    .002    .007  12.474  11.340  -4.319 101.096
   503  42.512   8.945    .005  28.640    .006   -.009  12.488  11.307  -3.826 100.068
   504  46.904   9.162    .015  25.915    .030    .005  11.766  10.167  -4.564  99.400
   505  45.437   8.648    .002  26.965    .005    .005  11.666  10.935  -4.276  99.388
AVER:   44.035   9.007    .006  27.791    .011   -.001  12.225  11.060  -4.141  99.994
SDEV:    1.800    .332    .005   1.110    .010    .009    .396    .464    .349    .623
SERR:     .735    .135    .002    .453    .004    .004    .162    .190    .143
%RSD:     4.09    3.68   84.10    3.99   92.35-1540.98    3.24    4.20   -8.44
PUBL:   42.791   9.020    n.a.  28.700    n.a.    n.a.  12.100  11.180  -3.798  99.994
%VAR:     2.91  (-.15)     ---   -3.17     ---     ---    1.03   -1.07    9.03
DIFF:    1.244  (-.01)     ---   -.910     ---     ---    .125   -.120   -.343
Because, if you use the PFE option to calculate oxygen by stoichiometry in a standard sample (and the standard composition is already corrected for the oxygen-halogen correction in the standard database), the oxygen-halogen correction in PFE is incorrectly applied twice! Hopefully this makes sense after one has tried this out for themselves in PFE.
However (and I have not tested this) but I suspect if the standard composition in the standard database is *not* already oxygen-halogen corrected, then applying the oxygen-halogen correction to a standard sample in PFE will work fine.  I have my fluor-phlogopite standard entered both with and without the oxygen-halogen correction (std #282 and #284) in the default standard database.  One of these days I will get around to testing this!
The bottom line is, just use the oxygen-halogen correction for your unknowns and all will be fine!
« Last Edit: June 17, 2020, 11:21:01 AM by John Donovan »
« Reply #7 on: June 17, 2020, 02:23:04 PM »
Also, please remember that all corrections in CalcZAF/Probe for EPMA are performed during the matrix iteration procedures, including these oxygen-halogen calculations.  What these oxygen-halogen calculations are changing during the matrix iteration is not the halogen concentrations, but the "excess" oxygen concentration from stoichiometric oxygen.
Here is the code for the calculation of stoichiometric oxygen and the calculations for subtracting oxygen based on the halogen equivalent, and/or adding excess oxygen from Fe2O3.
' Calculate amount of stoichiometric oxygen and add to total
r1!(zaf.in0%) = 0#
For i% = 1 To zaf.in1%
r1!(zaf.in0%) = r1!(zaf.in0%) + r1!(i%) * zaf.p1!(i%)
Next i%
' Calculate equivalent oxygen from halogens and subtract from calculated oxygen if flagged
If UseOxygenFromHalogensCorrectionFlag Then
r1!(zaf.in0%) = r1!(zaf.in0%) - ConvertHalogensToOxygen(zaf.in1%, sample(1).Elsyms$(), sample(1).DisableQuantFlag%(), r1!())
If ierror Then Exit Sub
' Calculate excess oxygen from ferric Fe
If sample(1).FerrousFerricCalculationFlag Then
Call ConvertFerrousFerricRatioFromComposition(zaf.in0%, zaf.Z%(), zaf.atwts!(), r1!(), sample(1).numcat%(), sample(1).numoxd%(), zaf.p1!(), sample(1).DisableQuantFlag%(), sample(1).FerrousFerricTotalCations!, sample(1).FerrousFerricTotalOxygens!, tFerricToTotalIronRatio!, tFerricOxygen!, tFe_as_FeO!, tFe_as_Fe2O3!)
r1(zaf.in0%) = r1(zaf.in0%) + tFerricOxygen! / 100#
' Add to sum
zaf.ksum! = zaf.ksum! + r1!(zaf.in0%)
It's complicated!     ;D
The complete matrix correction code is available here:
« Last Edit: June 17, 2020, 07:25:12 PM by John Donovan »
« Reply #8 on: June 28, 2020, 11:47:39 AM »
Julien Allaz and I have been discussing the oxygen-halogen correction in PFE and wanted to share our recent efforts to improve the way PFE handles oxygen-halogen containing samples.  It's a little complicated.
First of all it should be noted that the program properly handles the oxygen-halogen correction on unknown samples for all situations, particularly when oxygen is calculated by stoichiometry. It also handles the situation where an unknown is calculated by simply specifying the amount of oxygen. But of course in that case, the oxygen-halogen correction has already been applied.
PFE also handles the situation where a standard is analyzed and the oxygen in the standard database was *not* already corrected for the oxygen-halogen equivalence (#282). As seen here:
St  282 Set   2 Fluor-phlogopite (synthetic), Results in Elemental Weight Percents
ELEM:        F      Si      Mg      Al       K       O
TYPE:     ANAL    SPEC    SPEC    SPEC    SPEC    CALC
BGDS:      LIN
TIME:    20.00     ---     ---     ---     ---     ---
BEAM:    29.99     ---     ---     ---     ---     ---
ELEM:        F      Si      Mg      Al       K       O   SUM
     6   8.453  20.002  17.307   6.404   9.281  38.218  99.665
     7   8.766  20.002  17.307   6.404   9.281  38.086  99.846
     8   8.536  20.002  17.307   6.404   9.281  38.183  99.713
AVER:    8.585  20.002  17.307   6.404   9.281  38.163  99.741
SDEV:     .162    .000    .000    .000    .000    .068    .094
SERR:     .094    .000    .000    .000    .000    .039
%RSD:     1.89     .00     .00     .00     .00     .18
PUBL:    9.020  20.002  17.307   6.404   9.281  41.775 103.789
%VAR:    -4.83     .00     .00     .00     .00   -8.65
DIFF:    -.435    .000    .000    .000    .000  -3.612
STDS:      835     ---     ---     ---     ---     ---
Note that the "published" (PUBL) amount of oxygen from the standard database is based on the assumed cation stoichiometry, and is not correct, but the halogen correction does make the proper correction.
Previously if the standard composition was already corrected for the oxygen-halogen correction as in the following standard (#284), the program performed a double correction of the oxygen-halogen equivalence (see above posts). However, Julien and I added some code that checks for a negative excess oxygen from the standard database, and if it is negative, it zeros out the deficit oxygen before entering the matrix correction routine. The new code is seen here:
If sample(1).Type% = 1 Then
analysis.WtPercents!(chan%) = ConvertTotalToExcessOxygen!(Int(1), sample(), stdsample())
If UseOxygenFromHalogensCorrectionFlag And sample(1).OxideOrElemental% = 1 Then
If analysis.WtPercents!(chan%) < 0# Then analysis.WtPercents!(chan%) = 0#  ' zero out oxygen equivalence of halogen from standard database
' For unknowns, use specified oxygen weight percent
Else
analysis.WtPercents!(chan%) = stdsample(1).ElmPercents!(ip%)
Here is an example of a standard, already corrected for the oxygen equivalence of halogens, analyzed with the oxygen-halogen correction:
St  284 Set   2 Fluor-phlogopite (halogen corrected), Results in Elemental Weight Percents
BGDS:      LIN
BEAM:    30.01     ---     ---     ---     ---     ---
     9   8.806  20.002  17.307   6.404   9.281  38.069  99.870
    10   8.832  20.002  17.307   6.404   9.281  38.059  99.885
    11   8.824  20.002  17.307   6.404   9.281  38.062  99.880
AVER:    8.821  20.002  17.307   6.404   9.281  38.063  99.878
SDEV:     .013    .000    .000    .000    .000    .006    .008
SERR:     .008    .000    .000    .000    .000    .003
%RSD:      .15     .00     .00     .00     .00     .01
PUBL:    9.020  20.002  17.307   6.404   9.281  37.980  99.994
%VAR:    -2.21     .00     .00     .00     .00     .22
DIFF:    -.199    .000    .000    .000    .000    .083
STKF:    .1715     ---     ---     ---     ---     ---
STCT:    59.95     ---     ---     ---     ---     ---
Note that both the published concentration of oxygen is corrected for in the standard database, and the stoichiometric oxygen in the standard is properly corrected as expected.
Part of the reason for this issue is that many, many years ago we decided to allow the user to handle the deficit (or excess) oxygen in the standard compositional database, but then include that deficit (or excess) oxygen in the matrix calculation. For example, when analyzing magnetite, the default is to include the excess oxygen by simply loading the speciifed oxygen concentration from the standard database.
But (and here it gets complicated), if the user decides to simply calculate oxygen by stoichiometry for this standard (instead of loading the specified oxygen from the standard database as is the default), the program automatically includes the deficit (or excess) oxygen from the standard database.
But if the standard composition in the standard database is already corrected for the deficit (or excess) oxygen, this results in an double correction during the matrix correction. The new code above should prevent this double correction.
Re: Halogen/Sulfur correction
« Reply #9 on: March 03, 2023, 08:36:32 AM »
This isn't a post exactly on the halogen-oxygen equivalence correction, but it is very closely related so I thought I would post it here.
That is, just as the halogen-oxygen correction handles the excess oxygen from when halogens are present and oxygen is being calculated by stoichiometry, e.g., when analyzing chlor-apatites or fluor-phlogopites and the halogens are replacing some of the oxygen in the material as described in the previous posts, we can also have a situation where (reduced) sulfur can also replace stoichiometric oxygen in some materials, as recently pointed out to me by Julien Allaz.
Here is an example of an experimental glass from Julien Allaz and Stefan Pitsch with a significant amount (~4 wt%) of sulfur is present and reduced to a -2 charge valance (specified from the Elements/Cations dialog) and where oxygen is calculated by stoichiometry:
Note the high totals due to the fact that sulfur is actually replacing some of the oxygen in the material. So what do we do? Well beginning with the latest version of the Probe for EPMA software, we can now specify a correction for sulfur as an anion, just as we have previously discussed for halogens as seen here from the Analytical | Analysis Option menu dialog:
Next we re-calculate our analysis and we now obtain the following (much improved) results:
Wait! There's more... in addition we can also output these sulfur as an anion calculations by selecting these highlighted options in the "user specified output" format export dialog. Which is available by right clicking the sample list in the Analyze! window or from the Output menu:
And here are the exported results in Excel:
Also, for completeness sake we added this same calculation flag to the free CalcZAF software as shown here:
This whole thing might seem a bit esoteric for some, but we at Probe Software thought it was worth implementing and we thank Julien and Stefan for pointing out this issue.  This feature is now available in version 13.2.7 of Probe for EPMA and can be obtained from the Help | Update Probe for EPMA menu as usual.
« Last Edit: March 03, 2023, 12:47:02 PM by John Donovan »
« Reply #10 on: March 04, 2023, 10:04:27 AM »
In addition to setting the sulfur charge valence to a negative number for the sulfur-oxygen correction, it also occurs to me that one should also adjust the sulfur cation ratio if sulfur is an anion as shown here:
« Last Edit: March 04, 2023, 11:43:55 AM by John Donovan »