I'll chime in here, as I get a lot of use out of our SXES (-LR rather than -ER though).
Regarding the WDS vs EDS port mounting, it depends on what chamber you have. The 8x30 series chambers (8230 and 8530F) and earlier have the SXES fitted to either WDS port 1 or 5 (i.e the two ports either side of the airlock), whereas I've been told the new iHP200F has a modified EDS port so that you can fit the SXES there instead (with a higher take off angle). You can see this arrangement on the front cover of the JEOL SXES brochure (from
https://www.jeol.com/download_catalogues.php#sem_fe_op).
If you're looking for a new machine, then you'll be asking for the SXES-EREP (or the SXES-LREP), where the EP = "
EDS
Port".
I've spent quite a lot of time getting my head around the best way of using the SXES-LR, as its quite a different beast compared to the WDS and EDS. The -LR (low range) comes with the JS50XL and the JS200N diffraction gratings, and these really are low energy (the energy of the C Ka is way too high for example, I have to use the 2nd order reflection or higher). The JS300N grating in the SXES-ER(/EP) looks to be pretty similar to the JS200N, so it'll be really for light element/soft X-ray use (and the all the complications that entails).
The JS2000 has a waaaay bigger energy range than the other three diffraction crystals, ranging from the smaller d spacing LDE crystals up through the TAP range and in to the bottom of the PET energy range. The CSIRO folks had a talk a while back where they compared the spectral resolution of the JS2000 grating across its range and found it was favourable relative to TAP for most of that range, but is notably worse than PET at the high energy end.
The big draw back of the SXES that I can think of are:
(i) time - and lots of it! - the SXES CCD is pretty slow. This is negated slightly by counting all elements and all backgrounds simultaneously, but you'll still be counting for 100s if not 1000s of msec per spectrum at the very least. This makes for some very long maps. Quantitative points would be possible, but you'll have some software to write!
(ii) it's bloomin' cold. The SXES CCD sits unprotected in your sample chamber at -70C, ensuring that any contamination you have in your system will sit on the CCD pretty rapidly - you're effectvely cryo-pumping your sample chamber. You'll either need to use the liquid nitrogen anticontamination device permanently (which is a nuisance when you have to fill up the dewar twice a day, and letting it run dry will crash the vacuum), or design your own anti-contamination device (I've built an
in vacuo Peltier based system).
Overall, I think the SXES is worth it. It's a big step forward in light element analysis, with much higher spectral resolution and essentially no deadtime to worry about. I don't think it's a
replacement for WDS and GFPC, but it's another useful detector alongside the WDS and EDS.