Author Topic: Missing peaks from thin film  (Read 8296 times)

Farqhuit

  • Graduate
  • **
  • Posts: 8
Missing peaks from thin film
« on: July 16, 2018, 06:48:00 AM »
We were having trouble identifying features on a sample, including various thin film coatings, so as a last resort we ran EDS mapping to help us get oriented, but suspecting that there would be very little signal from the thin films. We assigned some low energy peaks to Ir, even though some higher energy peaks were absent, as we knew Ir was present as a thin film and no other element matched the energy more closely. We were pleasantly surprised that Ir was accurately mapped (see image). Comparing the peaks we measured to the peaks predicted by Inca (attached), the L peaks up around 10 keV are barely present, and the M peaks down around 2 keV are in the wrong proportions relative to each other. Analysis performed at 12 kV.

The thin iridium film (light area) in the map image, has a thin silver underlayer and both are on a silica substrate. Thicknesses to be determined, probably around 500 nm for the Ir, and much thinner for the silver.

We suspect that the 12 kV beam passed right through the Ir/Ag generating very little signal, and that 1.7 keV X-rays from the Si beneath have excited a fluorescence in the Ir giving a strong Ir peak at around 1.55 keV. All other Ir peaks are from a weak interaction with the electron beam. Does this sound plausible? If not, we'd love to hear what you think is going on. I guess it might be useful to run at 3 kV and see if the relative size of the M peaks changes. Angling the sample might also have an effect.

Disclaimer: our EDS system requires calibration as discussed elsewhere on the forum, but can be persuaded to produce spectra and maps.

Thanks for reading, we look forward to any comments.

JonF

  • Professor
  • ****
  • Posts: 149
Re: Missing peaks from thin film
« Reply #1 on: July 16, 2018, 10:05:28 AM »
Hi Farqhuit,

  I just ran a quick Casino simulation on your case, using a variety of kV and a sequence of 500nm Ir on top of 100nm Ag and then a Si substrate.
According to my simulation (using a suggested Ir density of 22.5 g/cc), the electron beam will actually struggle to get through the Ir at all.

You're seeing a strong Ir M emission. The Ir L emission will be very low intensity due to the low overvoltage being used (Casino suggests the simulated emission depth for Ir L is <50nm) - meaning the electron beam isn't hitting enough Ir atoms before the electron beam has decelerated below the energy it would need to knock out Ir L shell electrons.

The 'ratio' of Ir M lines you're seeing actually looks to me more like an overlap of the Ir M and Si K emission lines. This is slightly puzzling as we shouldn't really be seeing the Si K underneath the Ir and Ag layers, which makes me wonder if the spectrum in your post is actually a sum spectrum of the map you also show, in which case it is seeing the right-hand side Si layer. You mention it is an INCA simulation - I dont have an Oxford instruments EDS but would be interested to hear what settings you put in to simulate that spectra? Does INCA simulate layers? You might be better off using NIST DTSA-II for that?

Rather than dropping the kV, I'd look at increasing it - you're not going to generate layer thicknesses if the electron beam never gets to the other side! Increasing the kV will also increase the likelihood of seeing the Ir L lines, and breaking down in to the Si layer (a small Ag signal may also be visible).
« Last Edit: July 16, 2018, 10:10:36 AM by JonF »

Probeman

  • Emeritus
  • *****
  • Posts: 2839
  • Never sleeps...
    • John Donovan
Re: Missing peaks from thin film
« Reply #2 on: July 16, 2018, 10:10:18 AM »
Hi Jonathan,
You beat me to it!     :)

Just for kicks you might want to post your simulated spectrum.  I think you are probably right about the Ir M and Si K lines overlapping each other.  And maybe also try a calculation at an even lower voltage to see if the possible Si Ka emission line disappears.
john
The only stupid question is the one not asked!

jrminter

  • Professor
  • ****
  • Posts: 72
Re: Missing peaks from thin film
« Reply #3 on: July 16, 2018, 10:46:59 AM »
I know you have had trouble with this system. I have some suggestions that may give you more capabilities until you can get these resolved. This is complementary to the suggestion to use Casino. Casino will let you model. DTSA will let you calibrate and model. If you want to use DTSA to model, let me know because I can supply some scripts. I would at least learn to use DTSA to calibrate your spectra as explained below.

1. Download and install DTSA on your PC (https://www.cstl.nist.gov/div837/837.02/epq/dtsa2/index.html). Create a DTSA detector configuration that matches the parameters of your detector. Eventually you will want versions that match the time constants and other detector parameters for all the configurations you use.

2. Run a high count spectrum of a piece of Cu at an accelerating voltage that gives you decent counts in both the Cu K and L lines. Export this spectrum and the spectra from your thin film sample taken with the same detector parameters as .msa files. The Inca software can do this. Use the Cu spectrum to calibrate your "DTSA detector configuration" using the built-in "Calibration Alien". You could also track the stability of your energy calibration using the DTSA "Quality Control Alien".

3. Then you can use this "calibrated" detector to assign your peak IDs.



 

jrminter

  • Professor
  • ****
  • Posts: 72
Re: Missing peaks from thin film
« Reply #4 on: July 16, 2018, 04:25:24 PM »
I had some time to run some Monte Carlo simulations on Farqhuit's system. I agree with JonF that 500 nm of Ir is hardly "thin". I used DTSA-II Kelvin to do the simulations. Farqhuit described the substrate as "silica". So following JonF, I simulated the following systems at 30 kV: 100 nm Ir and 50 nm Ag, 200 nm Ir 100 nm Ag, 400 nm Ir 200 nm Ag, and  500 nm Ir 200 nm Ag on SiO2. I am attaching a png of the results.

I used a Python (Jython) script that runs with "out of the box" DTSA-II Kelvin. I am attaching the script. You can download that from the DTSA-II website. A couple of notes of caution:

1. Python/Jython uses indentation to signify different levels. Spaces are the default (Pep8). Do not mix tabs and spaces. It will not end well.

2. I modeled the system using a detector configuration file from my FEI Sirion FEG SEM with an Oxford 80 sq. mm. SDD. I really miss having daily access to that system now that I am retired. I included an option to use the "Si(Li)" detector that DTSA-II comes with. Ideally you will configure a customized detector for your system.

3. Depending on the thickness of the layers, 10000 trajectories took between 10 and 18 minutes on my late 2013 MacBook Pro with a SSD.

I don't think a top-down analysis of this system is the right approach. I think it is a case that when you have a hammer, everything looks like a nail. X-ray absorption by Ir and Ag will be a big issue. I spent a lot of time analyzing multi-layer metallic electroplated lines and multi-layered Al printing plates. We had great success by cutting reasonably thick sections (2 micron) Focused Ion Beam lift-out sections. We would mount
these on highly polished carbon planchettes. We put down a very thin rubber cement layer from toluene to get the section to stick to the planchette. We would lightly C coat the specimen and analyze at low voltage.

I would note that all the elements we need to see have transitions below 3 kV. If we analyze the section at 7 kV, all the transitions are well excited. I was able to get very nice maps using this procedure from our Oxford system with the drift correction. My clients were quite pleased with these results.

Farqhuit

  • Graduate
  • **
  • Posts: 8
Re: Missing peaks from thin film
« Reply #5 on: July 17, 2018, 04:25:27 AM »
Thank you all for the detailed responses.

JonF, you are correct, the yellow spectrum is a sum spectrum from the map. I did try to upload the same spectrum without the simulated lines, and also a log scale version, but it doesn't seem to have selected properly. I will try again with this post. I didn't even think to simulate with Casino, I am so used to seeing several microns of penetration even at limited kV, but then I am thinking of much lighter metals (Al etc.).  The INCA simulation was not fed any parameters at all by me, other than what is built in. I am afraid you are crediting me with more sophistication than is justified! There is a feature in INCA that suggests elements when you click on a peak in the spectrum. Selecting an element from the list of suggestions overlays the purple and green lines onto the spectrum. I would think the default settings are assuming a smooth, polished sample and certainly not thin films and layers. i am still finding my way with the technique and the software, with very limited documentation (and limited brains). Between us, we assigned the small peak between the two main Ir M lines as the Si K line. The Ir peak suggested by INCA that coincides with this is one of the smaller ones. It turns out the substrate is coated with an (unspecified) alloy (it is a crystal from a thin film monitor head), so the Si signal looks very similar either side of the edge of the Ir. I will bear in mind that penetration of the beam in Ir is far lower than I assumed, but I also may have overestimated the Ir thickness, less than 100 nm now looks more likely. Interesting point about the shallow L depth and the low overvoltage - noted, thank you.

Probeman, as above, the only simulation I have is the INCA stock one. I am unsure if INCA varies this spectrum to match the microscope settings at the time. I suspect not.

Jrminter, your suggestions regarding DTSA are gratefully received. It looks slightly daunting, but I will give it a go. Previous discussions have highlighted the need for an Oxford engineer to calibrate the system, but this is not going to happen any time soon due to shortage of cash. It would be handy if I could generate what I need to feed back into INCA to stop all the error messages and enable some more peak ID functionality, but I am not sure if DTSA will help me in this. I must pluck up the nerve to hit the calibrate button on INCA and see what it asks for. Your simulations are interesting. Even with very thin layers, the 2 keV Ir peak is a fair bit larger than the 1.5 keV one, whereas my 1.5 keV dominates. Maybe there is some coincident peak contributing. Thanks for the script, I'll try to pick it apart and identify the bits I need to adjust and give it a whirl - my Python knowledge is very thin. Your point about hammer and nails is very true ("Can't you just stick it in the SEM?").

I take from this lots of extra ideas that I an call on when interpreting what we see. I'll be less inclined to make assumptions about beam penetration, and will make more use of the simulation tools (I'd forgotten that Casino gives the M and L depths). If we get any more results that are relevant to the discussion I'll post them here.

Thank you again for your responses, they are much appreciated. I really don't mind admitting that the extent to which I know what I'm talking about is limited. It's great that you are willing to engage anyway.

« Last Edit: July 17, 2018, 08:05:56 AM by Farqhuit »

Farqhuit

  • Graduate
  • **
  • Posts: 8
Re: Missing peaks from thin film
« Reply #6 on: July 17, 2018, 04:32:03 AM »
Missing spectra from original post.

Probeman

  • Emeritus
  • *****
  • Posts: 2839
  • Never sleeps...
    • John Donovan
Re: Missing peaks from thin film
« Reply #7 on: July 17, 2018, 07:34:46 AM »
Probeman, as above, the only simulation I have is the INCA stock one. I am unsure if INCA varies this spectrum to match the microscope settings at the time. I suspect not.

You can download CalcZAF for free from here:

http://probesoftware.com/download/CalcZAF.msi

It automatically installs CalcZAF, Standard and the complete Penepma apps for Monte Carlo simulations of bulk samples, thin films and couples. Here is a short explanation for simulating thin films in Penepma using the GUI in Standard.exe:

http://probesoftware.com/smf/index.php?topic=57.0
The only stupid question is the one not asked!

Farqhuit

  • Graduate
  • **
  • Posts: 8
Re: Missing peaks from thin film
« Reply #8 on: July 17, 2018, 08:53:09 AM »
Thanks Probeman,

Under the Analytical menu in Standard, the lower three entries including PENEPMA are greyed out. Is there another way of accessing the PENEPMA GUI?

Cheers,

Dave

Probeman

  • Emeritus
  • *****
  • Posts: 2839
  • Never sleeps...
    • John Donovan
Re: Missing peaks from thin film
« Reply #9 on: July 17, 2018, 08:56:22 AM »
Thanks Probeman,

Under the Analytical menu in Standard, the lower three entries including PENEPMA are greyed out. Is there another way of accessing the PENEPMA GUI?

Hi Dave,
Simply open any standard database using the File | Open menu. Penepma needs compositions to calculate spectra...

You can use the default standard database (standard.mdb) or a standard database of your own devising.
john
The only stupid question is the one not asked!

jrminter

  • Professor
  • ****
  • Posts: 72
Trilayer simulation with Penepma from CalcZAF
« Reply #10 on: July 18, 2018, 11:21:57 AM »
Probeman, your suggestion to try this with CalcZAF intrigued me.  Farhuit's example was really a tri-layer system with Ir on Ag on silica. As I read the examples on the penepma part of the forum I saw your edits to do a trilayer system from a bilayer simulation where one layer was air. I tried to adapt your method with the input file attached below (Ir-250-on-Ag-150-on-Silica.in). I used standard pure elements from the standards database for Ir and Ag and a SiO2 standard. I renamed the mat files to "Ir.mat", "Ag.mat", and "SiO2.mat" because it made the simulation a bit more readable. I created a trilayer.geo file (tl_250_150nm.geo) based on your example (attached below). I ran the simulation for an hour. I have done this before on bilayers and the noise wasn't as bad (see the png). I thought this capability would complement the simuations I do with DTSA for EDS...

I'd appreciate it if you would take a look at my edits to the '.in' and '.geo' files to see if anything strikes you as wrong. Thought another pair of eyes would be helpful before I set up a 24 hour run to see if the problems is too few counts...

Thanks for all your work on this...

Best regards,
John

jon_wade

  • Professor
  • ****
  • Posts: 82
Re: Missing peaks from thin film
« Reply #11 on: July 19, 2018, 07:12:38 AM »
heres your geometry file at 20kV. 
If I could get fortran to work again (bloody anaconda!) I'd turn it into something that looks like an EDS spectrum, but the lines are right.


a few things about the .in file - I haven't bothered to check, but the initial values for WCC and WCR seem a bit odd ( I set them to 1e3 for the sake of speed).  This was run on Penpema 2015

and the mean take off angle for your eds detector - is it 35 degrees? not that its critical here....

I'll run it again at 12 and quant the lot if you want. 

jrminter

  • Professor
  • ****
  • Posts: 72
Re: Missing peaks from thin film
« Reply #12 on: July 19, 2018, 08:33:59 AM »
Thank you, Jon Wade. I'd like to get this running under penepma12 (that comes with CalcZAF). Would you please attach your editied copies of the .in and .geom file. I made what I thought were the corrections you suggested and penepma12 crashed complaining about my Ir.mat file that penepma12 generated. Once I get something that works I can tweak it.  I have written R code that converts the penepma spectra to .msa format that I can compare to results from DTSA-II.

jon_wade

  • Professor
  • ****
  • Posts: 82
Re: Missing peaks from thin film
« Reply #13 on: July 19, 2018, 09:09:42 AM »
Hi John

Your .geo file looked ok, the .in file I used is attached.  I use the standalone version of Penepma, partly as I run it on a old mac pro, but partly as I find it gives more flexibility.  You may need to edit this - REFLIN may not be supported in the 2012 version and I don't recall how John's version supports multiple detectors.  As you probably know, you can make this run a lot faster if you're only interested in the higher energy lines

I'm running a bunch of sims now at 12 and 15 kV and I'll quant them tomorrow (assuming I remember ).

jrminter

  • Professor
  • ****
  • Posts: 72
Re: Missing peaks from thin film
« Reply #14 on: July 19, 2018, 10:53:54 AM »
Thanks for your help, Jon. I found and fixed my error. I the fixed version running now. Intermediate results look as expected (I have an R function that plots the intermediate spectra when i want.) At some point I would like to script the quantification. I'll have to figure out from which files to extract the results...

Best regards,
John Minter