Author Topic: Micro-XRF detector for EPMA  (Read 6930 times)

D.

  • Professor
  • ****
  • Posts: 45
Micro-XRF detector for EPMA
« on: March 19, 2018, 04:44:04 AM »
Hi All,

1) Has anyone tried to add a micro-XRF (SEM-XRF) system to an EPMA? Was it successful?

2) Is there an obvious reason not to try this?

3) What sort of primary-X-ray penetration depths are we talking about for exciting K and L lines for common rock-forming minerals.

Thanks,
D.

Probeman

  • Emeritus
  • *****
  • Posts: 2858
  • Never sleeps...
    • John Donovan
Re: Micro-XRF detector for EPMA
« Reply #1 on: March 19, 2018, 11:46:41 AM »
3) What sort of primary-X-ray penetration depths are we talking about for exciting K and L lines for common rock-forming minerals.

I don't know about questions 1 and 2, but you can get a rough idea of the x-ray penetration range by entering a composition in the Calculate Electron and X-Ray Ranges window in CalcZAF as described here:

http://probesoftware.com/smf/index.php?topic=86.msg309#msg309

I don't know what energy x-ray beam you can generate with these devices, but I suppose it depends on the target?  In any case, you'll want to enter a thickness (say 5 or 10 or 15 um) and the energy of the x-ray beam you are generating to see what sort of attenuation you get.  It's just Beers Law, but one can enter an arbitrary composition which is nice.
The only stupid question is the one not asked!

Anette von der Handt

  • Global Moderator
  • Professor
  • *****
  • Posts: 355
    • UMN Probelab
Re: Micro-XRF detector for EPMA
« Reply #2 on: March 19, 2018, 06:53:30 PM »
Hi,

you might want to check with Matthew Loocke (I think he is here on the forum, otherwise at the Shimadzu Institute for Research Technology in Arlington). He used to be in Cardiff where they have/he ran an analytical SEM that also included a micro-XRF system. He can probably give a lot of useful insight into it.

I looked into this at one point (and suggested the system later to Cardiff) and one side effect from my memory was that you have to give up one spectrometer. Also, the systems with smallest spot sizes (I think around 10 micron was the smallest) got real expensive very fast. But this was a few years ago and this may have all well changed.
Against the dark, a tall white fountain played.

JonF

  • Professor
  • ****
  • Posts: 155
Re: Micro-XRF detector for EPMA
« Reply #3 on: March 22, 2018, 05:31:48 AM »
We've got one of the Bruker microXRF sources bolted on to our Quanta 650 and I've had similar ideas: WDS spectral resolution with the microXRF spatial resolution would be pretty interesting. The increased sensitivity by removing the bremsstrahlung would be ace.

A few thoughts I had:
  • Mounted on a WDS port would mean that you would be looking at a tangental cross section of the uXRF beam (meaning an oval shape on your sample surface) rather than circular, and a Gaussian distribution of the photons across that.
  • The Xray beam (coming in at the 'take off' angle to the sample surface rather than perpendicular to it) would travel not only through the grain you were looking at, but potentially also in to any neighbouring grains and possibly even in to the sample holder itself. The distance the beam travels would be dependent on the material that you were looking at and the energy of the primary Xray beam (ie whatever target you buy)
  • ZAF/PRZ would obviously be out of the window, but XRF quantification isn't exactly new so there must be an XRF-PRZ alternative(?)
  • Figuring out the geometric relationship between the point at which a characteristic X-ray was generated (remembering that the primary Xray beam would be travelling across as well as through the sample!) and the take off angle (and so extent of absorption by the sample) with respect to the position of the WDS crystal/detector would be... interesting!

An uXRF source mounted on top of an old EPMA where the column would be? That would be something interesting...


 


mloocke

  • Student
  • *
  • Posts: 3
  • Always thinking...
    • LSU Shared Instrumentation Facility
Re: Micro-XRF detector for EPMA
« Reply #4 on: March 30, 2018, 01:49:51 PM »
Most µXRF companies won’t have experience with putting them on an EPMA. My experience was with iXRF systems based out of Austin, Texas- the owner (Kenny Witherspoon, email me for his contact info) was very forthcoming with me regarding the lack of industry knowledge for operating a µXRF on an EPMA/SEM system and he was very eager to collaborate and try and develop new applications for their products.

I loved having the uXRF on the SEM in Cardiff. We had dual 150mm2 EDS detectors which were enabling us to resolve <10ppm concentrations for most elements Z>13 when using the µXRF. The only problem that we ran into was that there wasn’t much information in the literature regarding any experimental constraints on the actual penetration depths and interaction volumes. That being said, I found that these values could be modeled in the same way that we model electron-specimen interactions with relatively accurate results. I ran a series of tests before I left Cardiff on Ca in Olivine and Ti in quartz and the results were darn near spot on with various models that I had ran in CalcZAF.

The µXRF couples well with the EDS detectors on any instrument, but you need to make sure that you have room in your chamber for the µXRF to be adjusted for focusing the beam. It isn’t really all that well known how well the µXRF will couple with WDS detectors on an EPMA. We had a WDS on our SEM in Cardiff and my primary plan was to explore the possibilities of using the WDS and µXRF together with the E-beam for mixed analyses. I didn’t manage to get too far into that before I got the job offer to come to UT Arlington.

In regards to Jon's points:
"Mounted on a WDS port would mean that you would be looking at a tangental cross section of the uXRF beam (meaning an oval shape on your sample surface) rather than circular, and a Gaussian distribution of the photons across that." - This is true. We had our's mounted at a 32.5° takeoff angle which resulted in an oval-shaped spot roughly 10µm by 15 to 20 µm in size. We simply mapped out the location of the beam and overlayed the position and shape of the spot on our observation window.

"The Xray beam (coming in at the 'take off' angle to the sample surface rather than perpendicular to it) would travel not only through the grain you were looking at, but potentially also in to any neighbouring grains and possibly even in to the sample holder itself. The distance the beam travels would be dependent on the material that you were looking at and the energy of the primary Xray beam (ie whatever target you buy)" - Although this is true, this was not really a problem that we faced until we ran into a grain that was oriented perfectly to diffract our X-ray beam and cause a significant amount of secondary fluorescence. Occasionally we would get a little bit of an Al Ka signal fro the sample chamber, but most of the time we didn't see anything coming off of the sample holder or from surrounding grains aside from the expected secondary fluorescence.

"ZAF/PRZ would obviously be out of the window, but XRF quantification isn't exactly new so there must be an XRF-PRZ alternative(?)" - In the world of XRF they use a simple fundamental parameters (FP) method for quantification. It is relatively simple and is actually pretty good at standardless analysis. It is very easy to set up cal curves in conjunction with the FP method.

"Figuring out the geometric relationship between the point at which a characteristic X-ray was generated (remembering that the primary Xray beam would be travelling across as well as through the sample!) and the take off angle (and so extent of absorption by the sample) with respect to the position of the WDS crystal/detector would be... interesting!" - This was something that I was trying to work on before I left Cardiff, but I didn't get to make much headway on it.

Hopefully this helps. It has been a little while since I've really thought about any of this, but I can definitely dig up some of my data if anybody has any other questions.

Matthew Loocke, Research Professor
LSU Geology & Geophysics
E235 Howe-Russel-Kniffen Geoscience Complex
Baton Rouge, LA 70803
Mobile: (+1) 713 530 3064
Email: mloock1@lsu.edu

D.

  • Professor
  • ****
  • Posts: 45
Re: Micro-XRF detector for EPMA
« Reply #5 on: April 04, 2018, 01:17:58 AM »
Hi All,

After several exchanges with iXRF the last couple weeks, this was the conclusion:

"after further evaluation by our engineers, there's apparently not enough space to put an X-Beam in your JEOL Microprobe.  The "lens" of a microprobe does not look at the sample in the same way as a SEM lens does.  Most ports (maybe all) look through a hole in the "lens" onto the sample.  Unfortunately, there's no way to get close to the sample and gain a proper excitation."

Anette, thank you for suggesting Matthew. He shared his experience with me and then posted to the forum. (It sounds to me like SEM-XRF still needs a lot of development from the geo-application side to reach its full potential. )

Thank you all for sharing your knowledge, and bringing up the interesting points to consider.

(Always nice to dream about getting around the Bremsstrahlung!)

Deon.

Probeman

  • Emeritus
  • *****
  • Posts: 2858
  • Never sleeps...
    • John Donovan
Re: Micro-XRF detector for EPMA
« Reply #6 on: April 04, 2018, 09:03:38 AM »
I wonder if the Cameca instrument with its left side vacuum port would be a better candidate for a micro-XRF device?

Somewhere I have a cut away view of this SX100 side port, but I remember this port looks right over the top of the sample from the side.  I know Gatan uses this port to mount their parabolic CL collector.
The only stupid question is the one not asked!

jon_wade

  • Professor
  • ****
  • Posts: 82
Re: Micro-XRF detector for EPMA
« Reply #7 on: April 04, 2018, 03:02:46 PM »
Hi john - no in the Cameca the coaxial optics gets in the way. If it were to use this port, it would have to be retractable.

Besides which, I agree with all John F's comments in the micro XRF - yes, its fine for mapping, but quant and its a different question. 

I'm sure you'll be pleased to hear I  did once simulate a micro XRF set up in a beam-line using Penelope  to track secondary fluorescence. Its not exactly insubstantial!
« Last Edit: April 04, 2018, 05:26:04 PM by John Donovan »