Software => CalcZAF and Standard => Topic started by: Probeman on May 24, 2015, 06:27:28 PM
-
As we all know, precision is quite easy to estimate (just make some replicate measurements!), but ascertaining accuracy is more difficult and usually requires the analysis of so called "secondary" standards. These secondary standards can be any composition with a known (or zero) concentration of the element(s) of interest, but ideally they are similar in matrix (concentrations) to the unknown composition, and not assigned as a primary standard to the element(s) in question.
Evaluating accuracy for such secondary standards in Probe for EPMA is quite straight forward, since Probe for EPMA can "analyze" standards as though they were unknowns. By this I mean PFE automatically adds in elements not analyzed for (but present in the standard composition) to all standard sample acquisitions, so that the matrix correction can be calculated properly and then compared to the "published" values for that (secondary) standard as seen here:
St 305 Set 1 Labradorite (Lake Co.), Results in Elemental Weight Percents
ELEM: Na Si K Al Mg Fe Ca Mn Ti O H
TYPE: ANAL ANAL ANAL ANAL ANAL ANAL ANAL ANAL ANAL SPEC SPEC
BGDS: LIN LIN LIN MAN LIN MAN LIN LIN LIN
TIME: 30.00 30.00 20.00 30.00 30.00 30.00 30.00 20.00 30.00 --- ---
BEAM: 30.17 30.17 30.17 30.17 30.17 30.17 30.17 30.17 30.17 --- ---
ELEM: Na Si K Al Mg Fe Ca Mn Ti O H SUM
13 2.847 24.582 .112 16.343 .083 .347 9.520 .001 .026 46.823 .000 100.683
14 2.833 23.624 .101 16.355 .086 .314 9.544 -.003 .037 46.823 .000 99.714
15 2.842 23.794 .108 16.371 .078 .311 9.444 .007 -.001 46.823 .000 99.778
AVER: 2.841 24.000 .107 16.356 .082 .324 9.503 .002 .021 46.823 .000 100.058
SDEV: .007 .511 .006 .014 .004 .020 .052 .005 .019 .000 .000 .542
SERR: .004 .295 .003 .008 .002 .011 .030 .003 .011 .000 .000
%RSD: .25 2.13 5.22 .09 5.15 6.08 .55 306.38 92.79 .00 .00
PUBL: 2.841 23.957 .100 16.359 .084 .319 9.577 .000 n.a. 46.823 .000 100.060
%VAR: (-.01) .18 6.95 (-.02) -2.28 1.60 -.78 .00 --- .00 .00
DIFF: (.00) .043 .007 (.00) -.002 .005 -.074 .000 --- .000 .000
STDS: 305 358 374 305 12 395 358 25 22 --- ---
In the above example this labradorite standard (#305) is the primary (assigned) standard for Na and Al and is a secondary standard for the remaining elements. The line labled AVER: is the acquired analysis for the standard, the line labeled PUBL: is the expected analysis from the standard composition database and the line labeled %VAR is the relative percent error from the published concentration value.
Immediately it is evident that Si measured in labradorite (std #305), relative to diopside (std #358), is accurate to around several tenths of a percent, and Ca in labradorite relative to diopside again is accurate to better than a percent relative. The remaining minor and trace elements can also be examined to ascertain the ability to measure low (or zero) concentrations with his setup.
In CalcZAF, every sample is considered an unknown, so determining accuracy is slightly more work, but that simply means that one must import the analysis of a secondary standard and compare the results to the expected concentrations.
We might call this method of determining accuracy by using secondary standards an "external" method since we are checking ourselves using an external (secondary) standard. But there is another way to estimate accuracy, though secondary standards should always be acquired if at all possible!
I'll call this other accuracy assessment method an "internal" method, since we will be relying not on secondary standards, but on multiple matrix corrections. This is reason that John Armstrong called his off-line analysis program TRYZAF. Using it, one can "try" many different matrix corrections on the same dataset. This also true for CalcZAF.
Of course we might hope that these multiple matrix correction methods will always give the same answer, but of course they don't (that's science for you!). In any case, by "internal" I simply mean we can analyze the unknown composition using a number of matrix corrections methods that are available in TryZAF, CalcZAF (and PFE of course) as seen here:
(http://probesoftware.com/smf/oldpics/i59.tinypic.com/1zxb76w.jpg)
For example here is the previous labradorite (#305) composition exported from PFE and analyzed using CalcZAF with the Use All Matrix Corrections checkbox checked:
(http://probesoftware.com/smf/oldpics/i60.tinypic.com/2chu8ms.jpg)
After clicking the Calculate button, we wait for the app to cycle through all 10 matrix corrections (of course all these matrix correction methods are utilizing the same mass absorption coefficient (MAC) table, so that is another parameter that can be tested. CalcZAF (and PFE) can utilize 5 different MAC tables.
Summary of All Calculated (averaged) Matrix Corrections:
Labradorite (Lake Co.)
LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV
Elemental Weight Percents:
ELEM: Na Si K Al Mg Fe Ca Mn Ti O H TOTAL
1 2.841 24.002 .107 16.356 .082 .324 9.502 .002 .021 46.823 .000 100.060 Armstrong/Love Scott (default)
2 2.840 24.109 .107 16.353 .081 .331 9.519 .002 .021 46.823 .000 100.186 Conventional Philibert/Duncumb-Reed
3 2.840 24.036 .107 16.354 .082 .319 9.498 .002 .021 46.823 .000 100.082 Heinrich/Duncumb-Reed
4 2.840 24.057 .107 16.354 .082 .324 9.505 .002 .021 46.823 .000 100.115 Love-Scott I
5 2.841 24.006 .107 16.356 .082 .324 9.504 .002 .021 46.823 .000 100.065 Love-Scott II
6 2.840 24.089 .107 16.353 .080 .329 9.532 .002 .021 46.823 .000 100.176 Packwood Phi(pz) (EPQ-91)
7 2.838 24.316 .107 16.340 .082 .322 9.620 .002 .020 46.823 .000 100.470 Bastin (original) Phi(pz)
8 2.842 23.871 .107 16.362 .080 .329 9.510 .002 .021 46.823 .000 99.947 Bastin PROZA Phi(pz) (EPQ-91)
9 2.839 24.147 .107 16.350 .082 .330 9.525 .002 .021 46.823 .000 100.226 Pouchou and Pichoir-Full (Original)
10 2.840 24.116 .107 16.352 .082 .330 9.529 .002 .021 46.823 .000 100.201 Pouchou and Pichoir-Simplified (XPP)
AVER: 2.840 24.075 .107 16.353 .082 .326 9.524 .002 .021 46.823 .000 100.153
SDEV: .001 .115 .000 .005 .001 .004 .036 .000 .000 .000 .000 .139
SERR: .000 .036 .000 .002 .000 .001 .011 .000 .000 .000 .000
MIN: 2.838 23.871 .107 16.340 .080 .319 9.498 .002 .020 46.823 .000 99.947
MAX: 2.842 24.316 .107 16.362 .082 .331 9.620 .002 .021 46.823 .000 100.470
For this particular example, the variance of the 10 matrix correction methods is quite small, so if we are having accuracy issues it is not due to the matrix correction method selected!
However, other compositions will yield quite different results for the different matrix correction methods as seen here:
Summary of All Calculated (averaged) Matrix Corrections:
St 20 Set 1 ThSiO4 (Huttonite)
LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV
Elemental Weight Percents:
ELEM: Si Zr Th O TOTAL
1 9.191 -.069 70.784 19.746 99.652 Armstrong/Love Scott (default)
2 7.713 -.062 71.150 19.746 98.546 Conventional Philibert/Duncumb-Reed
3 8.634 -.065 71.467 19.746 99.782 Heinrich/Duncumb-Reed
4 8.334 -.064 71.395 19.746 99.410 Love-Scott I
5 8.370 -.064 71.400 19.746 99.452 Love-Scott II
6 7.253 -.059 71.034 19.746 97.973 Packwood Phi(pz) (EPQ-91)
7 9.212 -.069 71.556 19.746 100.444 Bastin (original) Phi(pz)
8 8.779 -.066 71.501 19.746 99.958 Bastin PROZA Phi(pz) (EPQ-91)
9 8.365 -.065 71.396 19.746 99.442 Pouchou and Pichoir-Full (Original)
10 8.170 -.063 71.344 19.746 99.196 Pouchou and Pichoir-Simplified (XPP)
AVER: 8.402 -.065 71.303 19.746 99.385
SDEV: .608 .003 .241 .000 .701
SERR: .192 .001 .076 .000
MIN: 7.253 -.069 70.784 19.746 97.973
MAX: 9.212 -.059 71.556 19.746 100.444
Now someone might ask: OK, which matrix correction method is correct? And the answer of course is "why all of them"! At least that is what their authors will claim!
But in fact we know (or should know!), that some matrix corrections perform better for large absorption corrections, some matrix corrections work well for large atomic number differences, etc, etc etc... but none are "universal" yet!
But if I wanted to hedge my bets, I would certainly utilize the Penepma alpha factor correction method. It's as good as our our current understanding of the electron solid physics involved. To run your analyses using Penepma, simply select the polynomial or non-linear alpha fit method and then check the Penepma boxes as seen here:
(http://probesoftware.com/smf/oldpics/i58.tinypic.com/2j6ptk.jpg)
The results from the Penepma alpha factor method is seen here:
ELEMENT si ka zr la th ma O Total
UNK KRAT .0684 -.0005 .6500
UNK WT% 9.088 -.065 71.585 19.746 100.354
UNK AT% 17.346 -.038 16.538 66.155 100.000
UNK BETA 1.3291 1.2013 1.1014
ALPITER 5.0000
This compares well to the ideal composition for ThSiO4:
ELEM: Si Zr Th SUM
PUBL: 8.665 n.a. 71.589 100.000
-
Dear Dr. Donovan,
I have successfully used the free-ware CALZAF to model the impact of distinct matrix correction programs and MAC tables on specific EPMA results produced in my lab. Now my question:
How and on which basis are the ZAF-factors for the standards are calculated? I am particularly interested in standards that contain "volatile" elements such as N (in BN). It would be helpful for me to get some description or references dealing with this topic.
Best regards and thanks in advance,
Thomas
-
Dear Dr. Donovan,
I have successfully used the free-ware CALZAF to model the impact of distinct matrix correction programs and MAC tables on specific EPMA results produced in my lab. Now my question:
How and on which basis are the ZAF-factors for the standards are calculated? I am particularly interested in standards that contain "volatile" elements such as N (in BN). It would be helpful for me to get some description or references dealing with this topic.
Best regards and thanks in advance,
Thomas
Hi Thomas,
Please call me John. Great questions! I'll start from the beginning so forgive me if you know any of this already.
The 10 matrix correction methods (and 5 or 6 mass absorption coefficient (MAC) reference tables) in CalcZAF (including several historical ZAF methods, plus a number of modern pr(z) methods) are utilized differently for standards and unknowns. For calculating unknowns we provide the measured (or modeled) background corrected intensities (and takeoff and keV) and CalcZAF calculates the resulting compositions based on the matrix physics in an iterative manner, usually converging rapidly on a final composition (because we don't know the physics until we know the final composition).
But for standard ZAF factors (also known as the standard k-factors, see the PFE Reference manual and click on the Glossary for ZAF for definitions of terms), we instead provide the standard compositions, and CalcZAF calculates the matrix physics using the same equations but essentially running them backwards (non iteratively because we already know the composition and therefore the physics).
So let's take an example of boron nitride, by clicking the Enter Composition as Formula String button and type in "BN" like this:
(https://probesoftware.com/smf/gallery/1_12_05_20_11_59_58.png)
then click the Calculate button and we get this output for the default Armstrong/Brown pr(z) method:
BN, sample 1
WARNING in ZAFSetZAF- the f(x) of B ka is .3887
Current Mass Absorption Coefficients From:
LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV
Z-LINE X-RAY Z-ABSOR MAC
N ka N 1.8125e+03
N ka B 1.5803e+04
B ka N 1.1196e+04
B ka B 3.3340e+03
ELEMENT ABSFAC ZEDFAC FINFAC STP-POW BKS-COR F(x)e
N ka 1.6872 3.8909 6.5646 .2469 .9608 .5927
B ka 2.5725 4.0830 10.5032 .2357 .9625 .3887
SAMPLE: 32767, TOA: 40, ITERATIONS: 0, Z-BAR: 6.128777
ELEMENT ABSCOR FLUCOR ZEDCOR ZAFCOR STP-POW BKS-COR F(x)u Ec Eo/Ec MACs
N ka 3.1206 1.0000 .9793 3.0560 .9847 .9945 .1899 .4000 37.5000 7906.92
B ka 2.0623 .9992 1.0297 2.1220 1.0203 1.0093 .1885 .1880 79.7872 7771.35
ELEMENT K-RAW K-VALUE ELEMWT% OXIDWT% ATOMIC% FORMULA KILOVOL
N ka .00000 .18468 56.439 ----- 50.000 .500 15.00
B ka .00000 .20528 43.561 ----- 50.000 .500 15.00
TOTAL: 100.000 ----- 100.000 1.000
The column marked "K-RAW" is of course all zeros because we provided no measured intensities, just the compositions. However, the column marked "K-VALUE" is the elemental k-ratios for those elements in that composition for those conditions, while the ZAFCOR values are the matrix corrections utilized in the standard k-factor. Again see the PFE Reference manual glossary for the equations that put these all together. Look for std-k-factor, k-ratio and ZAF definitions.
Also I don't quite see how nitrogen in BN is a "volatile" element, this material is quite refractory usually. But as for evaluating the different matrix corrections, you should definitely utilize the Use All Matrix Corrections checkbox as described in the first post in this topic. Also note that emissions such as boron and nitrogen are *very* dependent on the MACs. You should investigate the use of empirical MACs (see the Empirical MACs in the CalcZAF Analyutical menu), and also read about the use of empirical area peak factors (APFs) which handle the effect of chemical states on these low energy emissions. The PFE Reference manual contains this information.
A good topic on analyzing these light elements is also found here:
https://probesoftware.com/smf/index.php?topic=536.0
Please let me know if you have further questions.
-
Hi John,
thanks a lot for your prompt reply and very helpful comments! I already had a look on the glossary and I will continue to perform the calculation you suggested.
Of course, I did not mean that N in BN is volatile :). I thought that the std-k factors are based on pure elements (and nitrogen is a gas). My confusion is caused by the observation that I got unrealistic high Z-factors of up to 28 for N in the BN standard in my lab while these Z factors are generally about 1 in the calculations made by using CALCZAF.
Best regards,
Thomas
-
Ah, OK. But remember, in EPMA, so long as the distance is such that the electrons come to rest (e.g., an "infinitely" thick specimen), the density of the unknown (or standard) does not matter. The physics all works out just fine (the only place where density matters in EPMA is with "thin" samples, where a geometric correction must be applied along with the matrix correction, because some electrons are transmitted through the thin sample and therefore do not contribute to the measured sample signal).
When you say "unrealistic high Z-factors of up to 28 for N in the BN standard in my lab" are you referring to the std k-factors or the ZAF matrix factors? As you see from the previous output:
ELEMENT ABSCOR FLUCOR ZEDCOR ZAFCOR STP-POW BKS-COR F(x)u Ec Eo/Ec MACs
N ka 3.1206 1.0000 .9793 3.0560 .9847 .9945 .1899 .4000 37.5000 7906.92
B ka 2.0623 .9992 1.0297 2.1220 1.0203 1.0093 .1885 .1880 79.7872 7771.35
the ZAF matrix factors for B and N Ka in BN are 2.1 and 3 respectively. This is because in BN each contains significant concentrations of themselves and the ZAF factor for pure elements is always 1.0.
Again, although we are calling these "ZAF" factors, these factors are calculated using the Armstrong/Brown pr(z) method. Here is the same calculation using the PAP pr(z) method:
BN, sample 1
WARNING in ZAFSetZAF- the f(x) of B ka is .3556
Current Mass Absorption Coefficients From:
LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV
Z-LINE X-RAY Z-ABSOR MAC
N ka N 1.8125e+03
N ka B 1.5803e+04
B ka N 1.1196e+04
B ka B 3.3340e+03
ELEMENT ABSFAC ZEDFAC FINFAC STP-POW BKS-COR F(x)e
N ka 1.7572 .0407 .0715 23.5957 .9599 .5691
B ka 2.8123 .0660 .1857 14.7226 .9724 .3556
SAMPLE: 32767, TOA: 40, ITERATIONS: 0, Z-BAR: 6.128777
ELEMENT ABSCOR FLUCOR ZEDCOR ZAFCOR STP-POW BKS-COR F(x)u Ec Eo/Ec MACs
N ka 3.5450 1.0000 .9958 3.5301 1.0022 .9936 .1605 .4000 37.5000 7906.92
B ka 2.2589 .9992 1.0026 2.2631 .9950 1.0077 .1574 .1880 79.7872 7771.35
ELEMENT K-RAW K-VALUE ELEMWT% OXIDWT% ATOMIC% FORMULA KILOVOL
N ka .00000 .15988 56.439 ----- 50.000 .500 15.00
B ka .00000 .19248 43.561 ----- 50.000 .500 15.00
TOTAL: 100.000 ----- 100.000 1.000
So, as one can see, these factors do vary from one matrix correction method to another, especially for these low energy emissions (because the absorption correction terms are so large)!
Please feel free to share some of your CalcZAF calculations with us.
-
Hi John,
sorry for the delay but I was out of office, and thanks for your explanation.
I mean the N ZAFCOR values for the BN standard which are unusually high at my microprobe, and this is also resulting in very high ZAF-values. Of course, I already used all the correction procedures available in CALCZAF but I generally got values in the same order that you mentioned. So, what may be the reason for these "anomalies"?
Best regards,
Thomas
-
Hi Thomas,
No worries.
When you say "are unusually high at my microprobe" I assume you mean the JEOL or Cameca software is giving you much larger matrix correction factors for N Ka in BN (than those you obtain from CalcZAF)?
What MAC value for N Ka in B is being utilized? Can you provide us with an example of these "unusually high" matrix factors?
-
Hi John,
attached you will find one analysis of BN (as unknown). The MAC of N should be that of the LINEMU data base...
Best regards,
Thomas
-
Hi Thomas,
Someone more familiar with the JEOL quantitative output will have to comment on this example, because I have never used the JEOL software for quantification (here is what Thomas attached in the previous post, the "ZAF factors" in red I think are the "unusually high" values he is talking about):
Asynchronous Mode.
*******************************************************************************
UNK No. = 2 ./Labor/BN_test dated Sun Apr 26 01:09:20 2020
2 Elements WDS only X= 14.038 Y= 13.281 Z= 10.252
Acc. Voltage = 10.0 (kV) Probe Dia. = 0 Scan OFF
Probe Current = 3.010E-08 (A)
*******************************************************************************
Channel Element Acm Peak Peak cnt sec BG_L cnt sec BG_U cnt sec
1 LDEN N T 1 110.585 42377.1( 16.0) 1511.3( 8.0) 361.0( 8.0)
1 LDEB B T 1 128.873 42055.2( 16.0) 644.1( 8.0) 1889.5( 8.0)
Measurement over
Correction starts
Standard Data
Element Standard name Wt.(%) ZAF Fac. Z A F
1 N A_BN 56.4400 9.3141 28.0328 0.3323 1.0000
2 B A_BN 43.5600 26.3534 78.5310 0.3356 1.0000
Standard Intensity of WDS
Element Curr.(A) Net(cps) Bg-(cps) Bg+(cps) S.D.(%) Date
1 N 3.008E-08 2528.2 189.8 50.6 0.30 Apr 26 01:02 2020
2 B 3.008E-08 2457.8 76.0 243.9 0.31 Apr 26 01:02 2020
Unknown Specimen No. 2
Group : Labor Sample : BN_test
UNK No. : 2 Comment : A_BN_2
Stage : X= 14.0383 Y= 13.2815 Z= 10.2525
Acc. Voltage : 10.0 (kV) Probe Dia. : 0 Scan : Off
Dated on Apr 26 01:10 2020
WDS only No. of accumulation : 1
Curr.(A) : 3.010E-08
Element Peak(mm) Net(cps) Bg-(cps) Bg+(cps) S.D.(%) D.L.(ppm)
1 N 110.585 2531.5 188.9 45.1 0.52 853
2 B 128.873 2461.5 80.5 236.2 0.54 809
ZAF Metal
Element Wt.(%) Atom(%) K(%) K-raw(%) ZAF Z A F
N 56.479 49.9907 56.477 100.065 1.0000 1.0000 1.0000 1.0000
B 43.595 50.0093 43.596 100.082 1.0000 1.0000 1.0000 1.0000
-------------------------------------------------------------------------
Total 100.074 100.0000 100.073 200.147 Iteration = 4
UNK Measurement over
I do not know what the Z, A and F factors displayed in the above JEOL output represent, but multiplying the Z and A factors together I do get the "ZAF Fac." values in red. I can understand the "F" (fluorescence?) factors both being 1.000, but why would the "A" factors both be 0.33?
It appears to me that you are simply analyzing your BN standard as an unknown using your BN standard as the standard. If I do the same in CalcZAF (I created a "fake" BN standard for this example because my BN standard is not stoichiometric and contains some carbon and oxygen), I get this output calculating concentrations from counts:
STANDARD PARAMETERS (TOA= 40):
ELEMENT STDNUM STDCONC STDKFAC Z-BAR ABSCOR FLUCOR ZEDCOR ZAFCOR
N Ka 603 56.439 .2704 6.1288 2.1290 1.0000 .9805 2.0875
B Ka 603 43.561 .2560 6.1288 1.6604 .9993 1.0256 1.7018
ELEMENT STP-POW BKS-COR F(x)e F(x)s Eo Ec Eo/Ec
N Ka .9863 .9941 .7563 .3552 10.00 .4000 25.0000
B Ka 1.0179 1.0076 .5837 .3516 10.00 .1880 53.1915
SAMPLE: 0, TOA: 40, ITERATIONS: 10, Z-BAR: 6.12849
ELEMENT ABSCOR FLUCOR ZEDCOR ZAFCOR STP-POW BKS-COR F(x)u Ec Eo/Ec MACs uZAF/sZAF
N ka 2.1285 1.0000 .9805 2.0870 .9863 .9941 .3553 .4000 25.0000 7919.32 .9997528
B ka 1.6606 .9993 1.0257 1.7020 1.0179 1.0076 .3515 .1880 53.1915 7780.42 1.000138
ELEMENT K-RAW K-VALUE ELEMWT% OXIDWT% ATOMIC% FORMULA KILOVOL
N ka 1.00131 .27071 56.499 ----- 49.985 .500 10.00
B ka 1.00151 .25636 43.633 ----- 50.015 .500 10.00
TOTAL: 100.131 ----- 100.000 1.000
The JEOL output does not show the MAC values and I would not assume they are the "LINEMU" values. Can you output the JEOL MACs for B Ka in N and N Ka in B? Does anyone know where these values are found in the JEOL software?
-
Hi John,
yes, I know the MAC data bases for K alpha, L alpha and so on...in the JEOL software.
Here are the data of interest;
1) B as emitter: absorber B 3350, and N 11200, respectively;
2) N as emitter: absorber B 15800, and N 1810, respectively.
You are right, ZAFCor is calculated by multiplying Z, A and F. However, not only A values but particularly Z values seem to be very unusual to me...
Best regards,
Thomas
-
Hi John,
yes, I know the MAC data bases for K alpha, L alpha and so on...in the JEOL software.
Here are the data of interest;
1) B as emitter: absorber B 3350, and N 11200, respectively;
2) N as emitter: absorber B 15800, and N 1810, respectively.
These values are very close to the LINEMU vales as seen here:
Current Mass Absorption Coefficients From:
LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV
Z-LINE X-RAY Z-ABSOR MAC
N ka N 1.8125e+03
N ka B 1.5803e+04
B ka N 1.1196e+04
B ka B 3.3340e+03
However there is large variation in the literature as seen here in this output from CalcZAF using the Display MAC Emitter Absorber Pair menu, first B Ka in N:
MAC value for B Ka in N = 11196.24 (LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV)
MAC value for B Ka in N = 10570.00 (CITZMU Heinrich (1966) and Henke and Ebisu (1974))
MAC value for B Ka in N = .00 (MCMASTER McMaster (LLL, 1969) (modified by Rivers))
MAC value for B Ka in N = 10117.85 (MAC30 Heinrich (Fit to Goldstein tables, 1987))
MAC value for B Ka in N = 12596.53 (MACJTA Armstrong (FRAME equations, 1992))
MAC value for B Ka in N = 9566.52 (FFAST Chantler (NIST v 2.1, 2005))
MAC value for B Ka in N = 9566.52 (USERMAC User Defined MAC Table)
And here for N Ka in B:
MAC value for N Ka in B = 15802.96 (LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV)
MAC value for N Ka in B = 15810.00 (CITZMU Heinrich (1966) and Henke and Ebisu (1974))
MAC value for N Ka in B = .00 (MCMASTER McMaster (LLL, 1969) (modified by Rivers))
MAC value for N Ka in B = 16610.23 (MAC30 Heinrich (Fit to Goldstein tables, 1987))
MAC value for N Ka in B = 14326.82 (MACJTA Armstrong (FRAME equations, 1992))
MAC value for N Ka in B = 14705.91 (FFAST Chantler (NIST v 2.1, 2005))
MAC value for N Ka in B = 14705.91 (USERMAC User Defined MAC Table)
It is worth keeping in mind that when MACs are in the 10^4 range, a 1% change in the MAC corresponds roughly to a 1% change in the calculated composition.
You are right, ZAFCor is calculated by multiplying Z, A and F. However, not only A values but particularly Z values seem to be very unusual to me...
Exactly. Does anyone else know what these JEOL output values for Z and A below represent?
Standard Data
Element Standard name Wt.(%) ZAF Fac. Z A F
1 N A_BN 56.4400 9.3141 28.0328 0.3323 1.0000
2 B A_BN 43.5600 26.3534 78.5310 0.3356 1.0000
-
Exactly. Does anyone else know what these JEOL output values for Z and A below represent?
Standard Data
Element Standard name Wt.(%) ZAF Fac. Z A F
1 N A_BN 56.4400 9.3141 28.0328 0.3323 1.0000
2 B A_BN 43.5600 26.3534 78.5310 0.3356 1.0000
The Z factor is equal to the area under phi(rho*z), which is equal to (R/S)*(1/Q(E0)), where R is the backscatter factor, S is the stopping power, and Q(E0) is the ionization cross section. If you work through the calculation, you’ll see that (R/S)*(1/Q(E0)) for a given X-ray in the compound carries units of g/cm^2, so you shouldn’t expect it to look like a fraction (or at least a smaller number) until you take the ratio of it relative to the same quantity calculated for the pure element (since standard and unknown are the same in this case). The value given for A appears to be f(chi) for each respective X-ray in the mixture; this value makes sense for B Ka but it doesn’t look right for N Ka, which is very strongly absorbed by B. For X-rays of light elements, the value of f(chi) and the magnitude of the absorption correction are very heavily dependent on the accuracy of the model for phi(rho*z). Note that It’s critically important to carry units through calculations!!! But hardly anyone does, including JEOL.
-
Hi Brian,
Thank-you for chiming in, this is helpful information. I think it answers Thomas' questions quite nicely.
My take away would be, first the Z factor values are not in useful units, and second, the A factor appears to be in error, so we should perhaps just ignore these factors I suppose?
-
Hi Brian,
Thank-you for chiming in, this is helpful information. I think it answers Thomas' questions quite nicely.
My take away would be, first the Z factor values are not in useful units, and second, the A factor appears to be in error, so we should perhaps just ignore these factors I suppose?
The printed Z and A values in the JEOL output are for a given X-ray in BN, while CalcZAF and PfE print out ratios of these quantities to those for the pure elements. The JEOL software also prints out Z, A, and F relative to the standard. (It would be useful if CalcZAF and PfE did this, too.)
I haven't looked closely at calculations of f(chi) for B Ka and N Ka in BN, and so I'm not entirely certain that my assumption that f(chi) for N Ka in the mixture is smaller than that of B Ka is correct.
-
Here is some output for BN using PROZA96 with MACs determined by Bastin. So it looks like my assumption was incorrect (though, as expected, the absorption correction for N Ka in the mixture relative to the pure element is still much greater than for B Ka in the mixture relative to the pure element).
(https://probesoftware.com/smf/gallery/381_23_05_20_12_50_54.png)
-
Hi Brian and John,
thanks for the helpful comments!
Unfortunately, I was not able to find any detailed description of these correction procedures in the JEOL manual. But I will discuss this topic with the JEOL staff!
Regards, Thomas
-
Hi Brian and John,
I would like to announce some "open questions" after performing the calculations mentioned by Brian on May 22.
First, if the Z value of c. 28 for N in the BN standard would be equal to the area below the PRz curve (i.e. the "primary intensity") carrying units of g/cm2, then it cannot - in my opinion - represent a correction factor for the atomic number effect in the ZAF matrix correction. However, the "ZAF Fac" in the published file is calculated by ZxAxF as you can see.
I have further tried to calculate (R/S)x(1/Q(E10)) for N but was not successful to reproduce the value of about 28.
When I recalculated my BN data by using the CITZAF (i.e., PRz) method, I got values for the Z factor (given as (1/s-el)x(r-el)) of about 1 and not 28, i.e. quite different to the ZAF correction method..
May I ask you to explain this disreppancy?
Regards, Thomas
-
Hi Thomas,
I don't know exactly what the Z values printed out by the JEOL software represent. (I'm referring to the 28.0 for N Ka and 78.5 for B Ka in BN.) They are much too large to represent the actual atomic number corrections, and so I deduced that they are Z values for the X-ray of interest in the unknown before division by the respective value obtained from the standard using the same model. Without units attached to these quantities, they are difficult to interpret. Since only the ratios relative to the standard are used to calculate concentrations in the unknown, the units typically are not specified. This is bad practice, and it can create nightmares for those of us who want to figure out how the calculations are done. The A values presented are relatively easily identifiable as f(chi), which is dimensionless.
Also, I should point out that the area under phi(rho*z) wouldn't necessarily have to be stated in units of mg/cm^2, but it must carry units of mass per area (so it could be μg/km^2 or g/cm^2, for instance). I was a little sloppy in my original statement. Note that phi(rho*z) itself is dimensionless.
If you are interested in exploring how the Z and A factors are calculated realistically and precisely (and with identifiable units), then I suggest that you read through the presentation by Pouchou and Pichoir in "Electron Probe Quantitation" (aka "the green book") -- I've attached their paper; the basics are covered in pages 31-45. Note: it is not easy reading! Also, note that typos are present in equations 13 and 21. If you want, I can send you some Fortran code that I wrote based on this presentation.
-
Thomas,
Also note that the entire CalcZAF source code is available to all on Github as described here:
https://probesoftware.com/smf/index.php?topic=570.msg3245#msg3245
john
-
Hi Brian and John,
thanks for the comments and the links!
During the last weeks, I have tried to find out how to calculate in detail the Z and A correction based on Duncumb & Reed and Philibert by creating a simple excel sheet. I was able to reproduce the CALCZAF results but never the high JEOL numbers for the Z factors.....So, I have to go on to find out the final truth, hopefully ;)
And, Brian, it would be great to have your code for a test.
Regards, Thomas