I noticed that a certain lab has a Bruker SDD on one instrument and a Thermo SDD on another . I'm window shopping for an EDS and was wondering how these two compare in their performance, software functionality+ease of use, and just general opinions. Also, what was the ballpark price difference?
My lab has a Bruker SDD on my SX50 and a Thermo SDD on my SX100... but they are at least 5 years old or so.
That said, I don't think you can go wrong with either. They are both excellent detectors and compared to the days of yore when we were limited to 5 nA of beam current on Si(Li) detectors, this is paradise.
The main hardware difference between the two products that concerns EPMA instruments is that Bruker can input WDS ROIs and output EDS ROIs, while Thermo can only input WDS ROIs (for x-ray mapping purposes). Unfortunately neither vendor can quantify the WDS map intensities using this method. However, this will soon become a moot point as both vendors are implementing my suggestions for synchronized EDS-WDS EPMA x-ray mapping methods as described here:
http://probesoftware.com/smf/index.php?topic=400.msg3702#msg3702Thermo is already working on this synchronized mapping method and I expect Bruker will follow suit soon.
As for pricing you'll have to see what you are quoted. With the dollar so strong these days I would expect that Thermo would get you the best deal easily, but it depends obviously.
As for software, I tend to prefer the NSS interface. The Bruker GUI is very sexy but I prefer simplicity. For example, in Thermo NSS one only has to type in the sample "base name" in one place and it gets utilized for all subsequent operations. With Esprit one has to remember to drill down in multiple dialogs to edit the sample name or you end up with the previous sample name. Dale Newbury, Ed Vicenzi and I wrote a letter to Bruker a few years ago with a list of software requests and I think they have implemented some, though not all of our suggestions.
Also the report generation in NSS seems to be a little easier (no need for creating a report template), and NSS outputs directly to PowerPoint where it is easy to edit. Please note that both softwares are constantly changing and so my take on these issues may be out of date...
As for integration with Probe for EPMA both apps are very nicely integrated and will acquire a full EDS spectrum with each WDS analysis and furthermore, EDS elements can be added subsequently and included in the WDS quantification using full standard k-ratios, drift corrections and interference corrections. See here for more details:
http://probesoftware.com/smf/index.php?topic=79.msg286#msg286On a related note, one can also acquire a full CL spectrum (along with the EDS spectrum) with each WDS analysis as described here:
http://probesoftware.com/smf/index.php?topic=42.msg3855#msg3855However, this requires an Ocean Optics spectrometer already connected to the instrument as described. If you are interested in this feature please contact Probe Software for details.