Author Topic: Setting The Beam Current  (Read 6115 times)

John Donovan

  • Administrator
  • Emeritus
  • *****
  • Posts: 3304
  • Other duties as assigned...
    • Probe Software
Setting The Beam Current
« on: November 13, 2013, 10:07:28 AM »
As most of you may know, setting the beam current on a JEOL instrument is slightly more involved since there is no ROM level function as there is in the Cameca instruments.

Instead, applications have to adjust the beam current at the software level by adjusting the coarse and fine condenser lens values. In Probe for EPMA there are two modes for this beam current setting that are found in the Probewin.ini file in the PFE application folder in the [hardware] section as seen here:

JeolCondenserCoarseCalibrationMode=1        ; 0 = acquire calibration curve, 1 = use pre-defined calibration curve

Usually we want to use JeolCondenserCoarseCalibrationMode = 1 to save time rather than acquiring a new calibration curve for each time the beam current is set, but this does require that one routinely re-runs the beam current calibration every few months or so.

If you see the following message (or something similar) in your PFE log window:

8/10/2013 9:07:51 PM, Program: Probewin, Error: Calculated CC 4 is too low for beam current of 25. Setting to default low value and continuing coarse iteration, Procedure: JEOLSetBeamCurrentMDA

it probably means that the aperture is in the wrong position, the column alignment has changed or the aperture has gotten dirty. In any case, simply make sure the correct aperture position is specified in the Analytical Conditions dialog and click the Calibrate Set Beam Current button as seen here:



The calibration will run and be automatically saved and this calibration should be good for at least a few months.
« Last Edit: November 13, 2013, 06:42:22 PM by John Donovan »
John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Dan R

  • Professor
  • ****
  • Posts: 82
    • GE Research Materials Characterization
Re: Setting The Beam Current
« Reply #1 on: November 13, 2013, 12:47:21 PM »
I've also had to recalibrate occassionally when changing the accelarating voltage.
-Dan

Probeman

  • Emeritus
  • *****
  • Posts: 2856
  • Never sleeps...
    • John Donovan
Re: Setting The Beam Current
« Reply #2 on: November 13, 2013, 01:41:20 PM »
Hi Dan,
In my hearsay experience of JEOL instruments, that usually indicates that there is a aperture or column misalignment issue. The aperture is easy to adjust by us, but a column misalignment should be a service call.
The only stupid question is the one not asked!

John Donovan

  • Administrator
  • Emeritus
  • *****
  • Posts: 3304
  • Other duties as assigned...
    • Probe Software
Re: Setting The Beam Current
« Reply #3 on: March 19, 2014, 11:59:32 AM »
Here is a another typical error message which indicates that you might want to run a beam current calibration using the Calibrate Set Beam Current button under Analytical Conditions.



This calibration can be performed for each aperture in your instrument, though most use only one or two apertures for quant work.

John J. Donovan, Pres. 
(541) 343-3400

"Not Absolutely Certain, Yet Reliable"

Probeman

  • Emeritus
  • *****
  • Posts: 2856
  • Never sleeps...
    • John Donovan
Re: Setting The Beam Current
« Reply #4 on: March 20, 2014, 09:43:45 AM »
I wouldn't rule it out.

The best way to test this, is to run a number of replicate beam current measurements. If the FaradayWaitInTime isn't long enough, you'll see some bad current readings even just measuring the beam current.  Try using the Faraday.exe app and make sure that it inserts and removes the faraday cup on each measurement- it should if you check the "Cycle" checkbox.

If the aperture is dirty or the beam current calibration is out, you will more likely see the number of iterations exceeded in setting the beam current.
The only stupid question is the one not asked!