Update: This has been solved, It looks I got victim of Variable Pressure (which was out of order for many years, but got fixed, which I had not knew) mode on SEM. I am leaving below original message as anecdotal story.
Dear Collective Knowledge,
I am not sure that this is right sub-forum for this topic, As I could not found "General SEM Issues".
So I want to analyse the e-beam-extreme sensitive minerals, and SEM equipped with its multiple, large diameter EDS with very close proximity to the sample comes as best combination in the sense of counts for given beam exposure. EPMA-WDS TDI is out of the equation, as there are to many major elements, and TDI works only on the first 5 (with 5 WDS). SEM-multi EDS system makes TDI unnecessary, as it gives more counts with very short exposure, than few minute WDS.
The SEM problem with current measurements
All EPMA's has Faraday cup built-in column. That is a tremendous advantage - it has the constant length of conductor from cup to pico-amperometer. On SEM with stage mounted cup, the current reading will differ depending from where on the stage the cup (or set of standards with built-in cup) is mounted as the current path from cup to pico-amperometer will be different. In my case it is giving up to 5% difference in measured current! (Ouch!). That can propagate such error for standard-based EDS analysis.
Another problem is standard-set built-in faraday cups.
We have some SPI, MAC, etc reference material sets with faraday-cup built in (placed in center or at side of mount).
We polish these standard mounts once a year and do new carbon re-coating.
I guess this caused our built-in faraday cups to be contaminated. On AsB (angularly selective BSE) on SEM I can see some debris inside some of cups, but other cups looks clean. Another concern is that probably carbon had piled up from every coating after mount polishing, as it is impossible to polish-off carbon inside the cup. I guess those faraday-cups are junked at this point. I see that on 20kV faraday cup is giving higher current compared to beam hitting by the hole, albeit the difference is much less than what I see on EPMA when measuring absorbed and beam (in column FC) current. If I set SEM to 10kV I get lower current reading inside the stage-mounted faraday cup than outside (which is illogical). I think that shows that all my FC are contaminated, and I am doomed.
Has anyone a hint how to clean those FC?
Could there be another hidden problem in my SEM setup and how to troubleshoot that?
Would buying separate FC (stand alone, not mounted with standards) would solve my problem?
SEM: Zeiss Sigma (FE)
PS.: I start to understand the idea, why Bruker does not use the FC reading, but calibrates intensities to specified and measured X-ray intensity of material dedicated for that - in that way this problem with variations in FC measurements can be overcome. I think Au piece would perfectly fit for that porpoise as it is the most inactive metal; copper would oxidize, and would influence the counts at low voltages. The trouble is if I want to use DTSA-II which needs the beam current; hmmm... maybe I could simply define the ratio between Au and fictional beam current, and recalculate such net intensities into fictious current equivalent. If I would do standards and unknown with same fictitiously acquired beam current the quantification should work, I guess...