On our SEM we have an SDD EDS detector which is 30 sq. mm. On our EPMA we have a 10 sq. mm. The idea being that the SEM tends to operate at lower beam currents, while the EPMA tends to operate at higher beam currents. And we don't want to saturate the detector on our EPMA when running at high beam currents.
Though it's worth mentioning that we have modified the EDS aperture wheel on our Cameca SX100 to reduce the geometric efficiency of the 10 sq. mm detector for very high beam currents, as discussed here:
https://probesoftware.com/smf/index.php?topic=342.0Though generally we leave it "wide open". It should also be mentioned that utilizing a smaller aperture works just fine when performing standards based EDS analysis, because both the standards and unknowns are measured using the same EDS aperture.
But here's my question: does anyone have a 30 sq. mm SDD detector on their EPMA instrument? If so, at what beam currents (with a fully open aperture) does one see deadtimes increase beyond 60-80%?
That is, is it a good idea to install a 30 sq. mm SDD EDS detector on an EPMA instrument?