This is a very good question.
Ideally the EPMA lab should have temperature variations that are limited to +/- 0.5C but in the real world that is not always possible of course:
https://probesoftware.com/smf/index.php?topic=330.0If temperature control of the lab is not sufficient Probe for EPMA has a built in standard intensity drift correction which is designed to apply a linear interpolation between standardizations during an extended duration probe run:
https://probesoftware.com/smf/index.php?topic=168.0Therefore the standard intensity drift correction depends on the magnitude of the temperature drift in the lab, and is of course particularly important for PET crystals which have the largest thermal coefficient of expansion.
One could (in principle) include a standard analyzed as an unknown sample at repeated intervals in the automation list to track instrument drift, but there is no way to apply that drift to the standard (or unknown) intensities at analysis time. But I do know of users who have done this as a "accuracy check" for runs performed over many weeks or months.
So if the instrument drift in your lab cannot be adequately calibrated (linear interpolation) using the Run Standard Samples and Run Standard Samples (again) automation actions, what I would do is to utilize the Re-Standard Interval (hrs) feature shown here:
With this feature one can simply specify the re-standardization interval (in hours) and the software will automatically re-run the selected standards every X hours. Note that the software will only run the selected standards, so you don't have to select *all* the standards, perhaps just the standards assigned to the major elements. This is because trace elements aren't going to drift much (if at all), since background intensities tend to be very stable over time and in the case of off-peak backgrounds, they are measured for every analysis (unless using Nth point backgrounds in which case they are measured every N points).