Hi Dave,
I don't have data for a LDE3L and don't know anyone who has this monochromator. However, if the relative sensitivity changes between LDE1 and LDE5 in their respective L vs H form apply to the LDE3 as well, I would expect to see peak counts of 420-440 cps/nA and a p/b of 62-74 for Be Ka at 10kV on the LDE3L. For comparison, spec for LDEB I see peak counts of 70 cps/na and p/b of 60.
Nevertheless, I would expect that JEOL should be able to give you these numbers for LDE3L, no? I am also a little bit surprised to hear that the LDE1L gives you worse performance than the LDE1, based on specs it should give you 1.7 - 2.8 times better counts (but worse p/b).
Cheers,
Anette