Hi Emma,
I moved this to the Thermo board as maybe Steve Seddio can respond.
Error 5 is a Thermo NSS/Pathfinder socket timeout error. I've seen this error when I had some spectra with high deadtimes. It might also be from a Win 10 network timing issue.
But as far as net intensities go, just to speed things up I now utilize a recently added feature (Only Use EDS Stored EDS Net Intensities option in the Analysis Options dialog) by first putting PFE into off-line mode, then selecting all the samples from the Analyze! window, then clicking the Raw Data button. This extracts all the net intensities and these net intensities are now automatically saved to internal tables in the MDB file.
Then using this new Only Use EDS Stored EDS Net Intensities option, you can now only utilize these internally stored EDS net intensities for quantitative analysis, and everything runs *much* faster:
https://probesoftware.com/smf/index.php?topic=40.msg8637#msg8637This way one can take their probe MDB file and not have to connect to the Thermo software for net intensity calculations. Makes everything much more portable. This feature works for Thermo, Bruker and even JEOL EDS spectra.
Of course, when you add new elements (or new samples) for quantification by EDS, you'll need to turn this flag off and re-connect to Thermo to extract the new net intensities. Then you can turn on this "Only Use EDS Stored EDS Net Intensities" Option again.
So until you get this network timeout issue fixed, you can extract your net intensities on the other computer and bring it back over to the networked computer and then turn on the Only Use EDS Stored EDS Net Intensities option so it doesn't call the network interface.