I'm using a JXA-8200 Probe, with both JEOL and PfEPMA software. I'm really excited with Probe software due to the wide range of possibilities that this software allows me to perform. But I faced with some issues of direct oxygen measurements in oxide standards. Currently I'm using the LDE1 crystal (FCS WDS, 15kV, 20nA, probe diameter is 10-50 microns).
I have the problem of incorrect oxygen measurements in simple oxides (MgO, SnO2, etc). In particular, oxygen concentration is 2-3 wt.% higher in MgO, than standard concentration, while 15wt.% higher in SnO2.
I believe the lower oxygen concentration in SiO2 can be explained by the build-up of negative charge in a very nonconductive sample, but
how can we explain the increasing of oxygen concentration in simple oxides?
The discrepancy between measured and expected oxygen concentration in known standards can be reduced, but not completely, if we reduce the acceleration voltage to 2kV and measure the oxygen concentration only, and metal concentrations will be set as constants. In the table below you can see the normalised (cps/(1nA*1 sec)) oxygen counts, measured on Fe2O3, MgO, Al2O3, SiO2, SnO2 corresponding measured oxygen concentrations (wt.%) and relative deviations from the expected stoichiometric concentrations.
Fe2O3 MgO Al2O3 SiO2 SnO2
wt.% O in standards (stoichiometric) 30.06 39.70 47.08 53.26 21.24
cps/(1nA*1 sec) 2kV 4.7 6.1 6.6 6.6 4.1
6kV 31.4 39.7 43.4 44.1 16.2
15kv 54.2 61.9 64.9 60.5 13.9
(cps/(1nA*1 sec)) / wt.% O 2kV 0.16 0.15 0.14 0.12 0.19
6kV 1.04 1.00 0.92 0.83 0.76
15kv 1.80 1.56 1.38 1.14 0.65
wt.% O (standard is Fe2O3) 2kV 30.63 45.50 49.01 49.56 26.04
6kV 29.63 42.96 48.10 50.71 28.83
15kv 29.59 40.69 47.27 48.34 35.13
delta wt.% O, rel.% 2kV 1.91 14.61 4.11 -6.94 22.65
6kV -1.42 8.23 2.17 -4.79 35.75
15kv -1.54 2.50 0.42 -9.23 65.43
? charge
Fe2O3 was used as a standard for oxygen.
You can see the abnormal behaviour of oxygen intensity on SnO2, as well as huge discrepancies of oxygen determination at any acceleration voltage. The discrepancy for MgO is not so critical, but still questionable. The change of the matrix correction procedure, as well as change the oxygen standard did not help. Oxygen concentration in SnO2 can be measured correctly, if SnO2 is selected as a standard, but in this case the deviation of oxygen measurements for other oxides will be high.
My goal is to develop and experimental set-up such that the oxygen level of a multi-oxide system can be measured accurately. Any advice would be most welcome.
Thank you!