Re: Time Dependent Intensity (TDI) Corrections
« Reply #60 on: April 07, 2021, 12:00:00 PM »
Hi Andrew,
Interesting stuff.  I'm interested in this mineral, what is it?

But since the TDI fit is applied on a per point basis, I think looking at the statistics for each single point is a better way to evaluate this.
john
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Andrew Mott

  • Post Doc
  • ***
  • Posts: 11
Re: Time Dependent Intensity (TDI) Corrections
« Reply #61 on: April 07, 2021, 12:41:11 PM »
John,

It's alunite/natroalunite - (K,Na)Al3(SO4)2(OH)6

So you would recommend evaluating the models on a point by point basis rather than trying to be consistent with the model used?

Thanks again.

Andrew

John Donovan

  • Administrator
  • Emeritus
  • *****
  • Posts: 2691
  • Other duties as assigned...
    • Probe Software
Re: Time Dependent Intensity (TDI) Corrections
« Reply #62 on: April 07, 2021, 12:51:55 PM »
So you would recommend evaluating the models on a point by point basis rather than trying to be consistent with the model used?

Hi Andrew,
I think the model you use to evaluate the fit should be evaluated on a point by point basis since that is how the TDI correction is applied during the matrix iteration.

In your original examples it looked to me (based on the Relative Deviation % displayed) that the log-linear fit gave the smallest avg deviation.
john
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

AndrewLocock

  • Professor
  • ****
  • Posts: 65
    • University of Alberta Electron Microprobe Laboratory
Re: Time Dependent Intensity (TDI) Corrections
« Reply #63 on: April 07, 2021, 01:25:57 PM »
This may have been pointed out earlier, but:

The TDI intensities can be saved in Excel format, and then evaluated in Excel.
An example of Na intensities for a point in obsidian (15 kV, 6 nA, 5 micron beam diameter, nominally 30 s on peak) is in the PDF attachment.

Cheers,
Andrew Locock