Does anyone know how long it takes for a gas flow counter WDS, from beam on, to start counting?
I've measured C Ka peak counts and absorbed current for 70s at 0.5s intervals on C coated Bi samples. The absorbed current is showing a sharp initial increase in the first 1 - 2s. If this initial increase is due to a local reservoir of C under the beam spot being cracked and deposited (lowering the BSE coefficient under the beam and thus increasing the absorbed current) it isn't being reflected in the x-ray data. Either the absorbed current isn't showing a deposition of C (in which case what the heck is it showing), or the x-ray detector is somehow missing the first couple of seconds, even though the two sets of data are apparently being recorded simultaneously (using the JEOL chart recorder). For normal analyses (i.e. beam stable samples) such a delay wouldn't be an issue and would give the beam time to settle before counting starts, but for beam sensitive samples t=0 counts actually being t=2s counts might be important (some of those volatiles can move really fast in the first few seconds!)