I am trying to determine MACs for Cka in U using the approach outlined in the XMAC program of SAMx.  Unfortunately in UC, at higher voltages there is a U peak (N iv O iv)on the PC2 crystal that nearly completely obscures the Cka peak.  However, by measuring U on its usual crystal (PET or QTZ), I can know how much U is in the sample.  By knowing the U concentration in the sample, the program should be able to determine the size of the U peak (in cps/na) that must be obscuring the C peak and subtract it out.  The remaining counts should be those from CKa.
BUT……
1.   If the U obscures the C peak, I cannot empirically determine an APF, therefore I must use the integral.  Correct?
2.   In order to use the xMAC program, I need to know the background corrected Cka in cps/na in U without the added counts from U.  How do I find out the magnitude of the U overlap correction so that I know how much cps/na to remove from the C signal?  Is it available in the software?  Do you think this is the correct approach?
Thanks in advance for your thoughts!
Karen
Re: Light element MACs when there is an overlapping element
« Reply #1 on: April 12, 2019, 12:19:52 PM »
BUT……
Karen
Hi Karen,
Wow, this is a tough one I must say.  I saw it yesterday when you posted it, but needed to think about it overnight. I think I have some solutions for you.
For those that don't know, the XMAC program is a small app distributed by SAMX based on code by Pouchou and Pichoir that allows the user to calculate MACs (mass absorption coefficients) from intensity measurements made over range of beam energies in a compound containing the emitting element and the absorbing element.  It works best for low energy emission lines.  For these low energy emission lines it generally produces MAC values that are quite accurate. For a list of some empirically determined MACs generated by this app, see the Empirical MACs menu in CalcZAF or Probe for EPMA. Here's an example:
https://probesoftware.com/smf/index.php?topic=667.msg4064#msg4064
Now as far as the interference of U on C Ka in determining the MAC, I think the solution is going to be that you'll have to measure both C Ka and U Ma quantitatively at each beam energy and specify the interference correction(s).  How does this help, well it's not generally known but if you specify interference corrections in Probe for EPMA, the correction for spectral interference occurs during the matrix correction quantitatvely, so in the end you have not only concentrations corrected for interferences, but *also* you'll have k-ratios corrected for interference as well!  See note below.
By the way, this is why acquiring thin film intensities in Probe for EPMA works so well for spectral interferences on thin film samples. The k-ratios exported into STRATAGem are *already* corrected for spectral interferences!
Now XMAC doesn't care what units the intensities are in, so you could just take the spectral interference corrected k-ratios from Probe for EPMA and utilize those intensities for the empirical determination of your MAC.  It might end up being an interative process because you might want to enter the empirical MAC determined from XMAC into PFE to improve the accuracy of the calculation.
On the APF question for changes in the carbon peak shape, I think you should just utilize the integrated intensities option in Probe for EPMA so that you won't need to use APF values at all.  See the option Use Integrated Intensities in the Elements/Cations dialog. Here's an example using integrating intensities on sulfur and also aggregating multiple spectrometers:
https://probesoftware.com/smf/index.php?topic=42.msg4932#msg4932
Note: I looked at the (default) NIST x-ray database and I don't see a uranium line near C Ka (between 40 and 50 angstroms) , though I do see a 2nd order O Ka line.  I believe you but in what table are you seeing the U line near carbon Ka?  Is this the modified xray.mdb that you and Philipp developed for PFE?
« Last Edit: April 12, 2019, 12:47:47 PM by Probeman »
« Reply #2 on: April 13, 2019, 02:18:31 AM »
I’m actually analysing C in U right at the moment and talked about this very interference problem at the AMAS symposium a couple of months ago. You can’t completely separate the C and U lines but you can reduce the degree of interference by using the 2nd order C Ka line on either an LDE1/PC1 or a Pb-stearate. This puts the line at the top end of the spectrometer range, where the resolution is better. The stearate has the highest resolution of the light element crystals, and almost completely separates the two peaks but the intensity is really lousy (<200th that of the LDE2/PC2). Both still need overlap correction (really easy to set up in PfE, thanks John!). I’m using the LDE1/PC1 as a compromise between level of correction and useable count rate (~20th that of LDE2/PC2). Thanks to Ben Buse for suggesting this method. As John says, the k-ratio values that PfE outputs include the overlap correction.
I have a sort of linked question though: The interfering U line is the N6-O4 at 0.286keV, with a relative intensity of 0.01% (Bearden, 1967), but there should also be a N6-O5 line at 0.294keV which Bearden lists with a relative intensity of 1% (i.e. 100 times bigger than the N6-O4 line), but I see no evidence of this peak. My working theory is that the O5 shell in U isn’t normally occupied so the line can’t be fluoresced and the 1% intensity in Bearden is just a default value when the actual intensity hasn’t been measured. However, my understanding of how shells are occupied is very limited and I might be completely misunderstanding the diagrams. Can anyone confirm or deny this hypothesis?
Mike
« Reply #3 on: April 14, 2019, 12:03:46 AM »
OK, after discussions with Xavier, it looks like I am mistaken about the shell occupancies: For U the 5f shell is partially filled and this corresponds to the O6 and O7 levels (I was wrongly equating the 5f with the O4 level) so the N6-O5 transition should be possible. However, unlike Bearden, the Penelope database lists the intensity of this transition as 1/10th that of the N6-O4 which may explain why I haven't seen it. That does make more sense since why would Bearden list it if it wasn't a possible transition. I guess the 1% relative intensity is just a default value if the actual value isn't known. Since I have some DU in the probe at the moment I'll try a slow scan to see if I can tease this peak out of the background.
Mike
« Reply #4 on: April 14, 2019, 08:15:09 AM »
Xavier Llovet responded to both Mike and I and said we could post his comments. Here they are:
Hi John, Mike:
I've checked the electronic configuration of U and is 5f3 6d1 7s2 so in principle the O4 and O5 levels (5d) are filled. Those that are partially filled are the 5f levels, which would correspond to O6 and O7 levels. So if I'm correct, both two N6-O5 and N6-O4 lines should exist.
The problem could be in the relative intensity reported by Bearden. For instance, according to the PENELOPE database, the N6-O4 line (with energy 289 eV) is 10 times more intense that the N6-O5 line (with energy 298 eV), just the opposite than Bearden's data! You can find this information in the U.mat file:
15 20  0  7.40479E-05  2.89740E+02
15 21  0  4.72978E-06  2.98070E+02
(the transition "15 20 0" would correspond to the N6-O4 line and the transition "15 21 0" to the N6-O5 line)
So perhaps this could explain what you're seeing?
Regards,
Xavier
The (modified) NIST x-ray database included with Probe for EPMA does include some N emission lines, but not all apparently. I guess we need a KLMN x-ray database!
Pd MZ2               43.3622  .285930  1.00000        ES
Pd MZ1               43.3622  .285930  10.0000        ES
Ne KB1      III      43.3814  .857410  .500000        JD
Se SLB1``   V        43.4488  1.42680  .250000        JD
Dy M2-N1    V        43.4610  1.42640  1.25000        JD
Ni SLA3     III      43.4691  .855680  .500000        JD
La MB       III      43.5016  .855040  45.0000        JD
Bi N5-N6             43.5801  .284500  1.00000        ES
Eu M4-O2    IV       43.5993  1.13750  .064000        JD
Eu MA1      IV       43.6146  1.13710  64.0000        JD
Ag MG       II       43.6422  .568190  16.0000        JD
Eu MA2      IV       43.6492  1.13620  64.0000        JD
Tm MZ1      IV       43.6607  1.13590  3.84000        JD
Er M4-O2    V        43.6630  1.41980  .025000        JD
Tl N4-N6             43.6707  .283910  1.00000        ES
Tm MZ2      IV       43.6761  1.13550  .640000        JD
Se LB1      V        43.6876  1.41900  9.71200        JD
Ni LA2      III      43.7081  .851000  5.72500        JD
Ni LA1      III      43.7081  .851000  50.0000        JD
Mn Ln       II       43.7338  .567000  5.14300        JD
Ga L2-N3    IV       43.7724  1.13300  .173000        JD
Ce M2-N4    IV       43.8266  1.13160  5.12000        JD
Ga SLB1``   IV       43.8421  1.13120  .640000        JD
Ne KA1      III      43.8467  .848310  50.0000        JD
Ne KA2      III      43.8467  .848310  25.0000        JD
Ga LG5      IV       43.8886  1.13000  .141000        JD
Tm M3-N1    V        43.9260  1.41130  .250000        JD
Pr MG       IV       43.9938  1.12730  38.4000        JD
C  KA1               44.0023  .281770  100.000        ES        <-- carbon emission
C  KA2               44.0023  .281770  50.0000        ES        <-- carbon emission
Er MA1      V        44.0039  1.40880  25.0000        JD
Er MA2      V        44.0039  1.40880  25.0000        JD
Cs M2-N1    III      44.0049  .845260  7.00000        JD
Cs MZ1      II       44.0063  .563490  8.00000        JD
Ga LB1      IV       44.0837  1.12500  10.6910        JD
La M3-N1    III      44.1759  .841990  .650000        JD
Pd M2-N4    II       44.2016  .561000  .800000        JD
Gd MG       V        44.2205  1.40190  6.52500        JD
Zr M3-N1             44.2789  .280010  1.00000        ES
As Ll       IV       44.2805  1.12000  3.15500        JD
Sm MZ2      III      44.4259  .837250  .500000        JD
Ge LG3      V        44.4392  1.39500  .028000        JD
Mn Ll       II       44.5991  .556000  10.6000        JD
Se SLA4     V        44.6119  1.38960  .250000        JD
La MA1      III      44.6414  .833210  50.0000        JD
La MA2      III      44.6414  .833210  50.0000        JD
Eu M3-N1    IV       44.6432  1.11090  .640000        JD
Se SLA5     V        44.6569  1.38820  .250000        JD
As LB4      V        44.6633  1.38800  .650000        JD
As LB3      V        44.6633  1.38800  1.19200        JD
Ho SMB2     V        44.6859  1.38730  .250000        JD
Cu Ln       III      44.7068  .831990  1.37500        JD
Kr Ll       V        44.7601  1.38500  1.12700        JD
Ga SLA4     IV       44.7601  1.10800  .640000        JD
Sm MZ1      III      44.7865  .830510  3.00000        JD
Se SLA3     V        44.7924  1.38400  .250000        JD
Zn LB4      IV       44.8005  1.10700  1.51000        JD
Zn LB3      IV       44.8005  1.10700  .250000        JD
Ru M4-O2             44.8021  .276740  .010000        ES
Ga SLA5     IV       44.8086  1.10680  .640000        JD
W  MZ1      V        44.8118  1.38340  .336000        JD
Ho MB       V        44.8280  1.38290  14.8610        JD
La M4-O2    III      44.8313  .829680  .050000        JD
Sm M2-N4    V        44.8832  1.38120  .300000        JD
Se LA2      V        44.9548  1.37900  2.85500        JD
Se LA1      V        44.9548  1.37900  25.0000        JD
W  MZ2      V        44.9679  1.37860  1.12800        JD
Ga SLA3     IV       44.9956  1.10220  .640000        JD
Pb N5-N6             45.0021  .275510  1.00000        ES
Sm MB       IV       45.0815  1.10010  56.3200        JD
I  MG       III      45.1064  .824620  10.0000        JD
Ga LA2      IV       45.1677  1.09800  7.30900        JD
Ga LA1      IV       45.1677  1.09800  64.0000        JD
Sb M2-M4             45.2023  .274290  .010000        ES
Hg N4-N6             45.2023  .274290  1.00000        ES
Na SKB^4    IV       45.2584  1.09580  .640000        JD
Tb M2-N1    V        45.3495  1.36700  1.25000        JD
Te M2-N4    III      45.3511  .820170  .500000        JD
Nd M2-N1    IV       45.3993  1.09240  2.56000        JD
Er MZ1      IV       45.4867  1.09030  3.84000        JD
Er MZ2      IV       45.5118  1.08970  .640000        JD
Ho M4-O2    V        45.5728  1.36030  .025000        JD
Er M3-N1    V        45.5996  1.35950  .250000        JD
Sm MA1      IV       45.7173  1.08480  64.0000        JD
Sm MA2      IV       45.7173  1.08480  64.0000        JD
Sm M4-O2    IV       45.7216  1.08470  .064000        JD
Cd M2-N1    II       45.8018  .541400  4.00000        JD
Ho MA2      V        45.8085  1.35330  25.0000        JD
Ho MA1      V        45.8085  1.35330  25.0000        JD
Cu Ll       III      45.8639  .811000  2.07000        JD
In M3-N1    II       45.9503  .539650  .800000        JD
By the way, I also searched Philipp Poeml's modified replacement NIST x-ray database, which has additional emission lines for the actinide elements, and it doesn't have these N emission lines for U either.
« Last Edit: April 14, 2019, 08:43:11 AM by Probeman »
« Reply #5 on: April 14, 2019, 01:52:35 PM »
It did take some digging to work out what the overlapping peak was, none of the software tools showed any U lines in that region of the spectrum. We had to resort, as the late Douglas Adams put it, to twig-technology and dig out the hard copy x-ray tables.
wrigke
« Reply #6 on: April 15, 2019, 03:06:18 PM »
Thank you all for your suggestions.  I will let you know what I find out when I try these suggestions.
I do have a question about the overlapping line though.  My Bearden 1967 clearly describes the U line at 286 ev as "N4O4", not N6O4.  Is that a misprint in the article?
Karen
« Reply #7 on: April 24, 2019, 04:02:30 AM »
Hi Karen,
Sorry for the delay in replying. I got my info from the NIST x-ray database which does quote Bearden 1967 as one of its sources, but this gives the line as N6-O4. Looks like there's been a transcription error somewhere. Does anyone have the big Cameca book of x-ray tables to hand to see what that one says is the U-line at 43.4 A, 0.286 keV?
Mike
« Reply #8 on: April 24, 2019, 10:04:11 AM »
Hi Karen,
Mike
I only have the BRGM "little book" which only goes down to the U M lines, but I do have a White and Johnson (2nd edition, May 1970) book which reports:
U N6-O4    0.286 KeV    43.29999 angstroms
OK, I also found my Bearden 1964 book which lists:
U N6-O4    0.286 KeV     43.6 angstroms  (literally NVI OIV, but if my Roman numerals are correct that's N6-O4)
« Last Edit: April 24, 2019, 10:11:11 AM by Probeman »
« Reply #9 on: May 06, 2020, 01:06:35 PM »
Has anyone got a reference or a weblink for the NIST x-ray database. I want to cite it but I can’t remember where I picked my copy up from and a search isn’t turning it up.
jrminter
« Reply #11 on: May 08, 2020, 08:05:17 AM »
Unfortunately not. I did see that one but it’s only got K and L lines. The one I’m after was compiled by Chuck Fiori. Failing that, is there another citable comprehensive database? I’ve got Bearden but I’m hoping for something more recent and with more lines at the top end of the periodic table.
« Reply #12 on: May 08, 2020, 09:38:13 AM »
The table of emission lines that were first utilized by Probe Software for plotting KLM markers and other purposes were obtained from Dale Newbury and/or Nicholas Ritchie (I think!), back in 2000 or so (originally created by Chuck Fiori).
There are two versions.  The first version was a text file called MASTER.LIN. It contains 4985 (1st order) emission lines, along with absorption edges. See attachments below.  The file MASTER.TXT has a description of the data. Here is what it says:
This data base contains 4985 entries and includes all the  measurable
X-ray  lines,  satellites  and  absorption edges from under 100 eV to
over 120 keV. Additionally, most of the X-ray  lines  and  satellites
are  assigned  a  relative intensity (relative to the alpha-1 line in
each family).  The  data  base  was  assembled  primarily  from  four
sources:
1.)  B.L.  Doyle, W.F. Chambers, T.M. Christensen, J.M. Hall and G.H.
Pepper "SINE THETA SETTINGS FOR X-RAY SPECTROMETERS", Atomic Data and
Nucleur Data Tables Vol. 24, No 5, 1979.
2.)  E.W. White, G.V. Gibbs, G.G. Johnson Jr. and G.R. Zechman "X-RAY
WAVELENGTHS AND CRYSTAL INTERCHANGE SETTINGS  FOR  WAVELENGTH  GEARED
CURVED CRYSTAL SPECTROMETERS" Report of the Pennsylvania State Univ.,
1964.
3.) J.A. Bearden "X-RAY WAVELENGTHS AND X-RAY ATOMIC  ENERGY  LEVELS"
Rev. Mod. Phys., Vol. 39, No. 78, 1967.
4.)  J.A  Bearden  and A.F. Burr,"REEVALUATION OF X-RAY ATOMIC ENERGY
LEVELS", Rev. Mod. Phys., Vol. 31, No. 1, 1967.
Each X-ray line or edge series as a function of atomic number was fit
to  a  fourth  degree  polynomial.   The  fit was subtracted from the
appropriate data and the residuals plotted and examined. In this  way
rogue  entries  could be identified and corrected. The resulting data
base is considered to be sufficiently accurate  for  any  application
involving  the  Si(Li)  X-ray  detector and single crystal wavelength
spectrometers.
The  data  base  is  comprised  of  three  data  files:   MASTER.LIN,
MASTER.TRS,  and MASTER.ENG.   These three files are identical except
that they have been sorted in different ways.  MASTER.LIN  is  sorted
such  that  all  entries belonging to a particular transition such as
KA2 are grouped together in ascending atomic number.   MASTER.TRS  is
sorted  in such a manner that all the lines and edges associated with
a particular element  are  grouped  together.  These  groups  are  in
ascending  atomic  number.   And  finally,  MASTER.ENG  is  sorted in
ascending wavelength in Angstrom units.
Each data file is organized in the following manner: The first column
contains the atomic number and the second contains the atomic symbol.
The third column is the transistion or  edge.    Where  possible  the
transition  is  given  in  Siegbahn notation.      If the entry is an
absorption edge the forth column contains the letters ABS,  otherwise
the  column  is  blank.   the fifth column contains the wavelength in
Angstrom units.  The sixth column contains  the  relative  transition
probability  expressed  as  a percentage of the principal line within
each family ie K, L or M alpha 1. Absorption edge entries contain the
value  zero  for  this  column  with  the  anticipation  of  a future
inclusion of jump ratios. Finally, the last column gives a  code  for
the  source  of  the  entry.    If  the column is blank the source is
reference 2.  If the column contains the letter  "C"  the  source  is
reference  1.  If the letters "BB" appear, the source is reference 4.
The letters "W,F" mean that reference 2 was  used  but  the  relative
transition  probability  has been adjusted by Fiori.  Reference 3 was
used as a check since it is the source of  many  of  the  entries  of
reference 1.
In column 3 the notation KA1,2 means the entry is the weighted sum of
the KA1 and KA2 in the ratio 2 to 1.    For  low  atomic  number  the
entries  are  not  self  consistent  since the data is from different
sources.  If the column begins with the capital  letter  S  then  the
entry  is a satelite line due to doubly ionized atoms.   The relative
transition values for these  entries  are  only  valid  for  electron
excited specimens, and are, at best, estimates.
For more information call Chuck Fiori (301-496 2599) or write to   me
at Rm 3W-13 Bldg. 13, National  Institutes  of  Health, Bethesda, Md.
20205. In  any event,  if  you  find  this data useful please keep in
contact since we plan to write our work up formally and will no doubt
update and improve what is in the present files.
Th following  are  Siegbahn  to shell-transition notation conversions:
You will have to use your imagination to discover which arabic letters
we used to correspond to the Siegbahn Greek notation:
KA   =KA1+KA2+KA3
KA1,2=(2*KA1+KA2)/3
KA1  =K-L3
KA2  =K-L2
KA3  =K-L1
KB   =SUM(KBn)
KBX  =Metal
KB1  =K-M3
KB1' =KB1+KB3+KB5
KB2  =(K-N3)+(K-N2)
KB2' =K-N3
KB2''=K-N2
KB3  =K-M2
KB4  =(K-N4)+(K-N5)
KB5  =(K-M4)+(K-M5)
KB5' =K-M5
KB5''=K-M4
Kd1  =K-O3
Kd2  =K-O2
LA   =LA1+LA2
LA1  =L3-M5
LA2  =L3-M4
LB1  =L2-M4
LB10 =L1-M4
LB15 =L3-N4
LB17 =L2-M3
LB2  =L3-N5
LB3  =L1-M3
LB4  =L1-M2
LB5  =(L3-O4)+(L3-O5)
LB6  =L3-N1
LB7  =L3-O1
LB9  =L1-M5
LG1  =L2-N4
LG11 =L1-N5
LG2  =L1-N2
LG3  =L1-N3
LG4  =L1-O3
LG4' =L1-O2
LG6  =L2-O4
LG8  =L2-O1
Ll   =L3-M1
Ln   =L2-M1
Ls   =L3-M3
Lt   =L3-M2
Lu   =(L3-N6)+(L3-N7)
Lv   =L2-N6
MA1  =M5-N7
MA2  =M5-N6
MB   =M4-N6
MG   =M3-N5
MG2  =M3-N4
MZ1  =M5-N3
MZ2  =M4-N2
Md   =M2-N4
Me   =M3-O5
Eventually we discovered that a number of Mz and Mg lines were missing from Chuck's original table and so Nicholas Ritchie created a new table which is also attached below.  This table has 5713 emission lines, but no absorption edges.
But I don't have any documentation of this newer table, though Nicholas can help I am sure.
Finally, I attach below the XRAY.ELM file which is a text file I generated back in 2006, with all these lines and edges, and also higher order reflections with "nominal" intensities.
Hope this helps.
« Last Edit: May 08, 2020, 10:57:16 AM by John Donovan »
« Reply #13 on: May 08, 2020, 01:58:00 PM »