Author Topic: Flow Proportional Counter Backflow Gas Regulation  (Read 7644 times)
DavidAdams
« on: July 19, 2018, 09:10:27 am »
Here at the USGS Denver Microbeam Lab we are in the process of having our new JEOL 8530F+ Microprobe installed (YAY!). Everything has been going smoothly up until a couple of days ago when the engineer was checking the performance of the WDS detectors. Without going into details, the GFP counters aren't really working the way they need to be.  Because of the altitude in Denver the atmospheric pressure is significantly different than at sea level and that has a very noticeable affect on X-ray detection using the GFP counters versus the sealed Xenon detectors. It screws with the gain and bias voltages and P/B ratios. Changes in barometric pressure here has a very noticeable effect on the PHAs on our system much more so than I have seen on other instruments at lower altitudes. This is something that we've noticed for years on our old 8900 and we've recognized it and just dealt with it, but with the new probe being installed I wanted to get people's opinions and thoughts on ways to eliminate or minimize the effect of altitude and barometric pressure on the GFP counters.
My questions are: has anyone with a JEOL probe attempted some form of back pressure regulation at the end of the P-10 gas line from the GFP counters? Is this even possible or feasible? Is there some why to safely do this without blowing the detector windows that are under vacuum?
Any input/ideas/help would be greatly appreciated!
Dave
David Adams
The University of Auckland
Owen Neill
« Reply #1 on: July 19, 2018, 09:37:26 am »
Hi Dave,
If you can hook the tube to a butyl phthalate bubbler, that will help a lot. The glass piece (usually two hollow, parralel advil-capsule-looking glass pieces with hose barbs on either end, about 2x2") usually comes with the probe, although the butyl phthalate won't. You can get 100ml on Sigma Aldrich for <$20, and that's more than you'll ever need.
If the don't give you the glass case anymore, you can make one with a couple of beakers, just make sure to have a second empty vessel in case the window pops so you don't backstream it.
Edit: Found a picture (photocredit, U. Utrecht): http://nanosims.geo.uu.nl/nanosims-wiki/lib/exe/fetch.php/eprobe:microprobe_april_2014.jpg
. The bubbler I'm talking about is on the spectro on the left side (SP1 position), halfway between the black cap and the detector pre-amp. Connect the P-10 outlet hose from the last spectro to the barb on the left-hand glass tube, and fill the right-hand tube with butyl phthalate.
Good luck,
OKN
« Last Edit: July 19, 2018, 09:55:20 am by Owen Neill »
DavidAdams
« Reply #2 on: July 19, 2018, 10:10:50 am »
Hi Owen,
Thanks for the reply!
The dibuthyl phthalate in the bubbler is already something that I do on the the old 8900 probe and I also gave some to Ken for the new 8530 probe bubbler. Unfortunately, that does very little to curb the back pressure problems that I experience up here. I also experienced the atmospheric pressure effects when I was running the 8530F in Perth but the effect was less dramatic then what I'm seeing here in Denver.
Thanks again!
Dave
David Adams
« Reply #3 on: July 19, 2018, 11:47:55 am »
Hi Dave,
In Denver you are at roughly 85% sea level pressure, so I would expect that you can increase the back pressure by 15% and the detector windows should handle it fine.
DavidAdams
« Reply #4 on: July 19, 2018, 12:02:30 pm »
Thanks! That's definitely something I will try when JEOL hands over the 8530 to us. Hopefully that will be a simply easy fix if it works!!
Dave
David Adams
Owen Neill
« Reply #5 on: July 19, 2018, 01:05:39 pm »
Hi Dave,
OK, cool, and also interesting. The atmospheric effects I saw on the 8500 at WSU largely went away after I filled the bubbler, but Pullman is only ~2300ft ASL, so maybe not comparable. Maybe something simple like a de-humidifier for the lab?
Say hi to Ken for me,
OKN
« Reply #6 on: August 15, 2018, 12:27:39 am »
We got a bubbler from JEOL for our 8530F Plus as well.
I'm wondering why dibutyl phthalate is the fluid of choice (same for the bubblers on our Cameca SX100)? At least in Australia it is a class 9 dangerous substance, and the specified vapour pressure at room temperature is not particularly low either.
Wouldn't it be possible to use something like diffusion pump oil or even rotary pump oil? Many are non-hazardous, much lower vapour pressures. Santovac 5 is terribly expensive of course, but for a single small JEOL bubbler one wouldn't need much.
With the "clean vacuum system" of the newer generation machines, oil-free pumps, on-airlock plasma cleaners, cryo anticontamination etc, what would be the best bubbler fluid? Normally we probably wouldn't expect much back streaming, but if a counter window leaks or breaks...
Karsten
D.
« Reply #7 on: August 15, 2018, 05:31:54 am »
Sorry to hijack, but I've always wondered why the FE instruments come with bubblers and the 8230's don't. ? . I assume there is no technical reason?
D.
DavidAdams
« Reply #8 on: August 17, 2018, 05:28:03 am »
Karsten,
This is a good question and I return to my original post. If one put a back pressure regulator at the end of the gas flow chain between the last spectrometer and the bubbler I would guess that that would help solve any atmospheric pressure influences as well as help prevent any back streaming of any fluid (or anything else) in the bubbler in the event of a window failure.
D.
The FE 8530F in Perth, Western Australia didn't come with a bubbler either. I had to scrounge one up. Thankfully there was an old one in a drawer from a previous instrument. I just think JEOL randomly decides whether or not to send it with an instrument.
Dave
David Adams
« Reply #9 on: August 17, 2018, 05:40:04 am »
Thanks Dave, we probably got the bubbler because we asked for it. The information given to us was that it is optional, so only supplied on request.
« Reply #10 on: August 17, 2018, 09:13:34 am »
I would think just some cheap silicone based diffusion pump oil would be fine to use as a bubbler fluid, but I have no special knowledge in this area.  Except that at room temperature, silicon oil is not going to be out gassing anything!
« Last Edit: August 17, 2018, 09:15:17 am by Probeman »
« Reply #11 on: August 17, 2018, 02:09:26 pm »
Cameca instruments have a bubbler on the exit of each spectrometer. Do JEOL instruments only have a single bubbler?
The only advantage I can think of for separate bubblers is that it's easy to see which spectrometer has a broken detector window. Not that it happens very often, especially (never) since we added these "soft start" valves to our roughing pump line:
http://www.vatvalve.com/business/valves/catalog/H/311_1_A
https://probesoftware.com/smf/index.php?topic=120.msg487#msg487
« Reply #12 on: August 17, 2018, 05:36:51 pm »
Yes, by default only one bubbler on the JEOL as normally all the P10 channels are in series. I assume one reason for having separate regulators and bubblers on the Cameca would be that they use different pressures ("low" and "high" pressure counters) for different x-ray energy ranges. On the JEOL all P10 counters are at the same pressure as sealed Xe is used for high x-ray energies.
But I've seen at least one JEOL where the P10 line is split upstream from the spectros to have an individual supply for each P10 counter (and I think also a separate bubbler).
I was curious and did some testing on our 8530F+ where the three P10 spectrometers are currently in series. I moved the last spectro in the line on a major element peak, and then did the same for the first and second spectro while monitoring the count rate on the last spectro in the line. There is no noticeable change even at very high count rates on the first two spectrometers. So "recycling" of the gas further down the line does not appear to be affected by what's happening on spectrometers upstream.
I've also heard that one may see differences in count rates when changing the order of the spectrometer in the P10 gas chain on a JEOL instrument, i.e. the spectrometers further downstream can have lower count rates. But I'd assume that may be due to P10 leakage somewhere in the chain, e.g. a leaky detector window, or a tube not attached properly...?
Dave,
I don't have any experience with these back-pressure regulators. I can see how it might help in your case at high altitude, but do you think it would  also make a difference for us, being pretty much at sea level? I.e. in addition to preventing potential backstreaming, could they even out fluctuations in atmospheric pressure? There seem to be a range of designs available. How much of a pressure difference to atmospheric pressure would one have to set for them to operate properly? Obviously too much pressure on the detector windows etc would be a concern.
Karsten
DavidAdams
« Reply #13 on: August 20, 2018, 09:07:53 am »
Karsten,
It would be interesting to see if a back-pressure regulator would be useful for JEOL probes at lower altitudes. I think we're going to try one here to see if it will actually smooth out the pressure effects we see. Have you monitored your PHA bias voltages on your P10 counters over time particularly when there is a big change in the atmospheric pressure? In Denver the voltages on our flow counters can often be ≥±10V during large changes in the weather, which, as you can imaging, has a noticeable affect on the analyses being performed. I don't remember what I was seeing when I was in Perth, but I don't remember it being quite so dramatic. The bias voltage on the Xenon counters is almost always the same. The greatest variation on the xenon counters that I've seen here is ±2V and that could be easily attributed to non-pressure related changes. Usually, however, the bias voltage is exactly the same for weeks and months at a time. A back-pressure regulator should, in theory, stop any atmospheric pressure related fluctuations. There are some very low pressure regulator options available (0-10 psi) and from what I understand there doesn't need to be a large pressure differential in those regulators to operate properly. I'm sure that the JEOL engineers will know what the maximum pressure before window failure would be.
Dave
David Adams
« Reply #14 on: August 20, 2018, 10:28:16 am »
Hi Dave,
In Kingston, at ~100 m AMSL (and no P-10 back-pressure regulator), I see variations in bias voltage required to keep pulse amplitude distributions centered at 4 V using the gas-flow counters (but not with the sealed Xe counters, at least over short periods of time).  These swings can be of similar magnitude to the ones that you report and appear to correlate with variation in atmospheric pressure -- but I need to keep a better record of this in my log file to be absolutely certain.  Even if centering the PHA distribution at 4 V for a given count rate doesn't solve the problem of variation in deadtime with X-ray energy, it sure does provide a useful reference for monitoring fluctuations in the distribution due to various causes.
Brian
« Last Edit: August 20, 2018, 10:45:47 am by Brian Joy »