Author Topic: Calibrating Carbon Coated Standards to Samples with a Gold coat  (Read 4072 times)

Samuel Saltzman

  • Student
  • *
  • Posts: 3
Calibrating Carbon Coated Standards to Samples with a Gold coat
« on: September 25, 2015, 12:40:00 PM »
Hi,

Anyone have any good literature on applying carbon coated standards to gold coated unknowns?  Particularly how to quantify the change in x-ray absorption.

Sam

Probeman

  • Emeritus
  • *****
  • Posts: 2856
  • Never sleeps...
    • John Donovan
Re: Calibrating Carbon Coated Standards to Samples with a Gold coat
« Reply #1 on: September 25, 2015, 02:54:34 PM »
Hi,

Anyone have any good literature on applying carbon coated standards to gold coated unknowns?  Particularly how to quantify the change in x-ray absorption.

Sam

Hi Sam,
Both Probe for EPMA and also CalcZAF (our free EPMA utility) can correct for differences in coating between the standard and the unknown for quantification. See here:

http://probesoftware.com/smf/index.php?topic=265.0

The user reference manual that comes with CalcZAF (and PFE) documents what is being done for this "differential" coating correction, but bottom line is: it works!

You can download CalcZAF here for free:

http://probesoftware.com/Technical.html
The only stupid question is the one not asked!