Occasionally one may see motor time out errors reported by Probe for EPMA from the Sx100 spectrometer or stage motors. For example:
"Motor move on motor 2 (target= 76412, actual= 76413) timed out after 100 seconds..."
This error is caused by the instrument not completing the commanded motion within the specified timeout time and is caused by three different hardware issues:
1. The spectrometer glass optical encoder is dirty, causing the spectrometer to "skip" certain positions.
2. The spectrometer lead screw is dirty, causing the spectrometer to "settle" on one side or the other of the offending speck.
3. The stainless steel rod that slides in the brass detector body is dirty, causing the spectrometer to "stick" in certain positions.
All three issues can be easily fixed by careful cleaning of the surfaces by your service engineer.
Note that because the Peak-Sight software does *not* check for the motion complete flag, it will *not* report this error for spectrometer or stage moves that fail to complete.