Hi Phil,
I agree- but "failure modes" can still be instructive!
![Smiley :)](https://probesoftware.com/smf/Smileys/default/smiley.gif)
I've attached a text file that contains the results of running all my Penepma calculations through my binary alpha factor database looking for "xtreme" absorption or fluorescence cases. It is true that the fluorescence model is analytical using Salvat and Llovet's Fanal FORTRAN, where both the beam incident and boundary materials are the same... See:
http://probesoftware.com/smf/index.php?topic=47.0but the absorption and atomic number effects are modeled using the full Penepma Monte-Carlo code (Penfluor) at 11 different beam energies. Then I filtered those for between 5 and 30 keV and ZAF corrections greater than 8 (800% absorption) or less then 0.6 (40% fluorescence) and those are attached.
A few selections from the text file show that there are some interesting cases that could be empirically measured for comparison as seen here:
Warning: very large matrix correction (ZAF= 6.12903, K= 1.63158, C=10) for Mg ka in Zn at 26 keV : 12-30_40.txt...
Warning: very large matrix correction (ZAF= 6.42811, K= .777833, C=5) for Mg ka in Zn at 26 keV : 12-30_40.txt...
Warning: very large matrix correction (ZAF= 6.63289, K= .150764, C=1) for Mg ka in Zn at 26 keV : 12-30_40.txt...
Warning: very large matrix correction (ZAF= 6.44948, K= 1.55051, C=10) for Mg ka in Zn at 27 keV : 12-30_40.txt...
Warning: very large matrix correction (ZAF= 6.15973, K= .162345, C=1) for Al ka in Ge at 30 keV : 13-32_40.txt...
Warning: very large matrix correction (ZAF= .580743, K= 1.72193, C=1) for Si ka in Pb at 5 keV : 14-82_40.txt...
Warning: very large matrix correction (ZAF= .598947, K= 1.66960, C=1) for Si ka in U at 5 keV : 14-92_40.txt...
Warning: very large matrix correction (ZAF= .550399, K= 1.81687, C=1) for Ti kb in Fe at 7 keV : 22-26_40.txt...
Warning: very large matrix correction (ZAF= .392112, K= 2.55029, C=1) for Cr ka in Mn at 7 keV : 24-25_40.txt...
Warning: very large matrix correction (ZAF= .377312, K= 2.65033, C=1) for Cr kb in Mn at 7 keV : 24-25_40.txt...
Warning: very large matrix correction (ZAF= .489611, K= 2.04244, C=1) for Cr ka in Mn at 8 keV : 24-25_40.txt...
Warning: very large matrix correction (ZAF= .475308, K= 2.10390, C=1) for Cr kb in Mn at 8 keV : 24-25_40.txt...
Warning: very large matrix correction (ZAF= .567872, K= 1.76096, C=1) for Cr ka in Mn at 9 keV : 24-25_40.txt...
Warning: very large matrix correction (ZAF= .555012, K= 1.80176, C=1) for Cr kb in Mn at 9 keV : 24-25_40.txt...
Warning: very large matrix correction (ZAF= .369880, K= 2.70358, C=1) for Cr ka in Fe at 7 keV : 24-26_40.txt...
Warning: very large matrix correction (ZAF= .576173, K= 8.67795, C=5) for As ka in Pb at 13 keV : 33-82_40.txt...
Warning: very large matrix correction (ZAF= .561766, K= 1.78010, C=1) for As ka in Pb at 13 keV : 33-82_40.txt...
Lots more where these came from!
So, if you have a binary compound of any of these element pairs, but don't have one of these specific (1:99, 5:95, 90:10, 8-:20, 60:40, 50:50, etc.) stoichiometries, say 50:50 NaF (alas, because although I have already calculated a 50:50 binary, it's 50:50 by weight!), just let me know and I'll calculate the Penepma k-ratio for your actual specimen, and we'll have an empirical comparison!