I should also mention that to take advantage of the new detector "meta-data" in version 2.10 of Thermo Pathfinder Portal interface for integrated EDS and WDS in Probe for EPMA, you will also need to edit the "ThermoNSSVersionNumber" in the Probwin.ini file.
So after you download and install Thermo Pathfinder 2.10, you will need to open the Probewin.ini file in the C:\ProgramData\Probe Software\Probe for EPMA folder using any text editor (such as NotePad), and edit the following field in the [hardware] section:
ThermoNSSVersionNumber=6
Why is it 6? and not 2.10? Because we (Probe Software) have to keep supporting all the older Thermo NSS installations, and Thermo decided (all on their own) to re-set the version numbering back to 1.0 when they went to Pathfinder!
So to be clear here is the table for Thermo software version numbers:
ThermoNSSVersionNumber = 1 to 4 for Thermo NSS v. 1 to 4
ThermoNSSVersionNumber= 5 for Thermo Pathfinder v. 1
ThermoNSSVersionNumber = 6 for Thermo Pathfinder v. 2.10