Could you share with me some tricks which will help to measure integrated intensity (and APF as well) for peaks contain an extra one.
For instance, ka oxygen peak includes small Si peak on SiO2. So how I can ignore this small peak because background level is lower then "hight off-peak" of this small peak.
We use LDE, JEOL.
See addition.
Thank you
Rom,
I think you are confusing two different things.
When using the integrated intensity feature for quantitative analysis, the peak shift and/or shape changes are automatically compensated for, because you are obtaining the integrated intensity of each peak. So there is no need to also utilize the area peak factor (APF) corrections. But you'll still need to deal with background and interference corrections.
If instead you want to use the APF corrections, simply utilize the normal on-peak/off-peak acquisitions and then select whether to utilize the compound APFs included with PFE, or (more work) measure the specified APFs for your specific unknown relative to your specific standard. The differences between compound APFs and specified APFs are discussed here:
https://probesoftware.com/smf/index.php?topic=536.0However in both cases you will need to correct for, or avoid, spectral interferences. In the case of using integrated intensities you will need to identify the interfering peak, and then also acquire integrated intensities on an appropriate interference standard, and then assign the interference assignment as usual.
In the case of using normal on-peak/off-peak measurements and measuring a specified APF for your unknown, you would have to deconvolve the interfering peak(s) from both your unknown and standard integrated intensity scans to obtain an interference corrected specified APF, and then you would have to apply an interference correction as usual to the actual unknowns (and standard).
However, it should be noted that because the integrated intensity scans are a much wider spectral area (than simply measuring the on-peak intensity), you are much more likely to have spectral interferences with integrated intensity scans. Note also that you must avoid off-peak spectral interferences in both cases, but that is relatively easy to do because one can select different background measurement positions.
With all this in mind, I suggest you are probably better off using normal on-peak/off-peak measurements and the compound APFs that are already included in PFE. Also be sure to use a suitable background correction, e.g, an exponential or MPB background method to obtain properly corrected net intensities.
And of course after all that, one also has to consider the accuracy of the mass absorption coefficients (MAC) for low energy emission line measurements. But that is another topic!
In the meantime here are a number of "white papers" from many years ago which discusses some of these APF issues:
https://epmalab.uoregon.edu/reports/APF%20measurements.dochttps://epmalab.uoregon.edu/reports/APF%20Normalization.pdfhttps://epmalab.uoregon.edu/reports/Withers%20hydrous%20glass.pdfhttps://epmalab.uoregon.edu/reports/Preliminary%20work%20on%20MgB2%20and%20MgB4.pdfOf course the easy way to handle these peak shape/shift issues is to use a matrix matched standard for your unknown, but that is not always possible.