Hi John - We are still having some issues with our WDS/EDS integration. When we calibrate with EDS channels, everything seems to proceed as it should until completion, then the attached error message appears.
Also, we do not receive any data when we use EDS channels in ProbeImage either, which may be a related problem.
Hi Rob,
It should all be working. Many labs use the Bruker EDS with WDS in PFE and it works great for major and minor elements.
First thing is to update PFE using the Help menu (current version is 11.9.7) and create a new MDB and see if you still have an issue getting the net intensities back from Bruker. Also check your Probewin.ini file and make sure the EDSSpectraNetIntensityInterfaceType keyword is set properly like this:
EDSSpectraNetIntensityInterfaceType=2 ; 0 = Demo, 1 = Edax, 2 = Bruker, 3 = Oxford, 4 = Unused, 5 = Thermo NSS, 6 = JEOL
Next what version of Bruker Esprit are you using? Third, I assume your Bruker license has the LAN and API license options? I assume so because you can acquire the EDS spectra fine with PFE.
Fourth, your installation specialist is Paul Carpenter, and he should be able to help you, but he is at M&M this week. However, I will ping him and cc you.
On the EDS mapping acquisition with Bruker, we decided to skip the EDS ROI acquisition feature in Probe Image because these mapping intensities cannot easily be quantified, and instead we are currently implementing a full EDS spectrum imaging acquisition method in PI (with Thermo and JEOL also) integrated with the WDS mapping, which can be quantified in CalcImage. As soon as that is ready we will get that new version to you. In the meantime you can get a free update for Probe Image here:
http://probesoftware.com/ftp/PI-Update.zipI'll send you the unzip password by email.
john