I've got a good one for ya - here's two PHA scans, this time is the relative verification peaks for PET and LIF, but on the same spectrometer. PET looks a little weird but not obviously horrible, and then LIF has a nice little extra peak on the right side of the primary peak. It wasn't like this in the past, but we don't use LIF on that spectrometer very often (of our two LPET/LLIF spectrometers, this one is less sensitive for LIF than the other and vice versa for PET).

the PHA for Ca Ka/LPET is 1850/897 - not bad but a little low. The PHA for Fe Ka/LIF is 1849/400. The other LLIF crystal we have has a bias about 25 values higher.

I assume I need to adjust the gas flow again, ever so slightly.

Hi Nick,
I don't know what causes the "shelf" on the high energy side of the PHA distribution, but I do see this sometimes on my scans, maybe at high currents and PET crystals, but I'm not sure.

Check the attached images on this post (login to see attachments):

https://probesoftware.com/smf/index.php?topic=217.msg6899#msg6899

and also here, earlier in this topic:

https://probesoftware.com/smf/index.php?topic=31.msg7167#msg7167

Does anyone know what causes this?
john
« Last Edit: October 20, 2018, 03:17:15 PM by Probeman »
The only stupid question is the one not asked!

Probeman

  • Emeritus
  • *****
  • Posts: 2836
  • Never sleeps...
    • John Donovan
Re: PHA issues
« Reply #16 on: October 23, 2018, 04:14:26 PM »
I accidentally left the valve on the P-10 gas cylinder I changed yesterday closed, so the system had no flow overnight. I have no idea if air could have entered the system but I suspect not as all my flow detectors terminate in separate oil filled "bubblers" (as per Cameca).

When I came in this morning and started tuning up with a student, we found the PHA distributions on the high pressure detectors looking like this:



The PHA distribution on the 1 atm detectors looked fine even though the count rates were extremely low:



I don't know if this explains the "shelf" that we sometimes see on some PHA distributions, but I thought it worth sharing.
john
« Last Edit: October 23, 2018, 05:01:15 PM by Probeman »
The only stupid question is the one not asked!

dawncruth

  • Professor
  • ****
  • Posts: 76
Re: PHA issues
« Reply #17 on: August 21, 2019, 09:46:25 AM »
John -
You mention oil filled bubblers for your Camecas. My service tech installed a bubbler for the P-10 and just recommended "mineral spirits" or isopropyl alcohol as the fluid. The alcohol evaporates pretty quickly so I'm looking for an alternative solution. Which oil is in your bubbler?

Dawn

Probeman

  • Emeritus
  • *****
  • Posts: 2836
  • Never sleeps...
    • John Donovan
Re: PHA issues
« Reply #18 on: August 21, 2019, 11:06:31 AM »
Dave Adams mentions the bubbler fluid here (the trick is getting the right keywords!):

https://probesoftware.com/smf/index.php?topic=1109.msg7380#msg7380

john
The only stupid question is the one not asked!

dawncruth

  • Professor
  • ****
  • Posts: 76
Re: PHA issues
« Reply #19 on: August 21, 2019, 11:19:12 AM »
Thanks!

Probeman

  • Emeritus
  • *****
  • Posts: 2836
  • Never sleeps...
    • John Donovan
Re: PHA issues
« Reply #20 on: October 18, 2023, 03:16:19 PM »
My service engineer at UofO just reports this problem has been fixed:

Quote
On the SX100, if capacitor C1 (10nF, 3KV, ceramic disc) on any spectrometer board, starts leaking, users will see the count rate fall to near zero. The problem can be partially overcome by raising the bias voltage to compensate for the droop caused by the bad cap. But then they'll see jumpy peak scans (capacitor leakage is sporadic). On instruments approaching 20 years of age, this failure might become more prevalent.
The only stupid question is the one not asked!

sem-geologist

  • Professor
  • ****
  • Posts: 302
Re: PHA issues
« Reply #21 on: October 20, 2023, 12:48:42 AM »
My service engineer at UofO just reports this problem has been fixed:

Quote
On the SX100, if capacitor C1 (10nF, 3KV, ceramic disc) on any spectrometer board, starts leaking, users will see the count rate fall to near zero. The problem can be partially overcome by raising the bias voltage to compensate for the droop caused by the bad cap. But then they'll see jumpy peak scans (capacitor leakage is sporadic). On instruments approaching 20 years of age, this failure might become more prevalent.

Thanks for this info. I actually see on our SX100 2nd WDS spectrometer the capacitor is visibly cracked - the failure is a bit different from your reported as it clearly increases noise which can be mis-counted as pulses (there is no shorting (which causes leakage), but opening - (which creates more noise)). In this other failure mode the workaround is opposite then...
However, hearing all this I am now more curious: Had your service engineer replaced that capacitor?
I am asking as I would want to do replacement myself, but it is always a bit more comforting to know someone did that before.

zorch

  • Student
  • *
  • Posts: 2
Re: PHA issues
« Reply #22 on: October 20, 2023, 12:49:27 PM »
I did replace leaking C1 on the spectrometer board and the bias voltage at that capacitor rose about 20V. This eliminated the instability seen in the peak scans and allowed us to set the bias back to the old command value of 1830V for PET. While the bad cap was in circuit, we had to increase the bias command to 1850V to get a decent count rate. A cracked capacitor would also result in leakage current because exposed plates would arc past the insulating layers, creating a carbon track that would act much like a resistor. This would lead to a sag in bias voltage at that part of the circuit. I though that an additional leakage current might be taking place inside the P10 tube which carries the bias voltage to the detector. But pulling one end of R9 out of the circuit made no change in the voltage at C1. This result makes it clear that there is no problematic leakage taking place in that bias wire tube.

A very strange 2V rise in the voltage at C1 occurs when I pull one end of C2 (DC blocking signal coupling cap to the AMPTEK A203 IC). This apparent leakage of C2 occurs despite having cleaned the dust off with IPA. Since the AMPTEK part is a charge sensitive amplifier, one would expect some degree of DC voltage offset to occur in this devices output with leakage current from the bias supply. I'll replace C2 soon and let you know what I find. It should be noted that any DC offsets in this part of the circuit do not show up at the PHA or counting system because downstream capacitors C5 and C10 block DC voltages.

Of course all of the temporary circuit modifications described above are done with Electronics Power = OFF!

sem-geologist

  • Professor
  • ****
  • Posts: 302
Re: PHA issues
« Reply #23 on: November 02, 2023, 05:53:14 AM »
zorch,
this is really useful info!
Of course all of the temporary circuit modifications described above are done with Electronics Power = OFF!
So I suppose these soldering work was done with PCB mounted in place? What about measurements, how had You measured bias voltages with PCB being not powered? A hint: You don't need to power whole electronics off, just unplug the cable and wait for some time for capacitors to discharge.

I did replace leaking C1 on the spectrometer board and the bias voltage at that capacitor rose about 20V. This eliminated the instability seen in the peak scans and allowed us to set the bias back to the old command value of 1830V for PET. While the bad cap was in circuit, we had to increase the bias command to 1850V to get a decent count rate. A cracked capacitor would also result in leakage current because exposed plates would arc past the insulating layers, creating a carbon track that would act much like a resistor. This would lead to a sag in bias voltage at that part of the circuit.
Interestingly, We don't see described behavior, maybe it depends from how does the capacitor crack (parallel to layers vs perpendicular to layers). Also If bringing baseline up with gain, properly working system will give same amount of counts as canonical 1830V. I tested that and I can reduce it down by 100V. This does then less gas amplification, and relies on analog gain. Unfortunately with cracked capacitor that is no more possible as gain then catch the noise introduced by cracked capacitor (in our case), and doing that will saturate the counting system with false, noise pulses.

Now, Why I would want to reduce bias, one would ask?
I though that an additional leakage current might be taking place inside the P10 tube which carries the bias voltage to the detector. But pulling one end of R9 out of the circuit made no change in the voltage at C1. This result makes it clear that there is no problematic leakage taking place in that bias wire tube.
Indeed, you thought very well. There should be no perceptible leakage current in not irradiated detector. But in working irradiated mode the chamber will pass current from cathode to the anode which will be proportional to amplification and number of incident X-rays (plus other minor stuff). Actually measuring precise leakage current is near impossible due to cosmic rays (0 to 20 random incidents per second) and depending from the energy Townsend avalanches will have different amplitudes and pulsed current flow through the detector events, which could be missunderstood as leakage. The current will get significant with high pulse rate and high gas amplification (high bias voltage), and on Jeol side that causes significant PHA shift toward lower energies... but on Cameca that effect is less severe. Albeit it can be even more reduced by reducing bias, and increasing gain - the HV bias regulation has more chance to keep the set bias in check even at decently high count rate (i.e. 100kcps).
A very strange 2V rise in the voltage at C1 occurs when I pull one end of C2 (DC blocking signal coupling cap to the AMPTEK A203 IC). This apparent leakage of C2 occurs despite having cleaned the dust off with IPA. Since the AMPTEK part is a charge sensitive amplifier, one would expect some degree of DC voltage offset to occur in this devices output with leakage current from the bias supply. I'll replace C2 soon and let you know what I find. It should be noted that any DC offsets in this part of the circuit do not show up at the PHA or counting system because downstream capacitors C5 and C10 block DC voltages.
In case of Jeol probe it has high voltage cable from spectrometer HV supply somewhere far far away (in this sub- µs lenghting measured events - it is far far away, even if it is just one or two meters away). In case of Cameca probe, HV supply is integrated inside spectrometer as close as possible to where this HV is needed. Additionally to that the HV supply output is monitored and HV supply input is regulated with very fast OPAMP. In case there is any leak current, it is compensated so that output voltage of supply would stay the same proportional to set voltage reference. Thus reducing C2 leakage wont increase the bias current. Every capacitor has leakage current and that is normal and expected. 2V is only ~ 0.1% in this case, will You find better high Voltage capacitor with lower leakage?
Additionally leakage is very well known in X-ray detection preamplifiers, i.e. leakage through FET in EDS detector preamplifiers, and still they work. In case of this leakage, AMPTEK A203 IC will have few drains. input pin is connected with direct and reversed polarity diodes (inside IC), which will lower the leakage down to ~0.6V. The signal further is decoupled with capacitors for feedback and input to Charge sensitive preamplifier part (inside IC). thus that downplayed to 0.6V DC offset is ignored. Finally, differently to EDS preamplifiers, the feedback of CSP has built-in resistor which not only constantly discharges the built-in feedback capacitor, but also downplays further the DC offset or leakage current.